Generic or Specific Gauge R&R on Standard Equipment (ie.: Micrometer)

A

amilboa265

#1
We have conducted generic R&R studies on standard equipment ie Micrometer. A customer has requested that a R&R study be conducted using the same equipment (ie micrometer already conducted) but on there specific product. I do not see that this is a requirement in the MSA 4th edition manual, unless anyone else can correct me??
 
Elsmar Forum Sponsor
N

NumberCruncher

#2
Hi amilboa265

There is nothing in the MSA manual that requires you to use a specific customer's product.

The MSA study takes into account the variabilty in the person, the gauge and the part.

Let's say that you make several different product lines for different customers. Call those lines A and B. Then suppose that the process variation in line A is a bit bigger than line B.

Assumng that you have selected a representative range of parts from line A, your R&R results would look ok.

If you select a representative range of parts from line B, which has a lower process variation than line A, your R&R results may not look quite as good.

In this case, your customer may have a point.

If, on the other hand, you are making a standard product, and your customer simply buys from a common stock of goods, then personally, I fail to see the point of carrying out a second study.

How was the generic R&R study done?

The other problem is simply one politics. It's what your customer wants!

NC
 

Jim Wynne

Leader
Admin
#3
We have conducted generic R&R studies on standard equipment ie Micrometer. A customer has requested that a R&R study be conducted using the same equipment (ie micrometer already conducted) but on there specific product. I do not see that this is a requirement in the MSA 4th edition manual, unless anyone else can correct me??
Hi amilboa265

There is nothing in the MSA manual that requires you to use a specific customer's product.

The MSA study takes into account the variabilty in the person, the gauge and the part.

Let's say that you make several different product lines for different customers. Call those lines A and B. Then suppose that the process variation in line A is a bit bigger than line B.

Assumng that you have selected a representative range of parts from line A, your R&R results would look ok.

If you select a representative range of parts from line B, which has a lower process variation than line A, your R&R results may not look quite as good.

In this case, your customer may have a point.

If, on the other hand, you are making a standard product, and your customer simply buys from a common stock of goods, then personally, I fail to see the point of carrying out a second study.

How was the generic R&R study done?

The other problem is simply one politics. It's what your customer wants!

NC
If you do a "generic" gage study, on gage blocks for example, what you're showing in essence is that the measurement system is appropriate (or not) for measuring gage blocks. It's perfectly reasonable for a customer to expect GR&R to be done on the part feature(s) of interest.

There is no reason, however, that products and their characteristics shouldn't be arranged in "families" when it's sensible to do so. If the characteristics to be measured are sufficiently similar to make additional gage studies needlessly redundant, there shouldn't be a need to do additional studies. This is a matter of rational judgment and customer acquiescence, not something to be done to avoid useful work. Sometimes customers won't listen to reason and needless work has to be done, but if your rationale for GR&R economy is sound, you should at least discuss it with the customer.
 
A

amilboa265

#4
Thanks NC. We make reasonably standard product, all customer parts are bespoke, but variations on our core manufacturing capability, using micrometers and verniers to measure lengths and diameters.

The original GRR were 3 people, 10 parts, 3 repeats to get results for a type of micrometer, which we use across the company. This has always been accepted by all customers. Whilst i appreciate the political aspect, I do not want to set a precedent with this request, as it would probably not be sustainable to continually carry out GRR for each new part.

I really wanted to make sure that I hadn't missed anything in the MSA book. Thanks very much for your reply.
 

Jim Wynne

Leader
Admin
#5
Thanks NC. We make reasonably standard product, all customer parts are bespoke, but variations on our core manufacturing capability, using micrometers and verniers to measure lengths and diameters.

The original GRR were 3 people, 10 parts, 3 repeats to get results for a type of micrometer, which we use across the company. This has always been accepted by all customers. Whilst i appreciate the political aspect, I do not want to set a precedent with this request, as it would probably not be sustainable to continually carry out GRR for each new part.

I really wanted to make sure that I hadn't missed anything in the MSA book. Thanks very much for your reply.
I don't have the 4th Edition MSA manual handy, but I can tell you that nearly all of it is based on the assumption that R&R (and other analyses) will be done using parts and not standards. I don't know what you mean when you refer to "generic" studies, but if you mean that the studies weren't done using parts representative of your products, I wouldn't be surprised if a customer rejected them.
 
