I am not an MSA guru. I am not familier with Type 1 or Type 2 testing. In my company we do, however, have the issue of automatic versus manual gages (or measurement systems). In the case of automatice gages, our practice is that the operator is not a variable (as the tool is automated, the operator plays no part in measurement variation). So repeatability is done the same way whether manual or automatice. But reproducibility on automatic eqiupment is not between operators, but between shifts or days or equipment shut downs. Rather than multiple operators, the reproducibility portion would include sets of measurements taken on different shifts, or different times of day, etc.

------------------
 
Elsmar Forum Sponsor
M

mbruner

#2
I have not heard of type 1, type 2 in reference to GRR studies. The MSA manual specifically states that reproducibility is variation in the average measurements made by different appraisers. It also states there are some measurement systems where the reproducibility is considered negligible.We have interpreted this by doing a GRR study with one appraiser (a GR study if you will)for our automated equip. It kind of makes sense but I can't help that.
 
R

RNicholl

#3
R&R studies for Automatic equipment

I am currently using R&R studies recommended in the MSA manual for both automatic and manual gauges. However I have recently been told that I should be using Type 1 studies for automatic gauges and type 2 for manual gauges. The MSA manual makes no refrence to this, can anyone shed any light on the subject?

Thanks

Rob Nicholl
 
B

Batman

#4
Yes, with "automatic" measurement equipment you usually only determine repeatability.

However, be sure there is no operator influence available. If the "automatic" device requires different operators to "fixture" parts - as in a laser micrometer, CMM or Vision type system, or the operators must set-up and standardize a device prior to measuring parts, it should be considered that there are other potential sources of variation.
 
Thread starter Similar threads Forum Replies Date
B NDC - Gauge R&R Studies with results where NDC = 1 or 0 Manufacturing and Related Processes 1
M Normalising Data for Gauge R&R Studies with Minitab Using Minitab Software 14
B Gauge to Gauge (Gage) Studies Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 17
C Gauge R&R Studies Bad, but Process has a Cmk of over 7 IATF 16949 - Automotive Quality Systems Standard 8
R "Gauge (Gage) R&R Studies per MSA Manual" - Customer Audit Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 13
F Gauge R&R Studies on a CMM (Coordinate Measuring Machine) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
G Seeking a good reference on gauge R&R studies Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
F Calibrating Taper Sleev/Plug Gauge General Measurement Device and Calibration Topics 0
I How many gauges in a Go/No Go Gauge? General Measurement Device and Calibration Topics 7
J Standard used to determine if check fixture gauge is acceptable Reliability Analysis - Predictions, Testing and Standards 2
J Designing new gauge tracking software IATF 16949 - Automotive Quality Systems Standard 4
R Gauge R&R on pre-set torque wrench Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
C Aluminum galling on thread plug gauge Manufacturing and Related Processes 10
J Gauge R&R on a torque measurement Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 11
K I’m looking for a square probe for my linear height gauge Manufacturing and Related Processes 1
A Gauge Study for Small Lots Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
G Supplier Checking With Gauge Pin Problem Solving, Root Cause Fault and Failure Analysis 2
H Analog Pressure Gauge with no EN837 accuracy class information General Measurement Device and Calibration Topics 5
G Gauge R&R on multiple dimensions using 3D measurement system Capability, Accuracy and Stability - Processes, Machines, etc. 6
H Thread Plug Gauge Tolerances General Measurement Device and Calibration Topics 3
R Do Brands Matter When Purchasing Bore Gauge? General Measurement Device and Calibration Topics 2
D Using "Particle Size Standard" templates as gauges - How to avoid giving a gauge # while using for process control? General Measurement Device and Calibration Topics 2
D AS9100 Inspection records - Is there a requirement to have gauge ID and calibration status AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
G Comparing Gauge Blocks - Translating the calibration between the master set and the QA set General Measurement Device and Calibration Topics 2
F Gauge R&R & Calibration Intervals Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 14
Ron Rompen Source needed for Spline Form gauge (GO and NOGO members) General Measurement Device and Calibration Topics 3
S How to calculate the maximum allowable hysteresis of a pressure gauge General Measurement Device and Calibration Topics 1
S Gauge R and R - Product - Conveyor belt with threads that goes on Sprocket Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
R Feeler Gauge Calibration - What tolerance is used for calibrating feeler gauges? General Measurement Device and Calibration Topics 4
F Holes position inspection - Calculation of centre distance and plug sizes of a gauge Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
K Measurement Uncertainty Budget (MU) of for Ultrasonic Thickness gauge Measurement Uncertainty (MU) 1
R Radius Gauge R&R Study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
A Calibration of Pressure Gauge, How to Determine Calibration Tolerance General Measurement Device and Calibration Topics 1
S Thread Ring Gauge verification General Measurement Device and Calibration Topics 6
G Uncertainty Budget Examples for Caliper, Micrometer and Dial Gauge Measurement Uncertainty (MU) 3
S Pressure Gauge Calibration Uncertainty Calibration and Metrology Software and Hardware 5
N Correct reading for analog pressure gauge ISO 17025 related Discussions 2
Ron Rompen Gauge Design Standards Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
B Gauge CAD certification to Part CAD vs. Gauge certification to Part CAD General Measurement Device and Calibration Topics 1
shimonv Feeler Gauge as a Medical Device Accessory EU Medical Device Regulations 3
C Proper Thickness Gauge for Foam/Felt and Soft Materials General Measurement Device and Calibration Topics 5
J What kind of Tool and Gauge in IATF 16949 Clause 8.5.1.6? IATF 16949 - Automotive Quality Systems Standard 1
alonFAI Gauge R&R for a Peel Tester (tensile tester) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 11
R Height gauge to measure short hard to access heights General Measurement Device and Calibration Topics 3
V Why Gauge R&R 10% is acceptable for variable instruments ? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
B Can we calibrate a standard pin with a gauge? General Measurement Device and Calibration Topics 2
C How to perform Gauge R & R on Automated Test & Measurement Equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 8
I How to determine the thread wire size needed - Buttress Thread Gauge General Measurement Device and Calibration Topics 3
V Inactive Gauge Retention Time - PPAP Section 6 Requirements APQP and PPAP 1
Q Pin Gauge Setup - Measuring the OD of a Catheter Manufacturing and Related Processes 1

Similar threads

Top Bottom