Ground stud requirements on our medical device in IEC 60601

M

mouse7

#1
Hi all, appreciate your clarification on this point.
I cannot find a requirement for a ground stud on our medical device in 60601. Our chassis is plastic, do we need to add a ground stud just for testing purposes?
If we don't have a ground stud, how will leakage tests be conducted?

Thanks!
 
Elsmar Forum Sponsor
#2
Medical electrical equipment (MEE) may be Class II - double insulated. As such they would not have a 'Protective Earth' and would not be fitted with a ground stud.

(MEE for home healthcare settings is required to be Class II).

To measure leakage currents - for enclosure leakage (Touch Current) the measurement is taken between ground and foil applied to the outside of the enclosure. The size of the foil represents an adult open hand (or bigger if risk analysis demands). The foil is moved around the enclosure to find the 'worst case' positioning.

Internally powered MEE is not tested for Touch Current unless you have SIP/SOP.

The other tests are done in similar ways - using foil on the outside of the enclosure or insulated applied parts.
 

phase90

Involved In Discussions
#3
Along the same lines, if you have a product that is Class I with a plastic enclosure, does it need a ground stud?
 
#4
Along the same lines, if you have a product that is Class I with a plastic enclosure, does it need a ground stud?
Assuming that you are thinking of an external ground stud, then I would suggest that, no you do not have to have one.

However, electrical equipment, especially medical devices, are routinely tested often by a hospitals own engineering / equipment management service. A key test that they will try to do is earth bond integrity. If they cannot find a suitable earth point they may attempt to use something like the metal shell of a connector. As this test is usually done with 25 amps of current it would likely cause damage if that current was passed through unsuitable conductors, such as a PCB trace.

But if you provide an external earth point, with strong bonding then the test is easy to complete and no damage is caused.

As with just about everything in this game it's all down to risk analysis.
 
Thread starter Similar threads Forum Replies Date
T Accessible metal parts connected to secondary ground IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Metal Enclosure Connected to PCB Ground for ESD Reasons IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Sidney Vianna Informational ESG - Environment, Social & Governance issues gaining ground in the Business World Sustainability, Green Initiatives and Ecology 22
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M How to Start SPC from the ground up! Statistical Analysis Tools, Techniques and SPC 8
J Bees or Hornets Nesting in the Ground Coffee Break and Water Cooler Discussions 28
A Ground bond test when no ground is accessible IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601 Ground Bond Test Requirements for a Plastic Enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
somashekar Mangoes ground Air India flight ~~~ Travel - Hotels, Motels, Planes and Trains 1
G How many Airbus A380 planes must blow Rolls-Royce motors to ground the fleet? Coffee Break and Water Cooler Discussions 36
apestate Ground Tooling produced from OEM Blueprints - Part 21? Of WHAT? Federal Aviation Administration (FAA) Standards and Requirements 9
W Sunday Business Systems Ground Control? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
C How long does it take to clear back ground checking? Career and Occupation Discussions 7
J ISAGO - IATA Safety Audit for Ground Operations AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 12
C Applicability of AS9100 to Aircraft Ground Support Equipment AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 10
D Aeronautical regulations JAR145, JAR21G - Aircraft ground handling AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
M Paint booth prints - Painters using them as ?ground cloth? after the job is shipped Document Control Systems, Procedures, Forms and Templates 2
M Weld Stud Length Poka Yoke Nonconformance and Corrective Action 3
H What does the abbreviation for ?POAG? terminal/stud stand for? Manufacturing and Related Processes 3
J Screw Thread Inspection Standards? Double Ended Stud Inspection Criteria Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
CarolX I am looking for some information on stud welding Manufacturing and Related Processes 28
F Product classification according to the 60601-2-57: Particular requirements [...] of non-laser light sources [...] IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
DuncanGibbons Safe-Life, Fail-Safe and Damage Tolerant approaches and their relevant material property requirements Federal Aviation Administration (FAA) Standards and Requirements 0
B Protective impedance as MOOP/MOPP - Requirements? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Sidney Vianna IAQG News New IAQG Resources - AS9104-3 Requirements for Aerospace Auditor Competency & Training Courses AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
D IATF Requirements if remote plant has no interface with OEMs who are part of IATF? IATF 16949 - Automotive Quality Systems Standard 6
D Cleanroom requirements for surgical apparel EU Medical Device Regulations 2
M Customer Specific Requirements - Packaging Spec IATF 16949 - Automotive Quality Systems Standard 10
M FDA Syringe Marking requirements US Food and Drug Administration (FDA) 4
R Flammability requirements for applied parts as per 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
B Acquired Medical Device Product Line - Documentation Requirements for Device Master Record ISO 13485:2016 - Medical Device Quality Management Systems 7
D Customer requirements on specific standards Various Other Specifications, Standards, and related Requirements 5
S State Licensing requirements for Medical Devices sold in the US directly to hospitals and tissue banks 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
R PPAP Requirements for Tool Repair or Refurbishment APQP and PPAP 5
I France clinical trial requirements EU Medical Device Regulations 1
J Outsourced Internal Audit requirements for Aerospace Suppliers AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 21
H Electronic care product on patient bed side - Flammability requirements (62368-1) Hospitals, Clinics & other Health Care Providers 0
D Digital Healthcare - QMS requirements Other US Medical Device Regulations 3
M GP-12 procedure form and requirements Customer and Company Specific Requirements 1
G How to implement H&S and Quality Control Requirements in Contract for Potential Supplier? Contract Review Process 6
D Customer Specific Requirements / Customer Requirements for Indirect Customers IATF 16949 - Automotive Quality Systems Standard 10
J Print Requirements Manufacturing and Related Processes 9
S Requirements for Certificate of Destruction for Surgical Instruments Medical Device and FDA Regulations and Standards News 2
R FAI & P.O. Requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 21
W MDD Distributor Registration Requirements Other Medical Device Regulations World-Wide 0
M FDA Requirements for Investigational Devices - Clinical Investigation & Shipping Medical Device and FDA Regulations and Standards News 0
B Requirements for LiPo Batteries in the EU ? CE Marking (Conformité Européene) / CB Scheme 7
pbojsen ISO 13485 Requirements versus FDA product classification and GMP exemptions - Audits ISO 13485:2016 - Medical Device Quality Management Systems 5
A MDR - Legacy Device Review Timeframe and Requirements EU Medical Device Regulations 3

Similar threads

Top Bottom