Hi all,

We are planning to put a settle plates in our cleanroom as part our monitoring plan. The question is how do I set a limits to cfu, how much is ok? Our clean room is ISO class 7. I can't find anything in the standards, I need to justify cfu limits somehow.

Thanks,
Alex.
 
Elsmar Forum Sponsor
M

MIREGMGR

#2
I can't find anything in the standards
Where are you looking? The ISO 14644 standard series includes CFU-count-based definitions for "cleanroom" levels.

Our clean room is ISO class 7.
Says who, according to what standards-based evaluation? If you or someone else has determined that it's ISO Class 7, someone must already have evaluated it in regard to ISO 14644. If you won't have any such evaluation-results available to you, you might want to stop considering it to conform to a particular ISO level until you've brought it into control, which includes regular measurement.
 
Thread starter Similar threads Forum Replies Date
J How to Set Microbial Limits in Food Supplements Food Safety - ISO 22000, HACCP (21 CFR 120) 3
E How to use the Standard Error and how to set the error limits? Reliability Analysis - Predictions, Testing and Standards 2
B How to set Cleanroom Alert and Action Limits as part of Microbiological Evaluation Quality Assurance and Compliance Software Tools and Solutions 8
S How to Set Acceptable Calibration Limits for Measurement Instruments? General Measurement Device and Calibration Topics 11
C SPC: How to set up control limits on SPC chart based on historical data? Statistical Analysis Tools, Techniques and SPC 7
F Pre-Control Chart - Limits are set at 2/3 of the specification Statistical Analysis Tools, Techniques and SPC 6
W Regarding Requirement for Hi pot & Leakage Current test Station Set-up IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
G CE marking of MD and IVD set EU Medical Device Regulations 3
qualprod To set frequency to review documents in ISO 9001 7.5? Document Control Systems, Procedures, Forms and Templates 13
M IATF 16949 8.5.1.3 Verification of job set-ups - Do we need secondary check? IATF 16949 - Automotive Quality Systems Standard 7
B Set RH spec from dewpoint spec General Measurement Device and Calibration Topics 2
S Is Optical Parallel Set required calibration? ISO 17025 related Discussions 1
R Gauge R&R on pre-set torque wrench Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
A How to set up Continuous CpK monitoring of an injection mold process Reliability Analysis - Predictions, Testing and Standards 7
N Internal Audit Schedule – Who gets to set the schedule? Internal Auditing 16
G Comparing Gauge Blocks - Translating the calibration between the master set and the QA set General Measurement Device and Calibration Topics 2
L How to evaluate the process capability of a data set that is non-normal (cannot be transformed and does not fit any known distribution)? Capability, Accuracy and Stability - Processes, Machines, etc. 12
N When to record values if a spec for a mfg process is set Manufacturing and Related Processes 3
Marc Time Zone - Set your Time Zone Elsmar Xenforo Forum Software Instructions and Help 1
D As a newcomer to QMS, I really appreciate how ISO is set up. ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
C IATF 16949 Cl 8.5.1.3 - Verification of job set up IATF 16949 - Automotive Quality Systems Standard 4
D How to set equipment calibration/verification criteria General Measurement Device and Calibration Topics 1
S GRR on ID with set of pin gauges Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
G What to Check - Set Up vs. In Process Inspections Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
N The known set of all root causes Nonconformance and Corrective Action 11
G What degrees of freedom means in uncertainty budget and how do I set the value Measurement Uncertainty (MU) 2
D What is the exact Temperature Coefficient Value to be set on a Conductivity Meter? General Measurement Device and Calibration Topics 2
G Training Employees about Quality - Philosophy / Mind set Misc. Quality Assurance and Business Systems Related Topics 1
renenatasha MDD Classification for Pressure Monitoring Set CE Marking (Conformité Européene) / CB Scheme 1
U Are all manufacturers required to set up eMDR system? Other US Medical Device Regulations 4
Sean Kelley Gage Block Verification or buy a new set (cost aspect) General Measurement Device and Calibration Topics 1
C Implementing a set of bore gages at a machine Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
S How to set up the Accupoint 2 ATP sanitation monitoring system Quality Assurance and Compliance Software Tools and Solutions 1
4 Quality vs Production on CMM Set Up Requirements Manufacturing and Related Processes 1
A What is the definition of Installation that requires Record Keeping vs Set-up 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
G AS9120 - What should we be expecting once we set forth into becoming certified? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
C ISO 9001:2008 Audit Major Finding - Dial Calipers not set to zero ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
A IEC 60601-2-24 Question: Free Flow Clamp on Set? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C Lowest Acceptable Grade for Master Gage Set Calibration and Metrology Software and Hardware 1
L How to set up a basic electronic test equipment calibration laboratory General Measurement Device and Calibration Topics 3
Q Inquire TS 16949 standard - Clause 7.5.1.3 Verification of 'Job Set-Up' IATF 16949 - Automotive Quality Systems Standard 7
M AS9100C Document Control for Machine Set Up Work Instructions AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
J Appropriate In-Process Monitoring Frequency for 2X per day Change Set-Up. Statistical Analysis Tools, Techniques and SPC 3
S Insufflation Tubing & Filter Set Product Classification question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
M New Calibration Laboratory Set Up Measurement Parameters and Equipment General Measurement Device and Calibration Topics 4
F How to set-up a Dual Company QMS (Quality Management System) Quality Manager and Management Related Issues 2
R Can the same Biocompatibility Report for Infusion Set be used for its Components? ISO 13485:2016 - Medical Device Quality Management Systems 7
insect warfare How is the Management Review structure of your company set up? Management Review Meetings and related Processes 5
AnaMariaVR2 Legal Flash: 8 Regulations set to impact the Pharma & Biotech Industry in 2013 US Food and Drug Administration (FDA) 0
F Receiving a new Gage Block Set with an old Certificate Calibration Frequency (Interval) 3

Similar threads

Top Bottom