How does IEC-60601-1 apply to a non-medical device in the patient vicinity?

#1
I am developing a system which is not a medical device. It is a data system for collection of vital signs from patient monitors in a hospital. It is not a medical device per the FDA's own definition (see MDDS rule).

As part of this new system, I want to put an off-the-shelf serial-to-ethernet converter at the patient bedside (such as the IOLAN SDS1 M Secure Device Server). This device will connect to the RS232 port on the patient monitor and transmit the RS232 data (encrypted, of course) across the network to my server.

I am confused about how 60601-1 applies to such a design. To save costs, I'd prefer to use a non-60601-compliant serial-to-ethernet converter. Could I just power it using a 60601-compliant power supply?

And the loaded question is: Would my system need to be tested for 60601-1 compliance in order to be sold to a hospital?
 
Elsmar Forum Sponsor

Peter Selvey

Leader
Super Moderator
#2
Technically a non-601 power supply is OK as long as additional measures are taken to make sure the leakage meets the 601 limits inside the patient environment. In the standard they talk about extra earth bonding or isolation transformers to limit the leakage, but one of the simplest methods would be to select a double insulated power supply that already has low leakage (<0.1mA to the secondary in normal condition), and make sure it is powered directly from a wall socket, not from expander board (i.e. via the instructions for use, installation section).

From memory the standard for power supplies might have a 0.25mA leakage but most manufacturers will stay well below that, so it's possible to find one that is <0.1mA off the shelf.

You could also just source a 601 power supply anyway, which should be below the 0.1mA limit.

The system itself does not need to meet 601 standard. However make sure the power supply meets the relevant standards for a power supply. And if the power supply has an instructions for use, make sure to include that as part of the package you supply to the end user.

This is a quick summary/interpretation of Clause 16 in the 601 standard.
 
Thread starter Similar threads Forum Replies Date
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
shimonv IEC 60601-1-8 - When does an error message become an alarm signal? IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
MDD_QNA IEC 60601-1 Ed2 vs Ed3 - Does a product that conforms to Ed3 also fulfill Ed2? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-27: What does "linear within ±20% of the full scale output" mean? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
E What does "other APPLIED PARTS" mean in IEC 60601-1 ed 3.0 8.7.4.7 b) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R Does long-term body worn medical monitoring device need to meet IEC 60601-1-11? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S IEC 60601-1 3rd Gap analysis template TO BUY - Does anyone know where? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
D Will buyers (hospitals) require IEC 60601, even when FDA does not? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R Does the Banana connector deviate from the IEC 60601-2-25? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
C Does IEC 60601-2-49 apply to Thoracic Bioimpedance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Does a 12V DC device need EMC (IEC 60601-1-2) Testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A Does Canade recognize IEC 60601-2-22? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
MDD_QNA QR Code Standard ISO/IEC 15417:2007 - Does anyone use it? Other Medical Device Related Standards 3
S Does IEC 62304 require documenting unresolved anomalies for all safety classes? IEC 62304 - Medical Device Software Life Cycle Processes 4
M Does Calibration to ISO/IEC 17025 conform to Z540.3? ISO 17025 related Discussions 1
M Does ISO/IEC 17021 Accreditation work for 13485 ISO 13485:2016 - Medical Device Quality Management Systems 2
L Does IEC 61010-1 apply for Surgical Microscopes? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Does ISO/IEC 17025:2005 require the Laboratory to have separate ISO 9001 Procedures ISO 17025 related Discussions 4
T Does anybody know about EN45014? (Replaced by ISO/IEC 17050-1) Other Medical Device Related Standards 1
A Sampling - What does it mean by Sampling in ISO:IEC 17025 ISO 17025 related Discussions 1
R Does ISO/IEC 17025 supersede ANSI/NCSL Z540 ISO 17025 related Discussions 6
A Does anybody have an audit check list for ISO/IEC 17025 to replace ISO/IEC Guide 25 ISO 17025 related Discussions 14
G When does containment event not require 100% sort? Nonconformance and Corrective Action 5
H Does this clause mean i must be in compliance with AS9100 or is 9001 enough? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
H Does ISO state in anyway that we MUST keep physical first piece parts? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
F Change in address for critical supplier - does NB need to be informed? CE Marking (Conformité Européene) / CB Scheme 19
S Brexit ISO 13485:2016 + Corrigendum - What does a UKCA DoC require? EU Medical Device Regulations 2
G Does anyone know about tobacco-free nicotine pouches? US Food and Drug Administration (FDA) 0
cgaro62 Does FDA apply to a non-medical 13485 certified custom manufacturing company? ISO 13485:2016 - Medical Device Quality Management Systems 11
R Does anyone use iQMS for their ISO based document control? Manufacturing and Related Processes 1
V How many hrs does it take to create a Technical File (TD)? EU Medical Device Regulations 4
M How does FDA determine OAI and VAI after inspection? ISO 13485:2016 - Medical Device Quality Management Systems 0
G Does TPI agencies comes under outsourced process as per Q1 Oil and Gas Industry Standards and Regulations 11
N FPY requested for an industry that does not have rework IATF 16949 - Automotive Quality Systems Standard 4
William55401 21 CFR 11 Password Aging - What does "periodically" revised mean in your org? Qualification and Validation (including 21 CFR Part 11) 3
I CQI-15 Welding System Assessment 2nd edition - Does not require Ultrasonic Welding anymore? Customer and Company Specific Requirements 0
Ajit Basrur Does "Refurbishment" fall under "Service"? ISO 13485:2016 - Medical Device Quality Management Systems 13
T Does marketing company require CE mark if manufacturer has CE mark on product? EU Medical Device Regulations 5
Watchcat REGULATORY WATCHCAT De Novo Post-Mortem 6 – How Does This Even Happen? Other US Medical Device Regulations 0
T What does VW mean by annual requalification? Customer and Company Specific Requirements 4
I Does BSI require suppliers to be ISO 9001 Certified? EU Medical Device Regulations 12
Watchcat Does Regulatory Affairs put the "no" in innovation? Other US Medical Device Regulations 14
Q Does anyone know what this call out means? Manufacturing and Related Processes 3
K COPLANARITY: Composite profile tolerance on multiple surfaces- what does" lower dimensional reference frame tolerance" control? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
P Does FDA require certification for quality system internal audit for auditor? Qualification and Validation (including 21 CFR Part 11) 1
H Does anyone has feedback on a notified body for small companies? EU Medical Device Regulations 3
M Does 4.5 - Alternative RISK CONTROL apply to the Particular Standards? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Does IVD need to integrate with hospital IT infrastructure? Medical Information Technology, Medical Software and Health Informatics 2
D What does a level 1 (PSW) PPAP actually promise? APQP and PPAP 19
R GS1 DataMatrix code does not meet GS1 Specification EU Medical Device Regulations 0

Similar threads

Top Bottom