How do Medical Device Standards vs Directive vs Regulation technically vary?
G
Gargi
Hello Elsmar,:bigwave:
How do Standards , Directive and Regulations (in medical device aspect) vary technically amongst each other.
I did a lot of search earlier but couldnt arrive at firm and accurate distinction.:nope:
Any help is genuinely appreciated.
Thanks and regards,
Gargi
Sarah Stec
Re: Standards v/s Directive v/s Regulation
Is your question as to the kind of information they contain?
If so, in my opinion, you can look at standards as more detailed or technical version of the directives and regulations. The directives and regulations give you the requirements; the standards point you in the direction (and sometimes outright tell you) of meeting the requirements (for harmonized standards). So, directives and regulations don't have a lot of technical information, whereas standards do.
As an example, a directive or regulation is going to say, "the device must be safe." If you think about it, this is actually an incredibly broad statement. So the standard is going to tell you "this is a safe voltage, this is how you test the device to make sure it's safe, this is how many samples you need to test, these are the results you're looking for, etc."
Directives and regulations are types of EU pieces of legislation (well, directives are, and regulations are universal, but we'll focus on the EU). A directive is directed towards the member states, and must be adopted and transposed by each member state into its own laws. For medical devices, most of the MDD was transposed exactly how it is, but the member states have differing provisions for things like device registration. A regulation, on the other hand, is effective upon all member states exactly how it is written; there is no need to transpose the regulation into member state's codes because it is already legally binding as it is.
There are a lot more differences between standards and the regulatory texts (who creates them, who follows them, who adopts them, harmonized vs. not-harmonized standards, effects of country-specific standards, etc.). Did you mean something else? :cool:
Ronen E
Problem Solver
Moderator
Generally, regulations are elaborations of laws or established hi-level policies. They're usually drafted not by the legislating authority (eg Parliament) but by an executive body (eg the FDA, the Minister of health etc.) empowered to do so in/by the relevant law.
In the EU, "Regulations" are a specific type of legislative instrument, as Sarah has explained. "Directives" are another type of legislative instrument (as Sarah has explained as well).
Published standards (assuming that this is what you were referring to) are exactly that - a formalised written expression of some (usually broad) agreement on a topic. Some regulations (and some EU directives) rely on specific standards instead of going into a lot of detail, but published standards don't necessarily relate to any regulatory / legal process or requirement. Some standards are just a reflection of a broadly-accepted "state of the art", while others are just tools (among others).
G
Gargi
Hi Sarah,
Thank you for detailed info.
However consider this case:
Directive 93/42/EEC (Medical device directive ) and Standard IEC 60601-1 (basic safety and essential requirements) both explain in detail the essential requirements of medical device.
I know that MDD is concerned with just Europe and IEC is a global standard.
Then on what grounds can I differentiate them.
I appreciate if you could clarify me on in this aspect
Thank and regards
Gargi
G
Gargi
Hi Ronen,
Thank and regards
Gargi
Marcelo
The most basic answer is, standards are generally voluntary and regulations (which include directives, ordinances and anything published by a government body) are mandatory.
Regulations usually details general expectations/requirements (what) and standards usually details application of those (how) or more detailed requirements related to the general expectations requirements.
A simple example, EU regulations are based on the concept of essential requirements, which are high level expectations. One essential requirement (in 12.6. Protection against electrical risks) is that "Devices must be designed and manufactured in such a way as to avoid, as far as possible, the risk of
accidental electric shocks during normal use and in single fault condition, provided the devices are installed correctly".
This essential requirement (as most of the others) is so generic that it's not easy to comply with. Anyway, you can comply with any good engineering solution that does fulfill the requirement.
One way to comply is to comply with the more detailed technical requirements of IEC/EN 60601. More specifically, you would need to comply with 8 * Protection against electrical HAZARDS from ME EQUIPMENT .
So, what the standard does in this case is to clarify, in a more explicitly way, how to comply with the more generic essential requirements.
For more on the link between directives and standards, please read the the 'Blue Guide' on the implementation of EU product rules.
G
Gargi
Thank you Marcello for that indepth yet precise distinction between standards and regulations.
I am really Glad that Elsmar reopend again.
I believe that ELSMAR IS PHOENIX that would always rise back in any adverse situation.
I wish ELSMAR to live long.
Best Regards,
Gargi
Sarah Stec
Marcelo's explanation is a good one. I would add a few corrections to what you say above to hopefully help your understanding:
- The MDD states (not explains) the essential requirements of medical devices. However, as Marcelo stated, these essential requirements are generic and very broad because they apply to all medical devices, from gauze to pacemakers. You could imagine that proving a piece of gauze or a bandage is safe and effective requires different information than proving a pacemaker is safe and effective. This is why there are standards.
- The harmonized standards (like EN/IEC 60601-1) explain in detail how the medical device for which the standard was written could meet the essential requirements.
-Another way to differentiate them: The MDD is mandatory to follow, whereas IEC 60601-1 is not. You can comply with IEC 60601-1, but if you don't follow the rest of the MDD requirements, then you aren't compliance with the MDD. Directives =/= Standards.
