Hello to all, I'm a newbie in a field of 60601-1 and 60950-1 standards. So, it's sometimes hard to understand (or "believe") tough requirements of 60601-1. Hope you will clarify some items to me. I'll be very thankful.
I have AC\DC power supply, which is intended to use in ME EQUIPMENT. I have a task to modify this power supply in accordance with requirements of 60601-1. You can see a simplified diagram of it in the figure below*. Input voltage is 230Vac, output voltage is variable and depends on the model and can be up to 1500Vdc.
I intend to assume the metal case and wires (input, output and interface) as ACCESSIBLE PARTS. In accordance with "8.1 Fundamental rule of protection against electric shock" I'm obliged to prevent unacceptable TOUCH CURRENTS and voltages on surfaces of ACCESSIBLE PARTS. My understanding of proper MOPs is shown with red on the diagram.
As I've understood, for MOPP I have to use Table 12 for creepages and clearances and Table 6 for solid insulation. So, for example, for 1900Vdc MOPP (400+1500Vdc MOP between MAINS PART and SECONDARY PART on my diagram) in accordance with Table 12 I have to provide 50mm creepage distance and 28,6mm clearance distance. The transformer between MAINS PART and SECONDARY PART should provide the same clearances and creepages and to have (in accordance with Table 6) 1900*sqrt2+5000=7679Vac dielectric strength. It's hard to imagine such a monstrosity, I mean transformer and PCB which are suits for requirements above. Are my rationales right?
I think, I have to add 1500Vdc 2xMop galvanic insulation between SECONDARY PART and control board (it's marked with "1?" on my diagram). In this case I have not to provide any insulation between parts marked "2?". Am I right? Is there any way to use protective impedance in feedback instead of galvanic insulation?
Thanks in advance.
*Unfortunately it's my first post and I have no permission to insert images in a body of the post. I've opened access to image in my Google disk. For seeing diagramm please search DiagramForElsmar27.09.17.pdf in Goole. It will be only one link. Hope it won't break forum's rules.
.
I have AC\DC power supply, which is intended to use in ME EQUIPMENT. I have a task to modify this power supply in accordance with requirements of 60601-1. You can see a simplified diagram of it in the figure below*. Input voltage is 230Vac, output voltage is variable and depends on the model and can be up to 1500Vdc.
I intend to assume the metal case and wires (input, output and interface) as ACCESSIBLE PARTS. In accordance with "8.1 Fundamental rule of protection against electric shock" I'm obliged to prevent unacceptable TOUCH CURRENTS and voltages on surfaces of ACCESSIBLE PARTS. My understanding of proper MOPs is shown with red on the diagram.
As I've understood, for MOPP I have to use Table 12 for creepages and clearances and Table 6 for solid insulation. So, for example, for 1900Vdc MOPP (400+1500Vdc MOP between MAINS PART and SECONDARY PART on my diagram) in accordance with Table 12 I have to provide 50mm creepage distance and 28,6mm clearance distance. The transformer between MAINS PART and SECONDARY PART should provide the same clearances and creepages and to have (in accordance with Table 6) 1900*sqrt2+5000=7679Vac dielectric strength. It's hard to imagine such a monstrosity, I mean transformer and PCB which are suits for requirements above. Are my rationales right?
I think, I have to add 1500Vdc 2xMop galvanic insulation between SECONDARY PART and control board (it's marked with "1?" on my diagram). In this case I have not to provide any insulation between parts marked "2?". Am I right? Is there any way to use protective impedance in feedback instead of galvanic insulation?
Thanks in advance.
*
.