Hand-Held dosing device has no PATIENT - Interpretation of the PATIENT definition

urticantwriter

Registered
Hi everyone,

For the past year I've been working in a team developing a HAND-HELD dosing device for people that today administer their prescription drugs with pills or capsules. The idea is to load the device with pellets and then dispense pellets on the demand of the "patient".

The reason for "patient" and not PATIENT is that the definition of patient in the IEC 60601-1 states:
  • PATIENT: living being (person or animal) undergoing a medical, surgical or dental procedure
    NOTE A PATIENT can be an OPERATOR
Is it your opinion that the administering of drugs through a device like ours really makes a medical procedure? And if it does, the PATIENT is rather an operator of the device when handling it, and a PATIENT in the procedure of taking the drugs from the device?

If anyone has any experience from similar cases or just feel confident in their interpretation of the PATIENT definition please let me know!

Kind Regards,
Oscar
 

urticantwriter

Registered
IMO ingesting medication is a medical procedure.

That might be so. But looking at our device it is really a question of storing and delivering the right amount for treatment. Sort of like what a jar is for tablets and capsules.

Thank you for your input Ronen E!
 

Ronen E

Problem Solver
Moderator
That might be so. But looking at our device it is really a question of storing and delivering the right amount for treatment. Sort of like what a jar is for tablets and capsules.

Thank you for your input Ronen E!
The terminology “dosing device” suggests that the device does more than just dispense or make the pellets available. A jar doesn’t do that.

Jar equivalence aside, let’s take your argument one step further. At the time of dispensing (or dosing) no medical procedure takes place, right? All that happens is pellets moving around. So the “patient” is not really a patient, just an operator. If that’s the case, there is no patient for that device (they only become real patients later, when the device is not in use), and the device has no medical purpose so it’s not a medical device and IEC 60601 doesn’t apply to it.
 

Conaider

Registered
Hi Oscar,

Is your end goal only to define what PATIENT is? Or is it to determine if your device falls within the scope of 60601-1? To me it would sound that this is a dispensing device and it wouldn't necessarily be considered MEDICAL ELECTRICAL EQUIPMENT (therefore out of scope). It does not have an APPLIED PART nor does it transfer energy to/from the PATIENT.

Of course, I don't know the details f the device so I might be wrong; still, something worth considering.
 
Thread starter Similar threads Forum Replies Date
P Application of 60601-1 and 60601-2-2 to a hand-held laparoscopic unstrument with motorisation in the handle IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
D Hand Held medical device - power supply requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 23
G Is impact test according to IEC 60601-1 applicable for HAND-HELD equipment if also classified as BODY-WORN? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
U PATIENT CONNECTION for HAND-HELD Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E Requirements for hand-held laser welding equipment Other ISO and International Standards and European Regulations 0
B Figure A.20 of IEC60601-1 about definition of "Hand-held" & "Portable" IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
D Hand-held Digital Tension Meter Calibration General Measurement Device and Calibration Topics 1
J Kanomax hand held Laser Particle Counter Calibration Calibration Frequency (Interval) 6
B Flammability Ratings for Internal Battery & Hand-Held ME Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
S Medical Device or Cosmetic Device in Korea? Hand Held Wrinkle Reduction LED Light Other Medical Device Regulations World-Wide 1
S Medical Device or Cosmetic Device in Japan? Hand Held Wrinkle Reduction LED Light Japan Medical Device Regulations 3
M Should a low cost hand held refractometer be in a calibration program? Calibration Frequency (Interval) 8
C Testing Agencies - Gas Meters and Hand Held Analyzers General Measurement Device and Calibration Topics 2
M 7.5 - Hand Written Record Transfer ISO 17025 related Discussions 4
E Artificial hand (IEC 60601-1-2) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
I Are suppliers required to hand over process validation reports? ISO 13485:2016 - Medical Device Quality Management Systems 20
J Machinery Directive : Emergency stop clause : definition of 'portable' or 'hand-guided' machinery Other ISO and International Standards and European Regulations 0
R Electronic and hand-written footnotes for GDP Document Control Systems, Procedures, Forms and Templates 1
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A Components of a medical device - Hardware robotic hand EU Medical Device Regulations 3
L Part 11 Certified to legally binding of hand written signatures Other Medical Device Regulations World-Wide 3
D Use of hand cream inside gloves in the cleanroom for medical devices Occupational Health & Safety Management Standards 7
S 21 CFR 11 - Require paper and hand signatures not electronic Other US Medical Device Regulations 2
L Hand-starting threaded fasteners Benchmarking 3
D Test Report written by hand or electronically ISO 13485:2016 - Medical Device Quality Management Systems 7
R Approving a New Hand Assembly Line at a Supplier Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
GURU SDI VDA 6.3 Hand Book - Two sections for Potential Analysis VDA Standards - Germany's Automotive Standards 1
K Preventing Shortage of Product Quantity - Hand Count and Scale Count Manufacturing and Related Processes 13
T Thread Gaging: Hand Tight Definition Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
B Hand Swabbing Employees - Guidelines of micro limits and procedures and frequency Food Safety - ISO 22000, HACCP (21 CFR 120) 2
R Silicone in Hand Lotion in Electronics Manufacturing Manufacturing and Related Processes 4
A Configuration Management and On-Hand/In-Transit Inventory Quality Manager and Management Related Issues 3
G I hope someone can give me a hand.. Various Other Specifications, Standards, and related Requirements 4
S Critical Limits for Hand Washing Food Safety - ISO 22000, HACCP (21 CFR 120) 13
A Microbial Limit (Bioburden)on dry nonwoven that used for making Hand sanitizing wipes ISO 13485:2016 - Medical Device Quality Management Systems 1
M Microbiological Standards (Limits) for Hand Swabs for Coliforms and S. Aureus Food Safety - ISO 22000, HACCP (21 CFR 120) 15
Icy Mountain Alcohol Based Hand Sanitizers in Electronics Manufacturing Manufacturing and Related Processes 10
M Ergonomics - Human Hand - Maximum force - Wheel/Flywheel - Help ASAP :) Human Factors and Ergonomics in Engineering 8
S Confusion DPU, PPM & DPMO vs. Sigma Level - ASQ Six Sigma Black Belt hand book Six Sigma 14
T Sanitary "Green" Restrooms - Hand Driers and Waterless Urinals - Food Manufacturing Sustainability, Green Initiatives and Ecology 20
R Hand tool to measure angles that has been machined into stock Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
A Hand-Out for Human Factors Training Training - Internal, External, Online and Distance Learning 7
F Production Employees and hand gage issues Human Factors and Ergonomics in Engineering 13
L Original Documentation - Need to maintain any original hand written documents ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 28
D Is CE Mark required for Hand Sanitizers to be sold in Europe ? EU Medical Device Regulations 6
M Hand Tool Selection to Remove Microjoints and Slivers off of Small Parts - Advice Manufacturing and Related Processes 6
Q Using Faxed Hand Signatures for Approval Records US Food and Drug Administration (FDA) 17
S How can I die cut foam - I'm thinking of a simple hand machine? Manufacturing and Related Processes 15
A Removing the liner from disposable nicotine patch without using hand ISO 13485:2016 - Medical Device Quality Management Systems 11
R CAPA (Corrective and Preventive Action) Discussions are Getting Out of Hand Nonconformance and Corrective Action 41

Similar threads

Top Bottom