Good day,
In regard to %GRR calculation:
when should I apply " 100x(GRR/ (Spec/6)) " --> for product control ?
when should I apply " 100x(GRR/TV)" ---->for SPC control ?
I don't get the point of MSA remark.

Would you pls clarify it for me ?
Tks
:bonk:
 
Elsmar Forum Sponsor

Miner

Forum Moderator
Staff member
Admin
#2
when should I apply " 100x(GRR/ (Spec/6)) " --> for product control ?
Use this formula when you use the gage for inspection. That is anywhere within the specification is acceptable.


when should I apply " 100x(GRR/TV)" ---->for SPC control?
Use this formula when you want to use the gage for Statistical Process Control to control the process to a target value.
 
T

tenaiwa

#3
Miner:
Thank you for your replying
Furtherly,I want to consult about sampling method under actual condition that
<1>Product spec is 0 ~ 10 (g)
<2>In fact, actual performance is indicated as pic attach (arond 90% locate at 0 ~3 g)

Should I take 10 samples randomly which means most of them located at 0 ~ 3 g ?
Or
Should I take 10 samples to cover 0 ~ 10 g as much as possible ?
:confused:

Tks and regards
 

Attachments

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#4
If you are going to use the gage for SPC, I recommend calculating the ndc as a function of the control limits:

ndc (cl)= (UCL-LCL)/GRR

And it should be>10

:cool:
 

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#5
Miner:
Thank you for your replying
Furtherly,I want to consult about sampling method under actual condition that
<1>Product spec is 0 ~ 10 (g)
<2>In fact, actual performance is indicated as pic attach (arond 90% locate at 0 ~3 g)

Should I take 10 samples randomly which means most of them located at 0 ~ 3 g ?
Or
Should I take 10 samples to cover 0 ~ 10 g as much as possible ?
:confused:

Tks and regards
Take 10 samples to cover 0 ~ 10 g as much as possible - over 10 if it is possible to occur. The point of the exercise is to see if the gage responds acceptably over the range it is going to be used.
 

Miner

Forum Moderator
Staff member
Admin
#6
Take 10 samples to cover 0 ~ 10 g as much as possible - over 10 if it is possible to occur. The point of the exercise is to see if the gage responds acceptably over the range it is going to be used.
The answer is "It depends."

If you will be using the gage for inspection purposes, bobdoering's suggestion is appropriate. However, if you plan to use the gage for process control (i.e., SPC), your samples need to represent the non-normal distribution that you show. Hopefully random samples will provide this representation. It is important that the variation of your samples is similar to the variation of your process.
 
T

tenaiwa

#7
Tks for replying from both of you.I see.
B.T.W
Does this sampling rules also apply during Cgk calculation ?
"For inspection-->sampling should cover spec.range"
"For SPC study-->sampling should present situation of actual process "


Auditor from Volkswagen raised Cgk study as a mondatory which I never met before.
Is there any thread where Cgk is described ?

Many thanks
 

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#8
Tks for replying from both of you.I see.
B.T.W
Does this sampling rules also apply during Cgk calculation ?
"For inspection-->sampling should cover spec.range"
"For SPC study-->sampling should present situation of actual process "


Auditor from Volkswagen raised Cgk study as a mondatory which I never met before.
Is there any thread where Cgk is described ?

Many thanks
There are, but you might also want to check this out:

Minitab: Rationale and Methodology for the Type 1 Gage Study
 

Bev D

Heretical Statistician
Staff member
Super Moderator
#9
If you are going to use the gage for SPC, I recommend calculating the ndc as a function of the control limits:

ndc (cl)= (UCL-LCL)/GRR

And it should be>10

:cool:
Bob are you sure?
That isn't the formula for NDC. plus control limits are - typically - limits on subgroup averages and measurement error is calcualted in terms of individual values - the two standard deviations are very different in range as well as meaning. (not to mention the entire math on standard deviations thing...)
 

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#10
Bob are you sure?
That isn't the formula for NDC. plus control limits are - typically - limits on subgroup averages and measurement error is calculated in terms of individual values - the two standard deviations are very different in range as well as meaning. (not to mention the entire math on standard deviations thing...)
True, that isn't the regular formula for NDC. And, you might have caught that I did neglect to put in the factor 1.41. My bad. It was early...

So, we should have ndc (cl)= [1.41(UCL-LCL)]/GRR

That is why I put the (cl) next to it - to delineate it from the standard ndc formula (it would have been a subscript).