I hope this is helpful!:cool:
J Medical device related standards or regulations - Disposed records log ISO 13485:2016 - Medical Device Quality Management Systems 2
Q Interpreting China Medical Device regulations/standards China Medical Device Regulations 1
C How medical device manufacturers are implementing standards like GDPR and HIPAA Other ISO and International Standards and European Regulations 5
S Keeping track of the changing Medical Device Standards Other Medical Device Related Standards 3
L Medical Device Electrical Safety Standards - Gap Analysis of Latest Revisions Other ISO and International Standards and European Regulations 2
R Standards Specific to Single Channel ECG Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C CFDA announced 120 new Medical Device Standards - July 2014 China Medical Device Regulations 0
N Regulatory and Safety Standards for Medical Device with Measuring Function IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Applicable Medical Device Standards for Home Use Blood Pressure Devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T The CFDA recently issued new Medical Device Industry Standards China Medical Device Regulations 0
M Revision of ISO 16142 - Clause on use of standards in the medical device life-cycle Other Medical Device Related Standards 3
Sam Lazzara Medical Device Standards Identification 101 article Other Medical Device Regulations World-Wide 2
P Specific requirements for filtering in Medical Device Standards IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
P Which Medical Device Standards to Buy IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
N Medical Device External Standards Control and Monitoring Other US Medical Device Regulations 10
M Service for Supersede Dates of Medical Device Standards? Other Medical Device Related Standards 3
R Medical Device Labeling Standards References 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
Marc Medical Device Standards - Difference between IEC and EN Standards Other Medical Device Related Standards 13
B Definition Diagnostic - Is "diagnostic" defined in the EU Medical Device Standards? Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 10
M Medical Device Regulatory Standards Update Services IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
S Canada - CSA C22.2 Medical Device Standards Other Medical Device Related Standards 1
D Standards concerning Medical Device Battery Charging DURING Use Other Medical Device Related Standards 3
Sam Lazzara Medical Device Harmonised Standards Update - May 2011 CE Marking (Conformité Européene) / CB Scheme 0
M Clarifying the many different Medical Device national and international standards Other Medical Device Related Standards 4
O Harmonized Standards on Medical Device IFU (Instructions for Use) ISO 13485:2016 - Medical Device Quality Management Systems 7
xcanals_tecno-med.es Revised List of Harmonized Standards for Medical Device Directives (MDD) in the EU EU Medical Device Regulations 48
B Define "Rework" Under Medical Device Standards ISO 13485:2016 - Medical Device Quality Management Systems 11
xcanals_tecno-med.es Revised list of harmonized standards for medical device directives in the EU (7/7/10) EU Medical Device Regulations 0
L ISO 13485 - Overwhelmed by various Medical Device standards ISO 13485:2016 - Medical Device Quality Management Systems 2
A Guidance Standards for Medical Device Manufacturer Pest Control Other Medical Device Related Standards 6
S Searching for applicable Standards for a Medical Device using electronics, software Other Medical Device Related Standards 11
R Medical Device: New List of Harmonized Standards in EU under MDD93/42/EEC EU Medical Device Regulations 6
M Revised list of harmonized standards for medical device directives in the EU EU Medical Device Regulations 1
M Implementation period for standards - CE renewal submission - Class II medical device EU Medical Device Regulations 3
S Standards used for Medical Device MDD Technical Files ISO 13485:2016 - Medical Device Quality Management Systems 25
S Relevant Packaging Standards for Medical Device Package testing ISO 13485:2016 - Medical Device Quality Management Systems 10
W New List of Medical Device Harmonized Standards - MDD CE Marking (Conformité Européene) / CB Scheme 5
S Environmental Control Standards - Sterile medical device manufacturing environment ISO 13485:2016 - Medical Device Quality Management Systems 3
M If there is any restriction on the transportation of Used Medical Device (with blood) from Canada to other countries? Canada Medical Device Regulations 4
M If there is any restriction on the transportation of Used Medical Device (with blood) from EU to other countries? EU Medical Device Regulations 2
adztesla Sterile packaging validation for medical device - ASTMD4332 and ISTA2A ISO 13485:2016 - Medical Device Quality Management Systems 3
P On-going stability study requirement for medical device US Food and Drug Administration (FDA) 2
N Regulations for Medical Device in Jamaica Other US Medical Device Regulations 0
M Classification of device according to China medical device regulations China Medical Device Regulations 4
adztesla Design change and Process change Class 3 medical device Medical Device and FDA Regulations and Standards News 2
M Importing a custom component for our medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
adztesla Adding new supplier/ Medical device design change ISO 13485:2016 - Medical Device Quality Management Systems 2
adztesla Medical device Drawing related change justification Other Medical Device Related Standards 0
O Medical Device With 2 Power Cords IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
Stoic Which medical device companies are using Recombinant Factor C (rFC) instead of Limulus Amebocyte Lysate (LAL) for endotoxin testing? Sustainability, Green Initiatives and Ecology 0