How does Non-Linear UUT affect TAR (Test Accuracy Ratio) Calculation?

E

Eddie Faulkner

#1
Hi,

How does Non-Linear UUT affect TAR calculation?

For example, I have an D/P Air Flow transmitter that performs the square root extraction from measuring D/P and outputs a flow CFM reading.

Parameters are:
INPUT: 0 to 1.3622 "WC
OUTPUT: 0 to 1760 CFM
Tolerance 17.6 CFM (Basically, 1% of Range)

What minimum test standard is needed in order to simply test the instrument with at least as good of standard as the UUT?

In this specific example, how does the square root calculation affect the TAR calculation?

Thanks,

Eddie
 
Elsmar Forum Sponsor

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#3
If using TAR instead if TUR, then you will first need to define the UUT over its range. That is, an initial curve based on manufacturer's specs. Then the calibration standards can be charted using specific data - with uncertainties - from the calibration of the standards.

Once you have that, and for simplicity that may need to be done by range breakdown, then the information can be put into Excel. That in turn will give a visual. Specific calculations for the points selected can easily be completed.

Now you are ready for actual calibration of the UUT, and then replot the points for the UUT. You can specify that the calibration provider give you the data for the points rather than a pass/fail.

You will also need to establish minimum acceptable TAR to know if you met requirements.

Hope this helps.
 

dgriffith

Quite Involved in Discussions
#4
Hi,

How does Non-Linear UUT affect TAR calculation?

For example, I have an D/P Air Flow transmitter that performs the square root extraction from measuring D/P and outputs a flow CFM reading.

Parameters are:
INPUT: 0 to 1.3622 "WC
OUTPUT: 0 to 1760 CFM
Tolerance 17.6 CFM (Basically, 1% of Range)

What minimum test standard is needed in order to simply test the instrument with at least as good of standard as the UUT?

In this specific example, how does the square root calculation affect the TAR calculation?

Thanks,

Eddie
Been thinking about this since it was posted. You can use a rotameter transfer standard calibrated to 0.5%, since most cannot be read with the human eye to better than 1%. Why waste money on a better calibration.
If the output units are correct, then the measured input differential pressure is likely correct. Since the result is in CFM, wouldn't you calibrate in CFM? Who cares about linearity? :mg:
Okay, we all do, but I don't see it as important here. It's already in the CFM result. When you compare CFM to CFM, you get CFM error. The rotameter is non-linear internally (volume vs. piston size), but the scale you read the piston against is linear.
Your TAR is the 0.5% vs. the 17.6 of range. But that's not the best way to do it, as TAR is fast becoming an obsolete parameter of quality. The TAR would be the 1% UUT tolerance / the rotameter 0.5% uncertainty, for a 2:1 TAR.
You certainly could find a better instrument at more cost, or you could have the cal done at a vendor that uses a primary standard.
:2cents:
 
