Gamma Sterilization Product Update

#1
I am updating a gamma sterilized product (carton with sterile pouch and IFU inside) to increase the size of the IFU by 2x. The original validation did not account for IFU size increases. Per ISO 11137-2, bioburden is the driver for the sterilization dose, not density or other factors. Since the bioburden in the sterile barrier did not change, does it make sense that the validation is not impacted and only the dose map would need to be verified due to the density increase?
 
Elsmar Forum Sponsor
#2
IFU bioburden will be double, so average bioburden will be increased. This changes verificatin dose. For methode 1, sterilization dose goes a bit higher.
 

planB

Super Moderator
#3
IFU bioburden might be only relevant if the IFU has to be terminally sterilised with an SAL of 10exp(6), e.g. is co-packaged into the sterile barrier system together with the actual medical device. The original request ("carton with sterile pouch and IFU inside") indicates that the IFU is not co-packed. Thus, only the density of the load presented to the irradiation cycle may change, and verification of the (continued) validity of the existing dose mapping may be appropriate. In case the dose-mapping experiment shows that the load density has a significant impact on the dose distribution in the load, further re-validation activities might be warranted according to ISO 11137-1, section 12.5.2, Assessment of change in product.

HTH,
 
Thread starter Similar threads Forum Replies Date
N Looking for Gamma sterilization facility in China ISO 13485:2016 - Medical Device Quality Management Systems 4
Q Use of parts in gamma sterilization validation - Reprocessing implants EU Medical Device Regulations 1
R ISO 11137 Gamma sterilization validation, dose setting, dose audit Other Medical Device Related Standards 1
D Gamma Sterilization of Coin Batteries (Primary Batteries) Other Medical Device Related Standards 3
K Gamma Ray Sterilization Quarterly Dose Auditing Supplier Requirements ISO 13485:2016 - Medical Device Quality Management Systems 5
A Moving from Ethylene Oxide to Gamma Sterilization - What is required? Other Medical Device Related Standards 10
K Sterilization for Hypodermic Needle - EtO or Gamma? ISO 13485:2016 - Medical Device Quality Management Systems 1
L ETO vs. Gamma Irradiation Sterilization of DBM (Demineralized Bone Matrix) Other Medical Device and Orthopedic Related Topics 3
D Sterilization Shelf Life Testing for a part that is gamma sterilized EU Medical Device Regulations 1
S Gamma Sterilization: Change in Bulk Density vs. Revalidation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
M Eto Certificate of Sterilization (vs. gamma irradiation) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
M E-beam Sterilization to replace Gamma - Any information will be helpful Other US Medical Device Regulations 2
Ajit Basrur Audit Checklist for Contract Gamma and ETO Sterilization needed ISO 13485:2016 - Medical Device Quality Management Systems 3
Q EtO vs. Gamma Sterilization - Stainless Steel and Titanium Orthopaedic Implants ISO 13485:2016 - Medical Device Quality Management Systems 7
S Gamma Sterilization Validation Documents - 13485:2003 section 7.5.1.2.1 ISO 13485:2016 - Medical Device Quality Management Systems 5
H Gamma Irradiation - dose audit and source replenishment Other Medical Device Related Standards 13
alonFAI Gamma Irradiation Process Validation Help Other Medical Device Related Standards 2
N Can we do a Partial Gamma Validation on a Similar Product (Syringes) ISO 13485:2016 - Medical Device Quality Management Systems 6
N Gamma Validation vs. Quarterly Dose Audit Differences ISO 13485:2016 - Medical Device Quality Management Systems 3
W Sterile Product or Not Sterile - The device has "Gamma Irradiated" printed on it 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
A Gamma Validation - What should a Third Party Irradiation/Gamma Company Supply EU Medical Device Regulations 2
M Does anyone have an electronic Gamma and Beta Table Statistical Analysis Tools, Techniques and SPC 4
I Quarterly Gamma Dose Audit Question Other Medical Device Related Standards 4
M Radiation (Gamma or E-beam) Dose Audit SOP - Quarterly dose audits based on ISO 11137 Other Medical Device Related Standards 8
F Non-Normal Distribution vs. Gamma Distribution Statistical Analysis Tools, Techniques and SPC 18
E ISO 11137 gamma sterilized device dose audit - Lead time? Other ISO and International Standards and European Regulations 7
R Change Control Required? Changes during the product gamma phase IATF 16949 - Automotive Quality Systems Standard 3
B Weibull Model - Reliability Analysis - Alpha, Beta & gamma value formulas Reliability Analysis - Predictions, Testing and Standards 3
M Gamma Sterility Validation - ISO 11137-2, 4.2 vs. FDA Requirements ISO 13485:2016 - Medical Device Quality Management Systems 6
chris1price Sterilization using beta radiation Other Medical Device Related Standards 2
A ISO 13485 Sterilization Clause Applicability ISO 13485:2016 - Medical Device Quality Management Systems 7
S Orthopedic Implants (Knee system) Sterilization Indicators Manufacturing and Related Processes 6
chris1price Contract Sterilization Service Providers ISO 13485:2016 - Medical Device Quality Management Systems 5
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
S Sterilization validation after changing sterilization process provider Qualification and Validation (including 21 CFR Part 11) 3
B Sterilization Validation Plan Other Medical Device Related Standards 3
R Scientific Operating Procedures For Sterilization Practices in India PSA/2012/1 Other Medical Device Regulations World-Wide 0
I Release checklist EO sterilization process Medical Device and FDA Regulations and Standards News 3
MrTetris Are GLP required for testing cytotoxicity and soil remaining after sterilization of MD? Other Medical Device Related Standards 8
Z EN 868-8 "required tests" - Rigid sterilization containers Other Medical Device Related Standards 0
Judy Abbott Increase the maximum allowable dose for E-beam sterilization Reliability Analysis - Predictions, Testing and Standards 2
H EO Sterilization Validation - Sterility Testing and Load Configuration Other Medical Device Related Standards 1
A Reprocessing, Cleaning, Disinfection and Sterilization Validation EU Medical Device Regulations 4
W EO sterilization and pallets Other Medical Device and Orthopedic Related Topics 4
M Informational Device Shortages Update – The US FDA Announces Two New Innovation Challenges on Device Sterilization Medical Device and FDA Regulations and Standards News 0
W EO sterilization - what about the pallet? Other Medical Device Related Standards 1
B Sterilization of a custom medical device Other Medical Device and Orthopedic Related Topics 1
M Informational US FDA – Device Shortages Update: Challenges to encourage the development of new approaches to device sterilization Medical Device and FDA Regulations and Standards News 0
xcanals_tecno-med.es Qualification of sterilization indicators as medical devices with the EU MDR 2017/745 EU Medical Device Regulations 7
G Pyrogen Test - Sterilization Validation - MDR 2017/745 Other ISO and International Standards and European Regulations 6

Similar threads

Top Bottom