Hi,
How does Non-Linear UUT affect TAR calculation?
For example, I have an D/P Air Flow transmitter that performs the square root extraction from measuring D/P and outputs a flow CFM reading.
Parameters are:
INPUT: 0 to 1.3622 "WC
OUTPUT: 0 to 1760 CFM
Tolerance 17.6 CFM (Basically, 1% of Range)
What minimum test standard is needed in order to simply test the instrument with at least as good of standard as the UUT?
In this specific example, how does the square root calculation affect the TAR calculation?
Thanks,
Eddie