Hi I have been studying about Gage R&R and did some experiments at my internship. and I ran into a question analyzing the results.

I see the general guideline for the AIAG method is that that percentage GR&R should be less than 10% or 30% to be marginal.

My question is, do i apply the same guideline when I'm using the ANOVA method (used MINITAB)? Do I see if my %StudyVariance is smaller than 10% or 30%? or is there some other specific guideline for the ANOVA method?

Thanks!!
 
Elsmar Forum Sponsor

Miner

Forum Moderator
Staff member
Admin
#2
There is no AIAG method. AIAG stands for the Automotive Industry Action Group who published the MSA manual for Ford, GM and Chrysler.

In the MSA manual, there are two optional methods: the Range method and the ANOVA method. Both methods will provide very similar results. The Range method uses simpler math, but the ANOVA method can detect a potential Operator x Part interaction.

The acceptance criteria is the same for both methods.

% Tolerance or P/T Ratio (gages used for inspection):

  • < 10% is ideal;
  • 10 - 20% is acceptable;
  • 20 - 30% is marginal but may acceptable if the characteristic measured is not critical/significant and better gaging is not economical or feasible;
  • > 30% unacceptable
% Study Variation or %GRR (gages used for process control or statistical studies):

  • < 10% is ideal;
  • 10 - 20% is acceptable;
  • 20 - 30% is marginal but may acceptable if the characteristic measured is not critical/significant and better gaging is not economical or feasible.
  • > 30% unacceptable
% Contribution (used to assess each component of measurement variation to prioritize measurement system improvements):

  • < 1% is ideal;
  • 1 - 9% is acceptable;
  • > 9% is unacceptable.
Number of Distinct Categories or ndc (for SPC):

  • > 10 Ideal
  • 5 -10 acceptable
  • < 5 unacceptable
 
Last edited:

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#3
Number of Distinct Categories or ndc (for SPC):

  • > 10 Ideal
  • 5 -10 acceptable
  • < 5 unacceptable
Actually, for SPC it should be 10 minimum calculated as:

1.41[(USL-LSL)/GRR]

That provides 5 distinct catagories on either side of the mean. Anything less than that is borderline attribute. 5 gives you 2.5 on either side of the mean. Now that is a weak chart.
 

Miner

Forum Moderator
Staff member
Admin
#4
Actually, for SPC it should be 10 minimum calculated as:

1.41[(USL-LSL)/GRR]

That provides 5 distinct catagories on either side of the mean. Anything less than that is borderline attribute. 5 gives you 2.5 on either side of the mean. Now that is a weak chart.
I am providing the criteria as stated by AIAG, which is very conservative.

See this article by Donald Wheeler on "An Honest Guage R&R Study". As Wheeler states the use of standard deviations (which are not additive) instead of variances (which are additive) artificially inflates the measurement variation.

Wheeler has also done an article (Good data, bad data and process behavior charts) modeling the effect of varying levels of measurement variation on the effectiveness of SPC. This simulation showed that process control charts were relatively robust to measurement variation.

Given that, your advice is good regarding quality measurement for SPC. However, in some circumstances, it could drive someone to invest in a more expensive gage than may be necessary for most processes. Precision machining may very well be an exception to this. You would know better than I on that.
 

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#5
I understand - oh so well - that it is the AIAG recommendation. And, I would say that for typical gaging ndc=5 may be fine. But, if you are trying to determine runs, other Western Electric rules - or any process activity - with a gage whose distinct categories indicate only 5 statistically valid "buckets", you have some pitiful resolution. I have seen SPC charts that jump in increments so large that they provide poor control limit analysis and overall poor control. It dilutes any predictive power you may have had. Point is, the people setting up the chart did not relate the statistical gage resolution to the resolution they thought they needed. Face it, they may have had 10 lines on their chart between the control limits, but with the gage they had, they only needed 5 lines. It would be a more accurate visual of their level of control.

