Hand Held medical device - power supply requirements

#1
Hello all,

First time posting on this forum, so thank you in advance for your comments and advice.

We've developed a Class II hand held device which is powered by a Lithium-ion rechargeable battery. The device will be supplied with a docking station that is connected to the mains supply. We're using a pogo-pin arrangement between the device and the station. The device cannot be used while docked.

My question is about the power supply that will be moulded into the docking station. Do we absolutely need to use an already certified 60601-1 power supply, or can we use a 60950-1 certified power supply (or other relevant standard) and have the docking station and embedded power supply tested to 60601-1 as an entire system? If the allowable leakage from the non-60601-1 certified power supply is too large, is our only option to move to a 60601-1 charger? Or can we incorporate our own isolation barriers into the design of the device / docking station that will be sufficient?

We're exploring all options as we're finding the cost of 60601-1 power supplies quite prohibitively expensive.

Thanks.
 
Elsmar Forum Sponsor
#2
Wondering if anyone has experienced a similar dilemma or found a definitive answer. Happy to provide more details if the details above are unclear?

Thanks.

David.
 

Peter Selvey

Staff member
Super Moderator
#3
I think the solution would be to treat the docking station as a separate "device" from the point of view of the regulations. If it's purpose made for your system, it's still technically a medical device (accessory) but since it doesn't contact the patient it can be assessed to other standards as appropriate for a normal battery charger. Take care though that sometimes those standards can be a bit obscure as well and can end up being expensive.
 
#4
Thanks for the reply Peter. Just to clarify, the device cannot be powered on while in the docking station, however, technically the patient can make contact with the applied part of the device while it’s docked and connected to the mains.
 

Peter Selvey

Staff member
Super Moderator
#5
Applied part is defined as having contact necessary for the intended use. So incidental contact is excluded. Someone could argue that it's not time related i.e. once an applied part always an applied part, but from first principles, applied parts have special rules based on the nature of contact which just don't apply in this case.
 
#6
Applied part is defined as having contact necessary for the intended use. So incidental contact is excluded. Someone could argue that it's not time related i.e. once an applied part always an applied part, but from first principles, applied parts have special rules based on the nature of contact which just don't apply in this case.
Appreciate again your input Peter. Further clarification from my side: when the device is charging, the charging base is like a cradle that the device is seated into. The applied part makes contact around the eye. Therefore, while docked, the device could in theory be lifted and the applied part placed on the eye. As mentioned before the device can not be turned on while charging. But the argument being made is that irrespective of the device being off, the device is connected to mains supply through it's connection with the docking station. The interaction to lift the device to the eye while docked would be very awkward and counter intuitive but I suppose possible. Can an argument be made that the intended use and design of the ME system make it highly unlikely (I concede not impossible) that the user will attempt to place the applied part on their eye while docked. It should also be noted that during normal operation, there is 90 second "start-up" phase after the user has powered on the device before it is operational.
 

Peter Selvey

Staff member
Super Moderator
#7
What you are describing is foreseeable misuse. By definition, an applied part does not take into account foreseeable misuse, incidental, accidental contact etc. An applied part is a part the contacts the patient during the intended use as described by the manufacturer. Obviously that does not include while the device is charging, which can be specified in the IFU and also protected against e.g. if power on is prevented by software/hardware while charging.

Misuse, incidental contact still needs to be taken into account, but it's a separate issue and the judgement is based on risk. In my opinion it's not an issue largely due to the nature of contact, it's just not a high risk type of contact that raises special concerns. Things like like ECG electrodes or catheters, infusion devices etc that make low impedance contact with the body over long periods, sometimes with small contact areas, sometimes for unconscious patients or patient's that cannot react; they need special consideration. None of those apply. And as you said, it would be fairly rare thing to do anyway, being awkward.

Keep in mind that:
(1) risk is based on the probability of harm, not the probability of contact or misuse
(2) the standard for battery chargers will still have excellent insulation, it's just not the super special high level we required for applied parts in medical devices, for the reasons given above (low impedance, long duration etc.).

It's not like someone will get electrocuted from the battery charger if they put it up on their eye ;)
 
#8
Thank Peter, the points you make are well considered and make perfect sense. I think my next step is to establish the potential harm that could be caused by misuse as described previously, while considering the short contact time and the fact that the user will be conscious and not attached to the device.
 
