How do you calculate Overall Residual Risk - Requirement of 14971 Risk Management
F
fuji033 - 2006
I am struggling to understand one requirement of 14971 Risk Management. How do you calculate overall residual risk otherwise known as aggregate risk. Any guidance?
Al Rosen
It's not a calculation. You have to weigh the severity against the probability of an occurrence. It is somewhat subjective.
Jerome
It's like Al Rosen says, but you can "make" it a calculation.
Depending on the techniques you use (I use FMEA) you can ascribe a value to the different risk elements.
To facilitate the (FMEA) risk assessment an estimate is needed of the
• Severity of the Effect (S) of a failure mode
• Likeliness of Occurrence (O) of a failure Cause
• Chance that the Effect or Failure mode is not Detected by means of current controls in place (Df)
These estimates are translated into a numerical value by means of a standardized approach. (I use a 1 to 5 rating for each)
The Risk (RPN) is subsequently quantified as follows
RPN = SEVERITY x OCCURRENCE x DETECTION Failure
RPN = S x O x Df
This risk product number (RPN) is now a numeric representation of the calculated risk.
If this number is above a preset cutoff point (I use 15) I oblige myself to take risk mitigation measures.
Then I calculate again what the RPN is when the risk mitigating measures are implemented.
This can be seen as the "calculated" Overall Residual Risk.
It's one of many way's to make something this subjective and almost abstract, tangible by means of numbers.
Hope it helps.
(just my 50 ct's)
Best regards,
Jerome
Sanne
Hey there,
I just had this discussion with the European Authorities (TÜV) during our last audit.
They don't accept for example RPNs (or whatever you use) as a measure for the "allover residual risk" since the RPN belongs only to a single risk and not to the allover product risk.
We even have a statement in our Risk Management Procedure saying that the allover residual risk is considered acceptable when all single risks are acceptable.
Even this is not enough for TÜV.
This time they didn't write down a finding (the reason: "We don't know how you could do it better"). But they were not satisfied and will look at this topic again next year :(
Does anyone have an idea how to solve this problem?
Cheers,
Sanne
Marcelo
I don´t. It´s eird that they do not know how you could do better and not accept what you do.
Please note that you can use some sort of general RPN or the like for the general - overall - risk, but you usually have to perform a different analysis for the overall risks.
Anyway, D.7 of ISO 14971 details some approachs to evaluate overall residual risks.
Burgmeister
I would contact your NB again and ask for their advise. If your methods aren't suitable they must have people whose methods are. If they can't offer guidance I would seriously consider changing to a more helpful NB.
Sanne
I guess I won't change anything for the moment- if they write a NC in the next audit, they have to provide me with guidance what they expect to see.
It seems like they were just re-trained on risk management because they wanted us to include "bleeding" as a potential hazard for implantation!
What a surprise that there will be blood in an operating theater during implantation of devices...
Anyway thank you all for your opinion and advise :)
davidespada - 2011
Risk Management is always a pain in the neck: everyone require it but nobody can show examples about how to do it!
The way we do it is the following:
We calculate the Risk Class by Impact x Likelihood for each hazard, we make a plan to reduce it and re-calculate the Risk Class.
Now, the question is (and I guess is what TUV ask for): is the Device safe or not? We assess the Risk Class of the Device (what I call Irreducible Residual Risk) by stating the the risk of the device simply equals the higher residual risk.
So if I have an hazard with a residual risk as "moderate", the risk of my device is moderate and so on.
Of course this is equal to say that "allover residual risk is considered acceptable when all single risks are acceptable", but you provide a quantitative analysis, which is what I guess TUV is asking for.