That said, if (for example - with enough exaggeration to make a point) your control limits are 10% of your specification, your PV represents that 10% of the specification and you have a true normal (or similar mean-central distribution) then ndc=5 could well be tolerable. But, if your distribution needs to be controlled via adjustments for any reason (not just precision machining), then I would recommend buying a more suitable gaging system to get statistically valid 10:1 within the control limits. You need some decent resolution to let "the voice of the process" (to use a Dr. Wheeler term) speak in clear terms - instead of mumbling.

I am also aware of the concern for the calculation for the GRR that Dr. Wheeler has brought up (I have read both articles). What I have not seen is a corrected ndc calculation from Dr. Wheeler from that analysis. Maybe I missed it. That could be a good improvement on the analysis. The concept of statistically significant resolution is critical to measurement system decisions. If there is an improvement, I am anxious to see it. Even so, I would go with a "corrected" (but still based on USL-LSL) ndc>=10 unless atribute (or near attribute) is "good enough".
 

Miner

Forum Moderator
Staff member
Admin
#6
Wheeler recommended the Intraclass Correlation, see "An Honest Guage R&R Study" section IV, Interpreting the Results. This is used to categorize the gage as a 1st through 4th class monitor. The table in Honest Step 11 then defines the capabilities of each class of monitor.
 
Last edited:

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#7
Wheeler recommended the Intraclass Correlation, see "An Honest Guage R&R Study" section IV, Interpreting the Results. This is used to categorize the gage as a 1st through 4th class monitor. The table in Honest Step 11 then defines the capabilities of each class of monitor.
OK, to utilize Wheeler's approach (which starts out by completely tossing out AIAG's approach - if the OP is willing to go there), then it appears:

The value of 10% of (USL-LSL) must fall between the Smallest Effective Measurement Increment is (0.2 PE) and the Largest Effective Measurement Increment (2 PE) - where PE is The Probable Error of a single measurement. That would give you 10 valid "buckets" between your control limits.

I interpret that he has replaced ndc with the area between the Smallest Effective Measurement Increment and the Largest Effective Measurement Increment.

Note: Do you like his Gage R&R logo? Scary, huh?
 
Last edited:

Bev D

Heretical Statistician
Staff member
Super Moderator
#8
Note: Do you like his Gage R&R logo? Scary, huh?
It's only scary becuase he's left out the specific identifyier to AIAG...as you all know I am not a user or fan of the 'traditional' approach perpetuated by AIAG...
 

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#9
It's only scary because he's left out the specific identifier to AIAG...as you all know I am not a user or fan of the 'traditional' approach perpetuated by AIAG...
That is understandable, and Wheeler has done a thorough job in describing his issue with the AIAG approach. I admit, at this point I have not had a chance to do a thorough comparison of the techniques.

But, the worst approach is not looking at the "statistical resolution" of a gage versus its intended output. People will do gage R&R, it will "pass", but it is not related at all to the resolution needed (10:1 minimum to the area between the control limits) to properly utilize SPC. It is a real missing link, especially for AIAG technique (and not too clear in Wheeler's writing, either).

Like most quality tools, gage analysis is not plug, chug and walk away. It needs some thought.
 
Last edited:
Q

quincy

#10
Do not use % study variation to assess whether or not the gage is acceptable. The study variation is an estimate of the width you need to capture 99% of your process measurements (using a multiplier of 5.15). Focus on 2 things. First, does the % R&R fall between 10% and 30% (better yet, is it below 10%)? Second, use the "Number of Distinct Categories" feature in Minitab to see if there is enough resolution in your measurement device to pick up the variation between parts. For example, if you get a value of 1, this tells you the measurement system is poor (it cannot differentiate between parts). A value of 4 or higher indicates that the measurement system is acceptable. A value of 3 says that the measurment device can segregate the parts into 3 unique groups - low, middle, high. Hope this helps.

