Greetings.
I am finding some issues on understanding the basics of correction to a measurement say for an example in voltage or current measurements.
I have been looking at the calibration of a multimeter , yes the HP34401A and i have set aside 4 factors in determining the uncertainty of the measurement.
First being the type A or the repetitive measurements , the others being the drift or inastability of the calibrator( we are using the 5720A fluke calibrator ) , the calibrator specs and the resolution of the multimeter itself to compute the expanded uncertainty for the measurement.
I know for the Fluke calibrator , there is a spec stating the accuracy such as the ppm of reading + floor . How do we quantify this value? Should we take the ppm value of the readings obtained by injecting the calibrator values( the reading of the dmm?) or the ppm of the nominal value? And say we send this calibrator to an external lab for annual calibration - do we take the CORRECTED reading? How do we apply correction to the reading? Do we calculate the accuracy based on ppm * measured or the corrected value?
I am a but lost here and i have seen and read some journals but i think i am still blur about this.
I checked with a few colleagues and some say , we should not include the corrected measurement values and instead take the value minus the value in the calibration certificate of the calibrator.
Can someone help or point me to the right direction?
And are there any clear examples of which i can get some knowledge in how to apply correction?
Thank you and have a nice day.
Elsmar Forum Sponsor
Stijloor
Staff member
Super Moderator
#2
Re: How do I extend Correction factor of Calibration cert to compute instrument accur
A Quick Bump!
Can someone help?
Thank you very much!!
Thread starter Similar threads Forum Replies Date
A CE Certificate Expiring and NB unable to extend / renew as not in their scope as a NB EU Medical Device Regulations 34
F ISO/CD 9001:2015 - The 6 mandatory requirement (procedures) are extend to 9 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
A Shelf Life Testing - We would like to extend the sterility expiration date Misc. Quality Assurance and Business Systems Related Topics 1
Stijloor Toyota to extend development time to ensure quality World News 0
Y To what extend should training materials be controlled? Document Control Systems, Procedures, Forms and Templates 27
D Determining Calibration Intervals - Extend or change the length of Calibration Frequency (Interval) 11
G What to do with correction factors overlooked by an accredited standard (6.4.11)? ISO 17025 related Discussions 5
G Rework, Reprocess, or Correction - A simple blurred letter on the legible etching ISO 13485:2016 - Medical Device Quality Management Systems 4
G Correction fluid found on documents that are reviewed Qualification and Validation (including 21 CFR Part 11) 11
Q Correction of records - FAR or DFAR? Various Other Specifications, Standards, and related Requirements 3
A Who pays for correction and removal actions? Other Medical Device Regulations World-Wide 2
P Documentation for correction, corrective action, mini CAPA Nonconformance and Corrective Action 9
Uriel Alejandro Evaluation of the effectiveness of correction actions General Auditing Discussions 15
D Non ASTM Partial Immersion LGT emergent stem correction General Measurement Device and Calibration Topics 3
I EO sterilization out of specification and correction Other Medical Device Regulations World-Wide 1
Jane's When audit NC requires only correction ISO 13485:2016 - Medical Device Quality Management Systems 16
J Who can make a Correction of Entry on GMP Records Other Medical Device and Orthopedic Related Topics 1
J ISO 9001:2015 - Clause for Correction of a QMS Nonconformity ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
C Medical Device Malfunction - Customer Complaint - Response vs. Correction Other US Medical Device Regulations 3
Q Corrective Action Effectiveness also for correction? Nonconformance and Corrective Action 11
M Does there always have to be a "correction?" Problem Solving, Root Cause Fault and Failure Analysis 16
K Definition Correction, Corrective Action and Preventive Action - Definition of terms Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 28
A Software Medical Device Bug Fixes: Correction or Corrective Action? Other US Medical Device Regulations 5
L Medical Device Reporting vs Recall/Correction: What action to take? US Food and Drug Administration (FDA) 5
A AS9100 and Record Correction Procedure AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
N Applying Correction Factors to Furnaces Manufacturing and Related Processes 5
M Calculation for Correction Deformation (Pitch Diameter) General Measurement Device and Calibration Topics 3
A Observations - Is Correction / Corrective Action a Must ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
D Correction and Degree of Error - Metrology Definitions and Differences General Measurement Device and Calibration Topics 1
B How do I calculate Correction to Caliper Calibration Data General Measurement Device and Calibration Topics 4
Q Correction Action as Preventive Action at another Plant? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
S Correction and Removal - Include Device used in Non-Production Environment? US Food and Drug Administration (FDA) 5
Q CAPA (Corrective Action and Preventive Action) after Correction always? Nonconformance and Corrective Action 10
B Recall vs. Field Correction - Field Corrections instead of a Recall US Food and Drug Administration (FDA) 25
Q Corrective Action - Disposition always Correction? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S 3 Wire Pitch Dia. Method - Measuring Force Correction & Lechowski?s method General Measurement Device and Calibration Topics 1
M Definition Corrective Action vs. Correction: Definitions Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 18
D NLEC (Non-Linear Error Correction) vs. SLEC (Segmented Linear) Calibration Technique Calibration and Metrology Software and Hardware 3
S Correction vs. Corrective Action - Differences Nonconformance and Corrective Action 30
F Zahn Cup Correction Chart - Boeing specification BSS 7234 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
C Medical Device Reporting procedure - Correction and Removal (21 CFR 806) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
V Can a non-conformance be closed with "correction"? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 15
E Correction and Corrective Action - ALWAYS DOCUMENT BOTH Nonconformance and Corrective Action 13
M What is the indication of "adjusted scale correction" to calibration of liq thermom.? General Measurement Device and Calibration Topics 5
H Definition Correction vs. Corrective Action vs. Preventive Action Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 2
G Whats the difference in applying error factor and correction factor? General Measurement Device and Calibration Topics 1
M Correction for Part/Master material difference in gage block comparison testing General Measurement Device and Calibration Topics 2
K Correction, Corrective Action, and Preventive Action differences - Internal Audits Internal Auditing 10
Z Apogee V2.5 - No correction mapping unless I pay $8,000+ Software Quality Assurance 6
D FDA Requirements about using Correction Fluid on Documents and Records US Food and Drug Administration (FDA) 8
Similar threads
Top Bottom