How Many Notified Bodies are "Enough" Already?

Watchcat

Trusted Information Resource
#1
The days to MDR are numbered, and that number grows smaller every day, a pace that is alarming for many in and around the industry. Everyone is asking how many, how many, how many notified bodies will have been designated under the new regulations this month, next month, by the end of April? End of April??? Yikes!

I’d like to consider a different question: How many notified bodies are “enough”? And for whom?

In theory, only one NB is needed...

[Full article attached]
 

Attachments

Elsmar Forum Sponsor

Ed Panek

QA RA Small Med Dev Company
Leader
Super Moderator
#2
I myself have reached out to all of them last week. One of them responded and said no new clients in 2020. The rest have not responded.
 

Ronen E

Problem Solver
Moderator
#3
Everyone is asking how many, how many, how many notified bodies will have been designated under the new regulations this month, next month, by the end of April? End of April???
Not really; or at least they shouldn't, because it's not relevant anymore for the question "What's going to happen on May 27, 2020?"

It takes on average 6 months from a successful certification application (manufacturer to NB) to issuance of a certificate, so any designation in 2020 is (on average) irrelevant for certification by May 26. Add to that that a NB doesn't start "pumping" certification on the day it's designated, and you realise that the "(on average)" can practically be dropped from my previous statement.

I have to credit Erik Vollebregt for this observation/analysis.
 

Watchcat

Trusted Information Resource
#4
I myself have reached out to all of them last week. One of them responded and said no new clients in 2020. The rest have not responded.
Yeah, I think pretty much everyone just has to hang onto their MDD NB and hope it gets MDR certified.

A strategy I might try is to keep a close watch on NANDO, and the minute the next certifications are announced, call those NBs. (I'm assuming you reached out only to the 8 that are already MDR certified and still located in the EU?) The newly designated will probably start by checking to see how many of their "existing" clients are still with them (some companies will withdraw from the EU) and how many of their devices still need to be certified by May 26 (and some companies with withdraw some devices), so there might be a small window of opportunity there.
 

Watchcat

Trusted Information Resource
#5
I have to credit Erik Vollebregt for this observation/analysis.
Actually, Erik's analysis isn't solid, but then he's a lawyer, not a statistician. My comments are on his blog, but the cliff notes are that MedTech Europe reported that some of its members said that MDD certificates had been taking 3-9 months before the crunch started, sometime in 2017. Erik managed to mangle this into an "average" of 6 months and then added "at least," but of course, if MedTech Europe's data were solid, 3-9 months meant they were taking at least 3 months, not at least 6.

One problem is that there is no universally accepted starting point for EU certification timelines. And no transparency, so it's all just rumor. I recently asked someone at an NB who was responsible for certification management about the MedTech Europe data, and they saw it the same way: "What do manufacturers think when they say "time for conformity assessment"? Is it time from offer approval till receiving the certificate? Or maybe the time from 1st day of the audit till report." And the different NBs have handled the certification process differently, so different companies don't all have the same starting points to choose from.

My own take is that this is totally uncharted territory, and how long things used to take are not likely to be predictive of how long they will take going forward. I am expecting at least some of the NBs to have a few tricks up their sleeves.
 

Ed Panek

QA RA Small Med Dev Company
Leader
Super Moderator
#6
Actually, Erik's analysis isn't solid, but then he's a lawyer, not a statistician. My comments are on his blog, but the cliff notes are that MedTech Europe reported that some of its members said that MDD certificates had been taking 3-9 months before the crunch started, sometime in 2017. Erik managed to mangle this into an "average" of 6 months and then added "at least," but of course, if MedTech Europe's data were solid, 3-9 months meant they were taking at least 3 months, not at least 6.

One problem is that there is no universally accepted starting point for EU certification timelines. And no transparency, so it's all just rumor. I recently asked someone at an NB who was responsible for certification management about the MedTech Europe data, and they saw it the same way: "What do manufacturers think when they say "time for conformity assessment"? Is it time from offer approval till receiving the certificate? Or maybe the time from 1st day of the audit till report." And the different NBs have handled the certification process differently, so different companies don't all have the same starting points to choose from.

My own take is that this is totally uncharted territory, and how long things used to take are not likely to be predictive of how long they will take going forward. I am expecting at least some of the NBs to have a few tricks up their sleeves.
In addition, with employees shifting companies and high workloads in general, its probable reviewers will be moved around to service higher status customers in time. This could lead to extra delays. As Watchcat mentioned, this is a watershed moment for the industry and the stability (foundation) will take time to settle.
 