Regards,

Q
 
Thread starter Similar threads Forum Replies Date
J Variation (using ANOVA method) Required for Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
Rameshwar25 GRR (Gage R&R) Nested Anova Method Example Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 40
C Excel spreadsheet containing Gage R&R with Anova Method wanted Excel .xls Spreadsheet Templates and Tools 22
R Gage R&R - Anova Method - Standard Deviations vs. Variances Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
R Is ANOVA Gage R&R the best method? Reproducibility issues (no inspector effect)? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 13
N % Tolerance - Type 1 study on the gages, then a gage R&R (ANOVA) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
B How to perform a MSA - GRR (Gage R&R) A&R/ANOVA Vs GUM/VDA Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
R P-value and interpretation of ANOVA Gage R & R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 11
L What is the the difference between the Gage R&R Nested Anova and ISO/REMCO N263? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
S Anova Gage R&R Calculations for 4 Operators Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
G GR&R (Gage R&R) Anova versus Xbar Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
I Minitab ANOVA Gage R&R Error (vs. EXCEL ANOVA GRR Calculation) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
L Good Excel .xls sheet for Gage R&R that calculates both Anova and Range Excel .xls Spreadsheet Templates and Tools 9
Rameshwar25 Gage R&R Anova - With Interaction or without Interaction? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
B Study Variation and Discrimination Ratio in Gage R&R ANOVA Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
E Gage R&R or Anova two way? Using Minitab Using Minitab Software 1
S p-Value(s) for Anova (Crossed) GRR (Gage R&R) Studies Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
R Interpreting Gage R&R (2 Way ANOVA) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
E Successful Gage R&R Anova or X/R? Need example Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
S Gage R&R Methods - ANOVA, AIAG, 6sigma Xbar/R, GM - Which to Use Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
C Sample Size of GR&R (Gage R&R) on Nested Anova Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
P Gage R&R Anova Report Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
C GR&R (Gage R&R) Anova spreadsheet is telling us that a gage is not working Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
T Variables Gage R&R - Minitab v14 - F statistic Calculation - ANOVA - Methods validity Using Minitab Software 2
B ANOVA Gage R&R % Tolerance and Preference Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
L Gage R&R TMV Acceptance Criteria Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
S MSA for attribute relation gage Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
E Zero part to part variation - Gage R&R project Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 15
M Gage R&R and right way to measure Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 16
L Gage RandR on automated equipment. IATF 16949 - Automotive Quality Systems Standard 7
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
D Difference between Test Method Validation and Gage R&R Qualification and Validation (including 21 CFR Part 11) 18
R Determining Uncertainty from Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
C Gage Block Wringing General Measurement Device and Calibration Topics 3
B Gage R&R with NDC=1 Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
S Type 1 Gage R&R or something else? Reliability Analysis - Predictions, Testing and Standards 6
G Should I perform Gage R&R only at the beginning of a new project? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
V Thread Plug Gage Pitch GO Diameter out of spec AQL - Acceptable Quality Level 8
G Gage R&R - Where am I going wrong? Part of a FAIR submission (Aerospace) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
R Gage R&R Excel templates Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
K Gage R&R with more than 3 appraisers Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
Anerol C Gage R&R Template AIAG 4th Edition IATF 16949 - Automotive Quality Systems Standard 3
T Attributes SPC study - Attributive control (Go gage) Statistical Analysis Tools, Techniques and SPC 5
S Capability or Gage R&R Study for Leak Tester? Reliability Analysis - Predictions, Testing and Standards 15
M Definitive answer on Type 1 vs Type 2 vs Type 3 Gage Study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
V Gage Management and Gage R&R Software General Measurement Device and Calibration Topics 1
D Gage R&R Study on Load Cells - Large chemical blending tanks IATF 16949 - Automotive Quality Systems Standard 1
B Minitab Type 1 Gage Study on True Position Question Measurement Uncertainty (MU) 1

Similar threads

Top Bottom