How does Non-Linear UUT affect TAR (Test Accuracy Ratio) Calculation?

E

Eddie Faulkner

#1
Hi,

How does Non-Linear UUT affect TAR calculation?

For example, I have an D/P Air Flow transmitter that performs the square root extraction from measuring D/P and outputs a flow CFM reading.

Parameters are:
INPUT: 0 to 1.3622 "WC
OUTPUT: 0 to 1760 CFM
Tolerance 17.6 CFM (Basically, 1% of Range)

What minimum test standard is needed in order to simply test the instrument with at least as good of standard as the UUT?

In this specific example, how does the square root calculation affect the TAR calculation?

Thanks,

Eddie
 
Elsmar Forum Sponsor

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#3
If using TAR instead if TUR, then you will first need to define the UUT over its range. That is, an initial curve based on manufacturer's specs. Then the calibration standards can be charted using specific data - with uncertainties - from the calibration of the standards.

Once you have that, and for simplicity that may need to be done by range breakdown, then the information can be put into Excel. That in turn will give a visual. Specific calculations for the points selected can easily be completed.

Now you are ready for actual calibration of the UUT, and then replot the points for the UUT. You can specify that the calibration provider give you the data for the points rather than a pass/fail.

You will also need to establish minimum acceptable TAR to know if you met requirements.

Hope this helps.
 

dgriffith

Quite Involved in Discussions
#4
Hi,

How does Non-Linear UUT affect TAR calculation?

For example, I have an D/P Air Flow transmitter that performs the square root extraction from measuring D/P and outputs a flow CFM reading.

Parameters are:
INPUT: 0 to 1.3622 "WC
OUTPUT: 0 to 1760 CFM
Tolerance 17.6 CFM (Basically, 1% of Range)

What minimum test standard is needed in order to simply test the instrument with at least as good of standard as the UUT?

In this specific example, how does the square root calculation affect the TAR calculation?

Thanks,

Eddie
Been thinking about this since it was posted. You can use a rotameter transfer standard calibrated to 0.5%, since most cannot be read with the human eye to better than 1%. Why waste money on a better calibration.
If the output units are correct, then the measured input differential pressure is likely correct. Since the result is in CFM, wouldn't you calibrate in CFM? Who cares about linearity? :mg:
Okay, we all do, but I don't see it as important here. It's already in the CFM result. When you compare CFM to CFM, you get CFM error. The rotameter is non-linear internally (volume vs. piston size), but the scale you read the piston against is linear.
Your TAR is the 0.5% vs. the 17.6 of range. But that's not the best way to do it, as TAR is fast becoming an obsolete parameter of quality. The TAR would be the 1% UUT tolerance / the rotameter 0.5% uncertainty, for a 2:1 TAR.
You certainly could find a better instrument at more cost, or you could have the cal done at a vendor that uses a primary standard.
:2cents:
 