Thread starter Similar threads Forum Replies Date
S Medical Device or Cosmetic Device in Korea? Hand Held Wrinkle Reduction LED Light Other Medical Device Regulations World-Wide 1
S Medical Device or Cosmetic Device in Japan? Hand Held Wrinkle Reduction LED Light Japan Medical Device Regulations 3
G Is impact test according to IEC 60601-1 applicable for HAND-HELD equipment if also classified as BODY-WORN? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
U Hand-Held dosing device has no PATIENT - Interpretation of the PATIENT definition IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
U PATIENT CONNECTION for HAND-HELD Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E Requirements for hand-held laser welding equipment Other ISO and International Standards and European Regulations 0
B Figure A.20 of IEC60601-1 about definition of "Hand-held" & "Portable" IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
D Hand-held Digital Tension Meter Calibration General Measurement Device and Calibration Topics 1
J Kanomax hand held Laser Particle Counter Calibration Calibration Frequency (Interval) 6
B Flammability Ratings for Internal Battery & Hand-Held ME Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
M Should a low cost hand held refractometer be in a calibration program? Calibration Frequency (Interval) 8
C Testing Agencies - Gas Meters and Hand Held Analyzers General Measurement Device and Calibration Topics 2
R Electronic and hand-written footnotes for GDP Document Control Systems, Procedures, Forms and Templates 1
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A Components of a medical device - Hardware robotic hand EU Medical Device Regulations 3
L Part 11 Certified to legally binding of hand written signatures Other Medical Device Regulations World-Wide 3
D Use of hand cream inside gloves in the cleanroom for medical devices Occupational Health & Safety Management Standards 7
S 21 CFR 11 - Require paper and hand signatures not electronic Other US Medical Device Regulations 2
L Hand-starting threaded fasteners Benchmarking 3
D Test Report written by hand or electronically ISO 13485:2016 - Medical Device Quality Management Systems 7
R Approving a New Hand Assembly Line at a Supplier Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
GURU SDI VDA 6.3 Hand Book - Two sections for Potential Analysis VDA Standards - Germany's Automotive Standards 1
K Preventing Shortage of Product Quantity - Hand Count and Scale Count Manufacturing and Related Processes 13
T Thread Gaging: Hand Tight Definition Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
B Hand Swabbing Employees - Guidelines of micro limits and procedures and frequency Food Safety - ISO 22000, HACCP (21 CFR 120) 2
R Silicone in Hand Lotion in Electronics Manufacturing Manufacturing and Related Processes 4
A Configuration Management and On-Hand/In-Transit Inventory Quality Manager and Management Related Issues 3
G I hope someone can give me a hand.. Various Other Specifications, Standards, and related Requirements 4
S Critical Limits for Hand Washing Food Safety - ISO 22000, HACCP (21 CFR 120) 13
A Microbial Limit (Bioburden)on dry nonwoven that used for making Hand sanitizing wipes ISO 13485:2016 - Medical Device Quality Management Systems 1
M Microbiological Standards (Limits) for Hand Swabs for Coliforms and S. Aureus Food Safety - ISO 22000, HACCP (21 CFR 120) 15
Icy Mountain Alcohol Based Hand Sanitizers in Electronics Manufacturing Manufacturing and Related Processes 10
M Ergonomics - Human Hand - Maximum force - Wheel/Flywheel - Help ASAP :) Human Factors and Ergonomics in Engineering 8
S Confusion DPU, PPM & DPMO vs. Sigma Level - ASQ Six Sigma Black Belt hand book Six Sigma 14
T Sanitary "Green" Restrooms - Hand Driers and Waterless Urinals - Food Manufacturing Sustainability, Green Initiatives and Ecology 20
R Hand tool to measure angles that has been machined into stock Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
A Hand-Out for Human Factors Training Training - Internal, External, Online and Distance Learning 7
F Production Employees and hand gage issues Human Factors and Ergonomics in Engineering 13
L Original Documentation - Need to maintain any original hand written documents ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 28
D Is CE Mark required for Hand Sanitizers to be sold in Europe ? EU Medical Device Regulations 6
M Hand Tool Selection to Remove Microjoints and Slivers off of Small Parts - Advice Manufacturing and Related Processes 6
Q Using Faxed Hand Signatures for Approval Records US Food and Drug Administration (FDA) 17
S How can I die cut foam - I'm thinking of a simple hand machine? Manufacturing and Related Processes 15
A Removing the liner from disposable nicotine patch without using hand ISO 13485:2016 - Medical Device Quality Management Systems 11
R CAPA (Corrective and Preventive Action) Discussions are Getting Out of Hand Nonconformance and Corrective Action 41
T Control Plans - Right hand side of the control plan was a flow chart FMEA and Control Plans 3
Q Qualifying Persons to Reflow Solder by hand - Circuit boards Various Other Specifications, Standards, and related Requirements 3
S Hand Solder Board/Component Temperature Requirements for Repair Various Other Specifications, Standards, and related Requirements 7
K Process Validation - Hand soldering of components & wires on PCB - Special Process? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
Q Hand processes - How can you error proof human systems with out automation? Preventive Action and Continuous Improvement 7

Similar threads

Top Bottom