Gage R&R lite? This form only uses two operators measuring five parts one time

Q

Qaware

#1
I recently came across a form for Gage R&R which I feel quite skeptic of. This form only uses two operators measuring five parts one time per operator. R-mean is then calculated, by using differences between operators, and multiplied with a constant=4.33, to get R&R-value. I think this looks strange. Can anyone help me understand???
 
Last edited by a moderator:
Elsmar Forum Sponsor
A

Atul Khandekar

#2
I think this is the 'Short Method' or "Range Method' given in the MSA manual. The Short method provides a quick approximation or the overall picture of measurement system variability. The measurement system variation is not decomposed into repeatability and reproducibility.
 
Q

Qaware

#3
constant?

Thank you for your answer, I still think that the amount of data is a little thin. Where does the constant come from, and how is it derived? How will it change if I would use the same method but added one operator?
 
A

Atul Khandekar

#4
Well, this is just a quick approximation. In the words of the MSA manual:

"This approach has the potential to detect an unacceptable measurement system (i.e. GRR > 30%) 80% of the time with a sample size of 5 and 90% of the time with a sample size of 10."

If you have the MSA manual, you'll find the factors in the d2* table depending upon the number of operators and number of parts used in the study.
 
Q

Qaware

#5
Thank you

Thank you very much for that answer, helps me alot. I don't have the MSA manual you are referring to. Who has published it?
_______________________________________________________________

I think I found it myself. MSA 3rd edition- published by AIAG. Correct?

Appreciate your help. :thanx:
 
Last edited by a moderator:
A

Atul Khandekar

#8
Casana said:
Atul,

May I ask where in the MSA does it define the 'Range Method"?
Ana,
This method is described on Page 97,98 of the AIAG MSA Manual- 3rd edition.
AIAG recommends GRR% of Process StDev.

Also see the section 8.3 of this Daimlerchrysler document:
https://gsp.extra.daimlerchrysler.com/mfg/amedd/gages/index.htm
This document recommends GRR% of Tolerance. I think this is the reference Qaware is talking about in the first post.

Hope this helps.
 

Jim Wynne

Staff member
Admin
#9
Atul Khandekar said:
Ana,
This method is described on Page 97,98 of the AIAG MSA Manual- 3rd edition.
AIAG recommends GRR% of Process StDev.

Also see the section 8.3 of this Daimlerchrysler document:
https://gsp.extra.daimlerchrysler.com/mfg/amedd/gages/index.htm
This document recommends GRR% of Tolerance. I think this is the reference Qaware is talking about in the first post.

Hope this helps.
I think it's important to stress the difference between process variation and within-study variation; imo the former is almost always preferable, if process variation is reliably known when the study is done.
Whether % of tolerance or % of variation (or both) is used is a matter of choice (that might, as Atul suggests, be dictated by the customer) but if you choose one or the other, you should be able to tell why.
 
A

Atul Khandekar

#10
JSW05 said:
I think it's important to stress the difference between process variation and within-study variation; imo the former is almost always preferable, if process variation is reliably known when the study is done.
:agree:
JSW05 said:
Whether % of tolerance or % of variation (or both) is used is a matter of choice (that might, as Atul suggests, be dictated by the customer) but if you choose one or the other, you should be able to tell why.
Typically based on what the measurement system is being used for: Process control/improvement or inspection.
 
Thread starter Similar threads Forum Replies Date
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
R Determining Uncertainty from Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
C Gage Block Wringing General Measurement Device and Calibration Topics 3
B Gage R&R with NDC=1 Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
S Type 1 Gage R&R or something else? Reliability Analysis - Predictions, Testing and Standards 6
G Should I perform Gage R&R only at the beginning of a new project? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
V Thread Plug Gage Pitch GO Diameter out of spec AQL - Acceptable Quality Level 8
G Gage R&R - Where am I going wrong? Part of a FAIR submission (Aerospace) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
R Gage R&R Excel templates Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
K Gage R&R with more than 3 appraisers Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
Anerol C Gage R&R Template AIAG 4th Edition IATF 16949 - Automotive Quality Systems Standard 3
T Attributes SPC study - Attributive control (Go gage) Statistical Analysis Tools, Techniques and SPC 5
S Capability or Gage R&R Study for Leak Tester? Reliability Analysis - Predictions, Testing and Standards 15
N % Tolerance - Type 1 study on the gages, then a gage R&R (ANOVA) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
M Definitive answer on Type 1 vs Type 2 vs Type 3 Gage Study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
V Gage Management and Gage R&R Software General Measurement Device and Calibration Topics 1
D Gage R&R Study on Load Cells - Large chemical blending tanks IATF 16949 - Automotive Quality Systems Standard 1
B Minitab Type 1 Gage Study on True Position Question Measurement Uncertainty (MU) 1
S Gage R&R on alloy Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
8 Need Help - Runout - Function Gage Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
L Gage R&R studies for identical Devices - Need to confirm the requirement to perform them Reliability Analysis - Predictions, Testing and Standards 2
G Stab Pin Size For Check Gage Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
E Functional Gage for TOP applied at RFS Design and Development of Products and Processes 5
F No reproducibility Error in Gage R&R? Reliability Analysis - Predictions, Testing and Standards 4
M Template for Attribute Gage R & R wanted Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
G Warning about Deltronic Gage Pins - The "Certification of Accuracy" document General Measurement Device and Calibration Topics 13
A Gage type 1 study on CMM Capability, Accuracy and Stability - Processes, Machines, etc. 2
Q Gage R&R for Instant Measurement Machine Manufacturing and Related Processes 6
M Gage R&R with large inter part variance - Torque Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
D Do we need normal data for gage r&r studies? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
D Do I need part variation while doing Destructive Variable Gage R&R MSA study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
S Gage R&R Standard Deviations Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
J Informational Gage R&R: Swapping operators Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
S Gage R & R for large amount of inspections IATF 16949 - Automotive Quality Systems Standard 10
S Ring Gage Thermal Expansion General Measurement Device and Calibration Topics 4
L Gage R&R for test precision or uncertainty prediction? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
S Minor diameter of threaded ring gage General Measurement Device and Calibration Topics 1
A Does Torque Gage Need Validation? Reliability Analysis - Predictions, Testing and Standards 4
J Ignoring a Failed Gage R&R and upcoming external audit IATF 16949 - Automotive Quality Systems Standard 22
R Scaling a Gage R&R - 2 holding fixtures each with 10 nests Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
B Gage R&R Acceptable (10-30%), deduct Total Variation from Tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
apestate Inexpensive Gage Blocks for small aerospace startup shop General Measurement Device and Calibration Topics 1
F Position tolerance check gage manufacturing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
B Informational Expanded Gage R&R Analysis using Minitab Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 18
S Interesting Discussion Threaded ring gage calibration and tendonitis in the elbows General Measurement Device and Calibration Topics 8
A Federal 700B-26 Gage Block Comparator is not reading properly General Measurement Device and Calibration Topics 0
O Gage Simulator for a Pilot at RFS Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
J Gage R&R without tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
B Variable Gage R&R - Two operators instead of three Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4

Similar threads

Top Bottom