Last edited:
Thread starter Similar threads Forum Replies Date
L How to evaluate the process capability of a data set that is non-normal (cannot be transformed and does not fit any known distribution)? Capability, Accuracy and Stability - Processes, Machines, etc. 12
N Does 13485 auditor have a right to inspect non-CE products? ISO 13485:2016 - Medical Device Quality Management Systems 5
R RoHS for Medical Devices - Does the directive apply to non-EEE accessories? RoHS, REACH, ELV, IMDS and Restricted Substances 1
M AS9100 First Article - Does 7.5.1.1 revC or 8.2.4.2 revB apply to Non Aerospace Goods AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 19
K How does an Internal Auditor document non-ethical or illegal practices Internal Auditing 10
B Does someone use APQP for non automotive industrie? IATF 16949 - Automotive Quality Systems Standard 2
T Non-Conformance Reports - Does the standard specifically say what an NCR is called? Nonconformance and Corrective Action 28
J Shipping from non-registered sites, how does it affect my registration? Registrars and Notified Bodies 7
B Non-sterile - Does anyone know of a universal symbol? ISO 13485:2016 - Medical Device Quality Management Systems 5
T Does anyone have any ideas/samples of non-conformance reports? Document Control Systems, Procedures, Forms and Templates 4
L Does a backdate form format can be changed if wrong revision is used? Document Control Systems, Procedures, Forms and Templates 8
B General Motors and Honda Alliance - What does this mean to suppliers? IATF 16949 - Automotive Quality Systems Standard 3
C ISO 13485 :2016 - CAPA - Does every CAPA need to be checked by regulations? ISO 13485:2016 - Medical Device Quality Management Systems 9
A Does ISO 9001:2015 cover all the requirements of ISO 10012:2003? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
N FDA UDI - Label vs. Labeling - Does the insert need to include UDI? Other US Medical Device Regulations 0
A Does anyone have a checklist of API Spec 650 13th Edition? Oil and Gas Industry Standards and Regulations 0
D Does Manufacture can submit CE mark application under MDD with NB for his New product after May 2020? EU Medical Device Regulations 3
A What does this sentence "this symbol shall be used in the orientation shown" mean in ISO 780:2015? Other Medical Device Related Standards 4
L Turkish Requirements - Does the Software need to be translated? CE Marking (Conformité Européene) / CB Scheme 2
R Where does IATF 16949 address Process mapping? IATF 16949 - Automotive Quality Systems Standard 3
J Does Pakistan Medical Device Import License allows parallel import? Other Medical Device Regulations World-Wide 0
BeaBea Interesting Discussion Where Does Marketing/ Advertisement of Products fit in to ISO 9001? Process Maps, Process Mapping and Turtle Diagrams 39
P Does anyone have a API Q1 Documentation Package? Quality Management System (QMS) Manuals 1
N What is our product classification? (Does Unclassified classification still exists) Other US Medical Device Regulations 14
C Does a CE mark infer meeting all applicable standards? CE Marking (Conformité Européene) / CB Scheme 4
N Small Company - Internal audit process - Who does the audit? Internal Auditing 16
J Does anyone have an excel IATF 16949 Internal Audit checklist I could use? IATF 16949 - Automotive Quality Systems Standard 7
A Does AS9100 require traceability to operators performing the work? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
I Does anybody use Detection in medical device Design FMEA? ISO 14971 - Medical Device Risk Management 18
A EN ISO 14971:2019 does not include the Annex Zs ISO 14971 - Medical Device Risk Management 4
Watchcat Does "Similar Device" = "Predicate"? EU Medical Device Regulations 7
C Document Control Stamps - Does anyone still stamp their documents? Document Control Systems, Procedures, Forms and Templates 24
Y Does Solidworks (2D/3D drafting modules) need validation? Other Medical Device and Orthopedic Related Topics 5
M Definition Open Audit - What does an Open Audit mean? Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
I Does training have to be written? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
P What does UPH/m² mean? Quality Tools, Improvement and Analysis 3
L "IATF-Compliant" IATF 16949:2016 certification? What does this mean? IATF 16949 - Automotive Quality Systems Standard 13
D Why does Official Journal list superseded standards? EU Medical Device Regulations 6
B Record Management - Does the QMS need to control templates of records? Records and Data - Quality, Legal and Other Evidence 17
MDD_QNA QR Code Standard ISO/IEC 15417:2007 - Does anyone use it? Other Medical Device Related Standards 3
R Does any here use an internal auditing tool that works on different platforms? Internal Auditing 3
Sidney Vianna IAQG SCMH explains "positive risk"..........but does it? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
D Where does "as far as possible" stop? FMEA - EN 14971 ISO 14971 - Medical Device Risk Management 29
W Does anyone have an API Q2 checklist for internal auditing? Oil and Gas Industry Standards and Regulations 1
G Does pitch/increment/resolution of a ruled scale apply to measurement uncertainty as line item? Measurement Uncertainty (MU) 10
L External power supplies: How close does the safety report have to match the end-use application? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
Watchcat Does 820.30 include the manufacturing process? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
Watchcat Does ISO 13485 7.3 include the manufacturing process? ISO 13485:2016 - Medical Device Quality Management Systems 14
R When does the FDA consider a component a medical device? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 17
B Does EN ISO 15223-1:2016 include the graphic symbols to be added to software and IFU? Other Medical Device Related Standards 9

Similar threads

Top Bottom