Ciao,
BR
Davide
S How to Calculate Overall FPY (First Pass Yield) Quality Tools, Improvement and Analysis 2
T Attribute MSA-AIAG 4th How to calculate sample size IATF 16949 - Automotive Quality Systems Standard 12
H How to calculate sample size and sample frequency Statistical Analysis Tools, Techniques and SPC 2
T Calculate FPY for Audit Results AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 9
D How to calculate the A4 constants in Median/Range control charts? Statistical Analysis Tools, Techniques and SPC 2
S How to calculate Effective Number of People for ISO 13485 Certification? General Auditing Discussions 2
A Need to calculate tolerance Intervals with a set of non-normal data and 3-Parameter Weibull distribution Using Minitab Software 0
U Calculate acceptance of equipment reading after calibration General Measurement Device and Calibration Topics 2
U Calculate acceptance of calibration reading Measurement Uncertainty (MU) 1
S How would you calculate this? Capability, Accuracy and Stability - Processes, Machines, etc. 4
S Bend allowance formula to calculate per degree per radii Manufacturing and Related Processes 2
P Commercial excel templates that calculate overhead, hourly rates, fee proposal Excel .xls Spreadsheet Templates and Tools 0
M How to calculate benefits? Moving some developed non controlled software/automation systems Service Industry Specific Topics 2
qualprod Calculate number of people in processes, considering cycle time/takt time = quantity of people Lean in Manufacturing and Service Industries 9
S How to Calculate the producer Risk AQL - Acceptable Quality Level 25
qualprod Methods to Calculate Available Production Rate ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 0
K Why calculate LCL in P-chart? Statistical Analysis Tools, Techniques and SPC 4
B How to calculate CPK with n = 1 Manufacturing and Related Processes 2
A Add-On software for VMC to calculate setting time, Idle Time, OEE etc.. Manufacturing and Related Processes 2
L How to calculate the Probability of Deficiency of the AQL of an Inspection AQL - Acceptable Quality Level 16
V How to Calculate Unbiasing Constant C4 Statistical Analysis Tools, Techniques and SPC 1
J How to Calculate OEE Indicator for a Workshop (15 Machines) Lean in Manufacturing and Service Industries 6
R How to calculate UCL (Upper COntrol Limit) and LCL (Lower Control Limit) Statistical Analysis Tools, Techniques and SPC 5
T I want to calculate reliability at 750 hours Reliability Analysis - Predictions, Testing and Standards 8
R How To Calculate & Print CDF and Lower Bound CDF value for some t(time) with Minitab? Using Minitab Software 1
N How to calculate the Premium Freight charges Service Industry Specific Topics 4
A How to calculate Mean and standard deviation without sample size in Minitab Using Minitab Software 5
R When do I need to calculate Process OEE? Manufacturing and Related Processes 12
kedarg6500 How to calculate Thread Pitch Diameter for non standard TPI like 5/8" 17.75 NF3 Manufacturing and Related Processes 3
C Definition and Way to Calculate some Metrics Lean in Manufacturing and Service Industries 6
M How calculate P value for Linearity and Bias using Excel Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 8
N Formula that will calculate the NoGo Major Diameter of an M20 x 1.5 6H Thread gage Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
M How to calculate Cg and Cgk for a 2-Dimensional Area Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
T Calculation Question - Calculate the runner weight for a part Manufacturing and Related Processes 2
M Spreadsheet on How to Calculate SMT Capacity wanted Manufacturing and Related Processes 2
F How to Calculate the unbiased Standard Deviation in Excel using Minitab Formula Capability, Accuracy and Stability - Processes, Machines, etc. 6
S Spreadsheet that will Calculate Audit Duration times based on IAF MD5 and MD11 Excel .xls Spreadsheet Templates and Tools 8
M Calculate PPMs by going back to Date of Production or Date Noticed Quality Tools, Improvement and Analysis 7
M How to calculate MTTF (Mean Time to Failure) Reliability Analysis - Predictions, Testing and Standards 1
H How to calculate the Lot Reject Rate Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
L How calculate the number of samples Statistical Analysis Tools, Techniques and SPC 2
A Simple 125 piece part Capabilty Excel .xls sheet to calculate Cpk Capability, Accuracy and Stability - Processes, Machines, etc. 3
R How to calculate MMC when is placed after datum Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
S Which one is better to calculate, Cpk or Ppk? Statistical Analysis Tools, Techniques and SPC 11
C How to calculate P value on Probability Plot Statistical Analysis Tools, Techniques and SPC 1
S Supplier Disruption Score and how to calculate it Quality Tools, Improvement and Analysis 4
D IAQG OASIS - New tool to calculate Audit Time AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
M How to calculate dimensions of trapezoid ? Quality Tools, Improvement and Analysis 3
J How to calculate Surveillance Audit Days in Corporate Audit Scheme IATF 16949 - Automotive Quality Systems Standard 4