Guys Hello

I have a question regarding process validation for Gamma Irradiation.
one part of the validation includes dose verification.
this is done by initially calculate an average Bioburden, then apply the calculated dose to the items and then perform a sterility test.
now, lets assume I manufacture my medical device in 2 different facilities (i.e. I have a second source) should I perform the Gamma Sterilization Process Validation twice? one time for products manufactured in facility 1 and another time for products manufactured in facility 2?
it does make some sense since part of the Gamma irradiation process validation includes calculation of the Average Bioburden level, and the Bioburden level is also depends on the specific facility were the products are manufactured.

thanks!!:thanx:
 
Elsmar Forum Sponsor

shimonv

Trusted Information Resource
#2
Hi Alon,
Yes, you should repeat the process validation for the second site.
You must demonstrate that the calculated dose is sufficient to sterilize items coming from the second site which, quite likely, will have a somewhat different Bioburden level. It's difficult to rationalize your way out of it.
Also down the road, as changes occur and not necessarily in both sites simultaneously, you will need to repeat the validation process.

-Shimon
 
Thread starter Similar threads Forum Replies Date
H Gamma Irradiation - dose audit and source replenishment Other Medical Device Related Standards 13
A Gamma Validation - What should a Third Party Irradiation/Gamma Company Supply EU Medical Device Regulations 2
L ETO vs. Gamma Irradiation Sterilization of DBM (Demineralized Bone Matrix) Other Medical Device and Orthopedic Related Topics 3
M Eto Certificate of Sterilization (vs. gamma irradiation) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
B Gamma Sterilization Other Medical Device Related Standards 1
B Gamma Quarterly Audit Medical Device and FDA Regulations and Standards News 1
N Looking for Gamma sterilization facility in China ISO 13485:2016 - Medical Device Quality Management Systems 4
E Gamma Sterilization Product Update Manufacturing and Related Processes 2
Q Use of parts in gamma sterilization validation - Reprocessing implants EU Medical Device Regulations 1
R ISO 11137 Gamma sterilization validation, dose setting, dose audit Other Medical Device Related Standards 1
D Gamma Sterilization of Coin Batteries (Primary Batteries) Other Medical Device Related Standards 3
N Can we do a Partial Gamma Validation on a Similar Product (Syringes) ISO 13485:2016 - Medical Device Quality Management Systems 6
N Gamma Validation vs. Quarterly Dose Audit Differences ISO 13485:2016 - Medical Device Quality Management Systems 3
K Gamma Ray Sterilization Quarterly Dose Auditing Supplier Requirements ISO 13485:2016 - Medical Device Quality Management Systems 5
W Sterile Product or Not Sterile - The device has "Gamma Irradiated" printed on it 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
A Moving from Ethylene Oxide to Gamma Sterilization - What is required? Other Medical Device Related Standards 10
K Sterilization for Hypodermic Needle - EtO or Gamma? ISO 13485:2016 - Medical Device Quality Management Systems 1
M Does anyone have an electronic Gamma and Beta Table Statistical Analysis Tools, Techniques and SPC 4
D Sterilization Shelf Life Testing for a part that is gamma sterilized EU Medical Device Regulations 1
S Gamma Sterilization: Change in Bulk Density vs. Revalidation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
M E-beam Sterilization to replace Gamma - Any information will be helpful Other US Medical Device Regulations 2
I Quarterly Gamma Dose Audit Question Other Medical Device Related Standards 4
M Radiation (Gamma or E-beam) Dose Audit SOP - Quarterly dose audits based on ISO 11137 Other Medical Device Related Standards 8
F Non-Normal Distribution vs. Gamma Distribution Statistical Analysis Tools, Techniques and SPC 18
E ISO 11137 gamma sterilized device dose audit - Lead time? Other ISO and International Standards and European Regulations 7
R Change Control Required? Changes during the product gamma phase IATF 16949 - Automotive Quality Systems Standard 3
B Weibull Model - Reliability Analysis - Alpha, Beta & gamma value formulas Reliability Analysis - Predictions, Testing and Standards 3
M Gamma Sterility Validation - ISO 11137-2, 4.2 vs. FDA Requirements ISO 13485:2016 - Medical Device Quality Management Systems 6
Ajit Basrur Audit Checklist for Contract Gamma and ETO Sterilization needed ISO 13485:2016 - Medical Device Quality Management Systems 3
Q EtO vs. Gamma Sterilization - Stainless Steel and Titanium Orthopaedic Implants ISO 13485:2016 - Medical Device Quality Management Systems 7
S Gamma Sterilization Validation Documents - 13485:2003 section 7.5.1.2.1 ISO 13485:2016 - Medical Device Quality Management Systems 5
F Irradiation Sterilization Validation Maximum Dose - VDmax 25 method (ISO 11137) EU Medical Device Regulations 2
P ISO 10993 - Sterile (Irradiation) Polypropylene Syringe Question Other Medical Device Related Standards 6
R Irradiation Sterilization Dose Setting and Dose Audit Flow Charts ISO 13485:2016 - Medical Device Quality Management Systems 0
I Quarterly to semiannual audit of irradiation sterilization Qualification and Validation (including 21 CFR Part 11) 2
S Effects of Irradiation Sterilization on Medical Electronics - Li Coin Cell Batteries IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
B How to satisfy clause 5.7.1.5 process validation for valve production API 6D Oil and Gas Industry Standards and Regulations 13
C Change Management Process ISO 14971 - Medical Device Risk Management 8
B Documents for production process IATF 16949 - Automotive Quality Systems Standard 8
J Process Capability - Determining the process capability of certain equipment Lean in Manufacturing and Service Industries 6
Z E-Coating process International Specifications? FMEA and Control Plans 3
W CQI-15 for new welding process IATF 16949 - Automotive Quality Systems Standard 3
D Using electronic lab notebooks in the design control process ISO 13485:2016 - Medical Device Quality Management Systems 3
M Process & Product Characteristic in Control Plan FMEA and Control Plans 17
S Consolidated Analysis Report for Moulding Process Manufacturing and Related Processes 1
P ISO 11607-2 Process Specification Other Medical Device Related Standards 1
Q Best off presentation of process flows Process Maps, Process Mapping and Turtle Diagrams 16
Q Logistics process flow ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
M Process FMEA for Laser Cutting/Press Brake/Welding FMEA and Control Plans 2
H Control of Change i.e. process and WI IATF 16949 - Automotive Quality Systems Standard 9

Similar threads

Top Bottom