I am fairly new to gage r&r studies, and I just finished running one on a ph meter, and got horrible results! (94.45 % Contribution and 97.18 % Study Variation)
10 samples of a Buffer solution were used (all were taken from the same source so they were assumed to be equal), three operators performed the measurements, two trials per operator. I also made sure the ph meter was calibrated and that all the proper procedures were followed when measuring.
I really wasnt expecting to get such bad results and I am not sure whether I did something wrong in the design of the test or if the measurement system is really that bad. I have attached a word document with the data, the minitab output for the crossed gage r&r and graphs. (wouldn't let me upload the minitab file...)
I would appreciate any comments/feedback/insights on what went wrong or any help on how to interpret what are the fallbacks of my measurement system. I noticed that basically all the variability is in repeatability, which doesnt seem that logical.
Appreciate your help,
Thank you!
Gabriela
bobdoering
Stop X-bar/R Madness!!
Re: Help in understanding my r&r results
You are probably jumping into GR&R right in the deep end!
I think your first problem is you need more variation in you samples - like 3 buffer solutions - not samples of the same buffer solution. You just don't have enough variation in your "samples" to see if you can detect the difference between them (which is kind of the point of the exercise).
What is the specification you are trying meet with the pH meter?
G
grs2828
thank you, that is a big help! I actually doubted for a long time on whether i should use the same buffer solution or not for all samples.
But this is where I get confused again. When you say to take the samples from 3 different buffer solutions, do you mean three completely different solutions (say one with ph of 5, another 14 and the third 10), or just three different batches of the same solution (all supposed to have the same ph)?
you need to have the full range of variation that you would expect - or you know to have of the ph of your buffer solutions. that is why knowing the allowable range is important. Now if your allowable range is extremley tight and you suspect that you have poor measurement repeatability my suggestion woudl be to make buffers that were just outside of the spec limits and use those in the experiment.
If you still get "lousy" results come back and we can help. all is not lost, you will just need to understand that yoru measurement system is more categorical (yes/no or high medium low) than it is exactly correct out to the nth decimal place.
For information regarding the selection of samples, see
Intro to Measurement System Analysis (MSA) of Continuous Data – Part 5a: R&R, Part Selection and
Intro to Measurement System Analysis (MSA) of Continuous Data – Part 5b: R&R for more information.
As pointed out, your part variation is not even measurable. If this reflects the actual process variation, your gage is not acceptable. Start looking at the gage resolution. Your pH meter
does not have sufficient resolution to be used for process control or statistical studies. This is based on the test for the Range chart: At least 5 distinct levels of measurement within the UCL, and nor more than 25% equal to zero.
If the part variation does not reflect the actual process variation, you need to obtain samples that do reflect it, or use the variation from a capability study. Minitab allows you to enter this information and will calculate another metric using this data.
Depending on your tolerances, it may be acceptable for inspection to a tolerance.
N Help with basic understanding of Competence requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
michellemmm Need help in understanding "Equipment performance in any attitude!! AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 9
K Verification of Purchased Product - Help Understanding Clause 7.4.3 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
S Need help understanding Ford's Phase 0 PPAP APQP and PPAP 1
B Help understanding the ISO/TS 16949 Clauses IATF 16949 - Automotive Quality Systems Standard 3
E Contract manufacturer seeking help understanding ISO 13485 Validation requirements ISO 13485:2016 - Medical Device Quality Management Systems 9
S Need help on "Country of Origin" Medical Device and FDA Regulations and Standards News 0
T Help with BS EN ISO - IEC 80079-34 2020 (Explosive atmospheres QMS) Other ISO and International Standards and European Regulations 0
D Supplier Quality level category help - high level ISO 13485:2016 - Medical Device Quality Management Systems 6
D Amendment Class II - Please confirm if the submission is by email (Canada) - Help Canada Medical Device Regulations 0
T ISO/IEC 17065 certification scheme Help Other ISO and International Standards and European Regulations 7
T Help to Suggest name for a new certification and inspection company Coffee Break and Water Cooler Discussions 7
SocalSurfer AS9100 new certificate, but need QMS software, help Quality Assurance and Compliance Software Tools and Solutions 2
L Manufacturing Process Audit Help IATF 16949 - Automotive Quality Systems Standard 6
M Case study help as per ISO 9001: 2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 18
A Beginners help with ISO3951-2 Combined control s-method n>5 what is Phi ?? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
MDD_QNA Medical Device Software - Is a Help Button required? IEC 62304 - Medical Device Software Life Cycle Processes 1
S Need help with analysing a survey on minitab Using Minitab Software 1
M Help me improve the definition for internal PPM Manufacturing and Related Processes 3
I Help: QFD House of Quality Quality Tools, Improvement and Analysis 1
C New to Site - Request Help with API Q2 Implementation Oil and Gas Industry Standards and Regulations 9
O Please, help with document on ISO/ TS 17758 and (IDF 87: 1979). Thank you. Manufacturing and Related Processes 1
A Class medical device (MDD) - Classification help EU Medical Device Regulations 1
K A proposal for the model Quality Management - I need help for the project ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
A Help with Drawing Review - ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
Q % Study variation low, % tolerance high - GR&R Interpretation help Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
R Probability - Need a help to solve the below question Statistical Analysis Tools, Techniques and SPC 5
M Thanks for the help Coffee Break and Water Cooler Discussions 0
Judy Abbott Can any one help me with an urgent BIOCOMPATIBILITY Impact assessment? Internal Auditing 9
A Help establishing regulatory requirements - Plastic dental cartridge syringe US Food and Drug Administration (FDA) 0
A Medical Device Contract Manufacturing Requirements and Information - Help wanted ISO 13485:2016 - Medical Device Quality Management Systems 5
8 Need Help - Runout - Function Gage Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
T Burn out - Tips to help combat QC burnout Career and Occupation Discussions 9
S New to FAIR, need help in filling it out AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
M Informational From RAPS: Danish Regulators Seek to Help Smaller Companies With EU MDR Compliance Medical Device and FDA Regulations and Standards News 0
M Reliability Testing - Weibull Analysis Help Reliability Analysis - Predictions, Testing and Standards 4
L Help me to decide which test to use - If each proportion belongs to the same population Using Minitab Software 7
M Informational US FDA issued the first warning letter for UDI violations to help ensure compliance Medical Device and FDA Regulations and Standards News 0
T QMS Template Help - Small machine shop/job shop Manufacturing and Related Processes 1
R Pls help --- Need expert advice on Video Measurement Measurement Uncertainty (MU) 0
B Will you please help me with my academic project? Repair & calibration service for test equipment ISO 17025 related Discussions 0
B [Help] Customer Engagement Survey for MBA Project - due Apr19 Manufacturing and Related Processes 0
B Help with Customer Engagement Survey for MBA Project - due Apr19 General Measurement Device and Calibration Topics 0
Q ISO 3310 Clarification Help - Aperture sizes for sieves used for particle sorting Other ISO and International Standards and European Regulations 2