Last edited:
Thread starter Similar threads Forum Replies Date
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
L How to evaluate the process capability of a data set that is non-normal (cannot be transformed and does not fit any known distribution)? Capability, Accuracy and Stability - Processes, Machines, etc. 12
N Does 13485 auditor have a right to inspect non-CE products? ISO 13485:2016 - Medical Device Quality Management Systems 5
R RoHS for Medical Devices - Does the directive apply to non-EEE accessories? RoHS, REACH, ELV, IMDS and Restricted Substances 1
M AS9100 First Article - Does 7.5.1.1 revC or 8.2.4.2 revB apply to Non Aerospace Goods AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 19
K How does an Internal Auditor document non-ethical or illegal practices Internal Auditing 10
B Does someone use APQP for non automotive industrie? IATF 16949 - Automotive Quality Systems Standard 2
T Non-Conformance Reports - Does the standard specifically say what an NCR is called? Nonconformance and Corrective Action 28
J Shipping from non-registered sites, how does it affect my registration? Registrars and Notified Bodies 7
B Non-sterile - Does anyone know of a universal symbol? ISO 13485:2016 - Medical Device Quality Management Systems 5
T Does anyone have any ideas/samples of non-conformance reports? Document Control Systems, Procedures, Forms and Templates 4
lanley liao Does all of the suppliers need to integrated into the supplier list qualified of the company? Oil and Gas Industry Standards and Regulations 2
0 To which part of 13485 does this refer? ISO 13485:2016 - Medical Device Quality Management Systems 3
A Medical Device Contract Manufacturer - Does the CM need to register with FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
J Records Control - Does each individual record need to be numbered? Records and Data - Quality, Legal and Other Evidence 2
lanley liao Does the customer`s trademark belong to customer-supplied property? Oil and Gas Industry Standards and Regulations 2
H How does a gas turbine work on diesel fuel? Oil and Gas Industry Standards and Regulations 12
G What does performance specification include? US Food and Drug Administration (FDA) 1
W Where does a coatings and paint company fall in IATF? IATF 16949 - Automotive Quality Systems Standard 5
A How much does a complete biocompatibility test package cost? Other ISO and International Standards and European Regulations 1
B Does anybody know how to get older versions of Minitab to work in Windows 10? Quality Tools, Improvement and Analysis 9
M Does the ISO 9001:2015 standard require a disaster recovery plan or emergency response plan ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
C Does an accessory need an IFU if it use is discussed in the Parent device IFU? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
S How long does it take to register a product with MHRA? UK Medical Device Regulations 3
M Quality Manual - Where does Revision History Section go? Document Control Systems, Procedures, Forms and Templates 8
U Does *anyone* know a lab that will test to EN 455-4 Medical Gloves shelf life determination? EU Medical Device Regulations 1
A Brexit When does the UK responsible person need to be in place? UK Medical Device Regulations 10
N Does anyone have experience of GB/T 34986-2017? China Medical Device Regulations 1
Z Does anyone have experience with EN ISO 17664 ? IEC 62366 - Medical Device Usability Engineering 9
F Does anyone have an ESD quality/cooler talk to share? Training - Internal, External, Online and Distance Learning 4
A What does this line from MDCG 2020-3 (MDR art. 120 substantial change) mean to you? EU Medical Device Regulations 4
D Change Approval Requirements - Does every change need formal customer approval? Design and Development of Products and Processes 17
T What does AS9100 mean when it says you must establish a process to do X? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 24
L Does a backdate form format can be changed if wrong revision is used? Document Control Systems, Procedures, Forms and Templates 8
B General Motors and Honda Alliance - What does this mean to suppliers? IATF 16949 - Automotive Quality Systems Standard 3
C ISO 13485 :2016 - CAPA - Does every CAPA need to be checked by regulations? ISO 13485:2016 - Medical Device Quality Management Systems 9
A Does ISO 9001:2015 cover all the requirements of ISO 10012:2003? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
N FDA UDI - Label vs. Labeling - Does the insert need to include UDI? Other US Medical Device Regulations 1
A Does anyone have a checklist of API Spec 650 13th Edition? Oil and Gas Industry Standards and Regulations 0
D Does Manufacture can submit CE mark application under MDD with NB for his New product after May 2020? EU Medical Device Regulations 3
A What does this sentence "this symbol shall be used in the orientation shown" mean in ISO 780:2015? Other Medical Device Related Standards 4
L Turkish Requirements - Does the Software need to be translated? CE Marking (Conformité Européene) / CB Scheme 2
R Where does IATF 16949 address Process mapping? IATF 16949 - Automotive Quality Systems Standard 3
J Does Pakistan Medical Device Import License allows parallel import? Other Medical Device Regulations World-Wide 0
BeaBea Interesting Discussion Where Does Marketing/ Advertisement of Products fit in to ISO 9001? Process Maps, Process Mapping and Turtle Diagrams 39
P Does anyone have a API Q1 Documentation Package? Quality Management System (QMS) Manuals 1
N What is our product classification? (Does Unclassified classification still exists) Other US Medical Device Regulations 14
C Does a CE mark infer meeting all applicable standards? CE Marking (Conformité Européene) / CB Scheme 4
N Small Company - Internal audit process - Who does the audit? Internal Auditing 16
J Does anyone have an excel IATF 16949 Internal Audit checklist I could use? IATF 16949 - Automotive Quality Systems Standard 7

Similar threads

Top Bottom