A

amilboa265

#6
By generic i meant that we have used one of our parts in the GRR study. Our products are all variations on a theme, so whilst the GRR study is not specific to one customer, the parts should be representative of what we supply to that customer. As i said, all our parts are bespoke to customer drawings, but they are all variations of a theme.
 

Jim Wynne

Leader
Admin
#7
By generic i meant that we have used one of our parts in the GRR study. Our products are all variations on a theme, so whilst the GRR study is not specific to one customer, the parts should be representative of what we supply to that customer. As i said, all our parts are bespoke to customer drawings, but they are all variations of a theme.
Again, I don't have the 4th Edition MSA manual handy, but it does account for doing "families" of gages. Perhaps someone here can provide the direct reference.
 

Golfman25

Trusted Information Resource
#8
I think Jim's first point is right on. If there is justification of similarly situation parts, i.e; families, then seperate studies are a waste (and don't we all want to reduce waste?). However, it might be easier to do the study than convince your customer. Unfortunately today, customers are more concerned that all the "boxes match" rather than some common sense thinking. Good luck.
 
N

NumberCruncher

#9
Again, I don't have the 4th Edition MSA manual handy, but it does account for doing "families" of gages. Perhaps someone here can provide the direct reference.
Hi Jim

I have searched through the MSA manual, and whilst I can't say that I have read every single word, I can't find any reference to families of gauges. Miner mentions them in one of his blogs.

Free - Miner's MSA "Blog" Series

Other than that, I can only find a document on the internet, "Ford Motor Company Customer-Specific Requirements For Use With ISO/TS 16949:2002

http: //www .iaob .org /pdfs/FordspecTS2nd_December 2007.pdf - DEAD 404 LINK UNLINKED

Page 18 states

"4.35
Measurement systems analysis (ISO/TS 16949:2002 cl. 7.6.1)


All gauges used for checking Ford components/parts per the control plan shall have a gauge R&R performed in accordance with the appropriate methods described by the latest AIAG Measurement Systems Analysis Manual (MSA) to determine measurement capability.

Any measurement equipment not meeting the specifications stipulated in the MSA must be approved by STA.

Use of family gauge studies per the MSA is permissible and must be approved by STA "

Where STA is

"Supplier Technical Assistance – Ford Motor Company's team dedicated to assist in the development of supplier processes."

I don't know if this is a unique Ford concession, or if it applies more widely. Perhaps Miner could enlighten us?

NC
 
Last edited by a moderator:
#10
Hi Jim

I have searched through the MSA manual, and whilst I can't say that I have read every single word, I can't find any reference to families of gauges. Miner mentions them in one of his blogs.

Free - Miner's MSA "Blog" Series

Other than that, I can only find a document on the internet, "Ford Motor Company Customer-Specific Requirements For Use With ISO/TS 16949:2002

http ://www .iaob. org/pdfs/FordspecTS2nd_December 2007.pdf - DEAD 404 LINK UNLINKED

Page 18 states

"4.35
Measurement systems analysis (ISO/TS 16949:2002 cl. 7.6.1)


All gauges used for checking Ford components/parts per the control plan shall have a gauge R&R performed in accordance with the appropriate methods described by the latest AIAG Measurement Systems Analysis Manual (MSA) to determine measurement capability.

Any measurement equipment not meeting the specifications stipulated in the MSA must be approved by STA.

Use of family gauge studies per the MSA is permissible and must be approved by STA "

Where STA is

"Supplier Technical Assistance – Ford Motor Company's team dedicated to assist in the development of supplier processes."

I don't know if this is a unique Ford concession, or if it applies more widely. Perhaps Miner could enlighten us?

NC
Where the Ford thing says "Use of family gauge studies per the MSA..." it most likely means per the MSA manual, but because I don't have the 4th edition manual I can't confirm it. Thanks for looking. In any event, family GR&R is common practice and should be done where appropriate and where customers agree to it.
 