Might be thinking this thing through a little too hard, as far as the standard deviations and such. NDC is the number of non-overlapping 97% confidence intervals that will span the expected product variation. But, I do not care about the range of product variation in SPC, I am concerned about the number of non-overlapping 97% confidence intervals that will span the expected product variation in control - which is between the control limits. If you can get 10 categories to fit within the control limits, then you have enough statistically significant divisions to adequately detect variation in that range. The calculation of control limits itself is not the issue - the fact that they have become a physical target range, is.

:cool:
 
Thread starter Similar threads Forum Replies Date
kedarg6500 Gage R&R - Is my sheet calculation wrong? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 21
G Gage R&R Calculation Spreadsheets - Which one here is valid? Quality Tools, Improvement and Analysis 4
I Minitab ANOVA Gage R&R Error (vs. EXCEL ANOVA GRR Calculation) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
A GR&R (Gage R&R) Calculation Excel .xls Spreadsheet Excel .xls Spreadsheet Templates and Tools 6
S Should we check the normality of measurements when the calculation of Gage RR is done Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
K During Gage R&R, is calculation of ndc value required? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 24
P Gage R&R (Repeatability and Reproducability): What are the calculation restrictions Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
T Variables Gage R&R - Minitab v14 - F statistic Calculation - ANOVA - Methods validity Using Minitab Software 2
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
L Gage R&R TMV Acceptance Criteria Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
S MSA for attribute relation gage Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
E Zero part to part variation - Gage R&R project Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 15
M Gage R&R and right way to measure Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 16
L Gage RandR on automated equipment. IATF 16949 - Automotive Quality Systems Standard 7
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
D Difference between Test Method Validation and Gage R&R Qualification and Validation (including 21 CFR Part 11) 18
R Determining Uncertainty from Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
C Gage Block Wringing General Measurement Device and Calibration Topics 3
B Gage R&R with NDC=1 Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
S Type 1 Gage R&R or something else? Reliability Analysis - Predictions, Testing and Standards 6
G Should I perform Gage R&R only at the beginning of a new project? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
V Thread Plug Gage Pitch GO Diameter out of spec AQL - Acceptable Quality Level 8
G Gage R&R - Where am I going wrong? Part of a FAIR submission (Aerospace) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
R Gage R&R Excel templates Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
K Gage R&R with more than 3 appraisers Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
Anerol C Gage R&R Template AIAG 4th Edition IATF 16949 - Automotive Quality Systems Standard 3
T Attributes SPC study - Attributive control (Go gage) Statistical Analysis Tools, Techniques and SPC 5
S Capability or Gage R&R Study for Leak Tester? Reliability Analysis - Predictions, Testing and Standards 15
N % Tolerance - Type 1 study on the gages, then a gage R&R (ANOVA) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
M Definitive answer on Type 1 vs Type 2 vs Type 3 Gage Study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
V Gage Management and Gage R&R Software General Measurement Device and Calibration Topics 1
D Gage R&R Study on Load Cells - Large chemical blending tanks IATF 16949 - Automotive Quality Systems Standard 1
B Minitab Type 1 Gage Study on True Position Question Measurement Uncertainty (MU) 1
S Gage R&R on alloy Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
8 Need Help - Runout - Function Gage Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
L Gage R&R studies for identical Devices - Need to confirm the requirement to perform them Reliability Analysis - Predictions, Testing and Standards 2
G Stab Pin Size For Check Gage Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
E Functional Gage for TOP applied at RFS Design and Development of Products and Processes 5
F No reproducibility Error in Gage R&R? Reliability Analysis - Predictions, Testing and Standards 4
M Template for Attribute Gage R & R wanted Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
G Warning about Deltronic Gage Pins - The "Certification of Accuracy" document General Measurement Device and Calibration Topics 13
A Gage type 1 study on CMM Capability, Accuracy and Stability - Processes, Machines, etc. 2
Q Gage R&R for Instant Measurement Machine Manufacturing and Related Processes 6
M Gage R&R with large inter part variance - Torque Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
D Do we need normal data for gage r&r studies? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
D Do I need part variation while doing Destructive Variable Gage R&R MSA study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
S Gage R&R Standard Deviations Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
J Informational Gage R&R: Swapping operators Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
S Gage R & R for large amount of inspections IATF 16949 - Automotive Quality Systems Standard 10

Similar threads

Top Bottom