GR&R (Gage R&R) on leak test equipment is there a deviation requirment?

H

haueyman

#1
I am running a GR&R on leak test equipment. I have two leak test equipment that test parts exactly the same:lol: or should be testing them the same. My question is when running a GR&R on these two machines how much of a deviation from one machine to the other should I expect.

The most that I have seen so far running stdev on excel was 7.35 and the least was 1.48 stdev. If any one could shed a little light on this it would be appreciated.

:thanks:
 
Last edited by a moderator:
Elsmar Forum Sponsor

Miner

Forum Moderator
Staff member
Admin
#2
Re: GR&R on leak test equipment is there a deviation requirment?

We really need more information to assist you.

What type of leak test equipment is this? Mass spec based using trace gases? In a sealed canister, a hand-held or robotic probe? Stationary or moving?

At a high level, you could perform a hypothesis test to determine whether the difference in results is statistically significant. The variation that you use in the equation should be the measurement variation, not the product variation.

However, before you do this, ensure that the gages have been calibrated, then determine whether each gage individually has an acceptable MSA. If they do not, correct the gage first.
 
H

haueyman

#3
Re: GR&R on leak test equipment is there a deviation requirment?

We leak test engine blocks and heads that sit on a table and are transfered or rotated into 3 stationary stations. Each testing a different cavity. I have ran the R&R on them and they pass as an individual gage but when I do a comparision between the two there is a big difference.

We use the same parts and test them the same way on the equipment.

I have never really looked into the difference between the two machines we always were concerned with the individual gage.
 

Miner

Forum Moderator
Staff member
Admin
#4
Re: GR&R on leak test equipment is there a deviation requirment?

Have the two gages been calibrated against a standard?

It would also help if you provide information about the type of leak test equipment you are using. Mass-spec, pressure decay? The more information that you provide, the less guessing we have to do to assist you.
 
H

haueyman

#5
Re: GR&R on leak test equipment is there a deviation requirment?

Well they are calibrated to a part that is closest to zero on each circut in the test.

The test is pressure decay on engine heads and blocks. We test the leakage by filling up the cavity (low pressure oil, water jacket and high pressure oil) then letting the cavity stabalize for 30 secs then test the air loss.

When running the GR&R according to supplier specs is 10% but the equipment was never able to hold that so they accepted a deviation or a max allowable in each circit. We are within the max allowable but I wans't sure if I should be concerned with the variation that I see from the first leak tester to the second leak tester. We have two leak testers that check the parts exactly the same we just have two for more throughput.
 

Miner

Forum Moderator
Staff member
Admin
#6
Re: GR&R on leak test equipment is there a deviation requirment?

Well they are calibrated to a part that is closest to zero on each circut in the test.

The test is pressure decay on engine heads and blocks. We test the leakage by filling up the cavity (low pressure oil, water jacket and high pressure oil) then letting the cavity stabalize for 30 secs then test the air loss.

When running the GR&R according to supplier specs is 10% but the equipment was never able to hold that so they accepted a deviation or a max allowable in each circit. We are within the max allowable but I wans't sure if I should be concerned with the variation that I see from the first leak tester to the second leak tester. We have two leak testers that check the parts exactly the same we just have two for more throughput.
Try the hypothesis test for significance between the two machines using the measurement variation determined by the MSA.

You may want to run an experiment where you vary the stabilization time to check the effect on the MSA.

If you can attach your data, we can assist in the analysis/interpretation.
 
Thread starter Similar threads Forum Replies Date
S Capability or Gage R&R Study for Leak Tester? Reliability Analysis - Predictions, Testing and Standards 15
F Leak Tester Gage R&R Sheet Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
O Capability or Gage R&R Study for Leak Tester Reliability Analysis - Predictions, Testing and Standards 11
T Gage Repeatability for an Automated Leak Tester Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
B MSA (Gage R&R) vs. Capability Cpk Study - Helium Leak Check Station Statistical Analysis Tools, Techniques and SPC 2
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
L Gage R&R TMV Acceptance Criteria Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
S MSA for attribute relation gage Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
E Zero part to part variation - Gage R&R project Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 15
M Gage R&R and right way to measure Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 16
L Gage RandR on automated equipment. IATF 16949 - Automotive Quality Systems Standard 7
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
D Difference between Test Method Validation and Gage R&R Qualification and Validation (including 21 CFR Part 11) 18
R Determining Uncertainty from Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
C Gage Block Wringing General Measurement Device and Calibration Topics 3
B Gage R&R with NDC=1 Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
S Type 1 Gage R&R or something else? Reliability Analysis - Predictions, Testing and Standards 6
G Should I perform Gage R&R only at the beginning of a new project? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
V Thread Plug Gage Pitch GO Diameter out of spec AQL - Acceptable Quality Level 8
G Gage R&R - Where am I going wrong? Part of a FAIR submission (Aerospace) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
R Gage R&R Excel templates Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
K Gage R&R with more than 3 appraisers Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
Anerol C Gage R&R Template AIAG 4th Edition IATF 16949 - Automotive Quality Systems Standard 3
T Attributes SPC study - Attributive control (Go gage) Statistical Analysis Tools, Techniques and SPC 5
N % Tolerance - Type 1 study on the gages, then a gage R&R (ANOVA) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
M Definitive answer on Type 1 vs Type 2 vs Type 3 Gage Study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
V Gage Management and Gage R&R Software General Measurement Device and Calibration Topics 1
D Gage R&R Study on Load Cells - Large chemical blending tanks IATF 16949 - Automotive Quality Systems Standard 1
B Minitab Type 1 Gage Study on True Position Question Measurement Uncertainty (MU) 1
S Gage R&R on alloy Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
8 Need Help - Runout - Function Gage Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
L Gage R&R studies for identical Devices - Need to confirm the requirement to perform them Reliability Analysis - Predictions, Testing and Standards 2
G Stab Pin Size For Check Gage Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
E Functional Gage for TOP applied at RFS Design and Development of Products and Processes 5
F No reproducibility Error in Gage R&R? Reliability Analysis - Predictions, Testing and Standards 4
M Template for Attribute Gage R & R wanted Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
G Warning about Deltronic Gage Pins - The "Certification of Accuracy" document General Measurement Device and Calibration Topics 13
A Gage type 1 study on CMM Capability, Accuracy and Stability - Processes, Machines, etc. 2
Q Gage R&R for Instant Measurement Machine Manufacturing and Related Processes 6
M Gage R&R with large inter part variance - Torque Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
D Do we need normal data for gage r&r studies? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
D Do I need part variation while doing Destructive Variable Gage R&R MSA study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
S Gage R&R Standard Deviations Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
J Informational Gage R&R: Swapping operators Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
S Gage R & R for large amount of inspections IATF 16949 - Automotive Quality Systems Standard 10
S Ring Gage Thermal Expansion General Measurement Device and Calibration Topics 4
L Gage R&R for test precision or uncertainty prediction? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
S Minor diameter of threaded ring gage General Measurement Device and Calibration Topics 1
A Does Torque Gage Need Validation? Reliability Analysis - Predictions, Testing and Standards 4

Similar threads

Top Bottom