Last edited by a moderator:
Thread starter Similar threads Forum Replies Date
C Appropriate ways to address Competency? Generic vs Specific ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
O Process Audit Checklist - Generic vs. Process Specific Process Audits and Layered Process Audits 17
N Clinical trials - Generic "parent" procedure vs. study-specific "child" versions Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2
J Generic or Specific Checklists for Internal Audits Internal Auditing 15
T Assessing risk where harm is indirect - Generic devices / accessories / intermediates ISO 14971 - Medical Device Risk Management 8
I How to determine "generic/me too" device? Japan Medical Device Regulations 0
K Generic preventive maintenance form wanted Document Control Systems, Procedures, Forms and Templates 2
V Generic IATF 16949 Audit Checklist wanted IATF 16949 - Automotive Quality Systems Standard 3
V Generic ISO 50001 checklist Various Other Specifications, Standards, and related Requirements 7
N Generic Rework/Reprocessing Procedure Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
V Generic requirements regarding capability study in automotive Capability, Accuracy and Stability - Processes, Machines, etc. 1
M Is IEC 60601 compliance possible when only generic power-supply is specified? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S Family / Generic type of FMEA & Control Plan and Symbols FMEA and Control Plans 3
Marc The Risks of Generic Drugs Coffee Break and Water Cooler Discussions 24
R Connection to a generic separate power supply IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
Marc Are you looking for ISO 31000 - Risk Management Principles and Generic Guidelines? ISO 14971 - Medical Device Risk Management 1
Marc Urine Spills Staining Image of Wockhardt?s Generic Drugs Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
Ajit Basrur FDA - Guidance for Industry - Generic Drug User Fee Amendments of 2012 - Q & A US Food and Drug Administration (FDA) 0
M Engineering Change Management Generic Procedure Manufacturing and Related Processes 3
D FDA MDDS - Listing Obligations for a Generic Data Interface US Food and Drug Administration (FDA) 3
R Do generic forms bought over the counter need to be controlled? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
A Should the FSMS have a Documented Generic Procedure for Verification and Validation Food Safety - ISO 22000, HACCP (21 CFR 120) 2
T Generic Guides for Procedure Writing Document Control Systems, Procedures, Forms and Templates 20
S Draft Guidance on Generic Drug User Fee Amendments (GDUFA) 2012 Other US Medical Device Regulations 0
C Control Plan Template (Generic) Document Control Systems, Procedures, Forms and Templates 0
AnaMariaVR2 Generic Drug User Fee Act (GDUFA) US Food and Drug Administration (FDA) 0
I Question on Declaration of Conformity for device subcategory or generic device groups EU Medical Device Regulations 2
P Generic Contract Manufacturing of Generic Prescription Medications Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
Hershal Couple of generic TickIT and ISO 27000 questions IEC 27001 - Information Security Management Systems (ISMS) 9
Q Numbering or Naming (Identification) of Generic Company Specifications? Document Control Systems, Procedures, Forms and Templates 3
S Complaint Process flow: Need a generic definition which would flag an "MDR" Customer Complaints 7
optomist1 Generic question regarding Two Way ANOVA Results Statistical Analysis Tools, Techniques and SPC 8
D Seeking Help: Nondisclosure vs. generic QMS documents Misc. Quality Assurance and Business Systems Related Topics 10
M Generic "Compliant Procedure" and "Non-conformance" forms needed Document Control Systems, Procedures, Forms and Templates 1
K Generic Product Development Documents that need Analysis in a Complaint Investigation Problem Solving, Root Cause Fault and Failure Analysis 2
D Process Documentation is extremely generic IATF 16949 - Automotive Quality Systems Standard 4
K Generic worksheet for calculating Supplier Chargeback and Cost of Poor Quality (COPQ) Supplier Quality Assurance and other Supplier Issues 2
Ajit Basrur Facts and Myths about Generic Drugs Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2
P Device Subcategory and Generic Device Group differences EU Medical Device Regulations 2
R Audit N/C - Quality Policy too generic? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
G Any rule to set Generic Process Target % (per cent) Manufacturing and Related Processes 2
P Generic Label Die Cutting Process flow Manufacturing and Related Processes 5
C Production Audit Checklist - Generic General Auditing Discussions 14
Ajit Basrur GIVE (Generic Initiative for Value and Efficiency) Launched by the FDA (US) Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
K Generic Term for Inspection Types FMEA and Control Plans 15
B Generic list of ISO9001 processes for an organisation (manufacturing set up) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
H FDA Site generic Supplier Related forms help needed ISO 13485:2016 - Medical Device Quality Management Systems 2
Q Does anyone have a generic Product Specification template? Document Control Systems, Procedures, Forms and Templates 7
G Is there a standard or generic set of PPAP documentation available? APQP and PPAP 4
G Generic Value Stream Map - Example of how Lean can be applied in a Call Center Lean in Manufacturing and Service Industries 8

Similar threads

Top Bottom