Watchcat

Trusted Information Resource
#7
to service higher status customers in time.
I was initially astonished to hear some of the claims about how long it takes to issue a certificate. Normally, you would expect the NBs to pitch their speed. The head of one MDR-designated NB claimed that it took a year, and emphasized that these are all "original certs." They are original certs on paper, but they are not MDD original certs, for a device, and sometimes a company, that is new to the NB. The NBs are servicing existing clients with existing devices, so they are already familiar with the companies and the devices.

But then I realized that, due to the lack of transparency, this positions the NBs to leave some customers waiting, without those customers realizing/complaining that higher status customers got their certs issued in a good bit less time than a year. Or maybe even 3 months.

I'm about to let the MDR go and concentrate on some FDA stuff, which is where my Watchcat interests lie, but, depending on my time, I might post something quasi-tongue-in-cheek in mid-May, maybe on the order of "Top ten ways to get your MDR certificate in 24 hours or less" or "We don't need no stinkin' certificates!" ;)
 

shimonv

Trusted Information Resource
#8
A colleague of mine is a BSI auditor. He informed me that they are not accepting new clients this year. He also wrote that it's the same with other NBs except that they don't say it.

It's going to be a difficult year for CE hopefulls.

Shimon
 

Watchcat

Trusted Information Resource
#9
They may not say it, but the word has gotten around, plus it's fairly straightforward math. BSI UK's first post after it received its designation (and still planned on using it) said it was prioritizing existing clients. This makes so much business sense, you would expect the other NBs to do the same. And you don't need a lot of transparency to guesstimate how long that is likely to take.
 

Sidney Vianna

Post Responsibly
Leader
Admin
#10
Fascinating and opportunity-rich for a thesis in Business (Mis)Management. The apparent massive asymmetrical positioning of supply and demand in the NB sector should make some people in the conformity assessment world think. But that is a challenge for many of them.

Isn't this a result of years of mismanagement in the sector? Commoditization of conformity assessment services led to buyers price shopping of NB's, what led to substandard assessments, forcing the Authorities to significantly tighten the control over NB's and, along the way, reduce the number of providers. The commoditization approach also meant that NB's cannot attract and retain the talented professionals they need in order to deliver the services in a timely manner.

Pressed by the Regulators on one side and the Market needs on the other, NB's, mostly risk-averse, are missing a huge business opportunity and the Market they serve, left in a conundrum.

Fascinating.
 
Thread starter Similar threads Forum Replies Date
N MDR review process by notified body - How many steps exist in the review process EU Medical Device Regulations 0
R How many process maps are adequate for a company for IATF 16949? Process Maps, Process Mapping and Turtle Diagrams 17
V How many hrs does it take to create a Technical File (TD)? EU Medical Device Regulations 4
G How many Design FMEAs do you have? FMEA and Control Plans 4
V Literature search over how many years? EU Medical Device Regulations 8
S How many tester quantity we need on the line based on the cycle time and peak volume Manufacturing and Related Processes 3
M R&R Studies - how many? per product? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
J How many majors is one company allowed ? IATF16949 Manufacturing and Related Processes 16
M Summative evaluation: how many sample per participant? IEC 62366 - Medical Device Usability Engineering 5
J How many hours for a NB to review Technical Documentation? EU Medical Device Regulations 6
T Too many NC's/ Deviations and CAPAs galore Nonconformance and Corrective Action 10
B How many nonconformances have you received on average during IATF audits? General Auditing Discussions 27
G Too many customer complaints Customer Complaints 16
I How many gauges in a Go/No Go Gauge? General Measurement Device and Calibration Topics 7
BeaBea How Many Processes should be created for each Department? Process Maps, Process Mapping and Turtle Diagrams 5
S How many mandays required for the below mentioned standards IEC 27001 - Information Security Management Systems (ISMS) 3
F How many signatures do we need on calibration certificates? ISO 17025 related Discussions 8
M Informational From RAPS – Operating System Vulnerabilities in Many Medical Devices, Germany’s BfArM and Device Firms Warn Medical Device and FDA Regulations and Standards News 0
R Validation of Medical Device Hardware containing Software - How many to Validate ISO 13485:2016 - Medical Device Quality Management Systems 1
K How Many Samples to send for Chemistry Testing Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
N MDSAP Participants 2019 - How many companies are participating YTD? ISO 13485:2016 - Medical Device Quality Management Systems 20
S How many of you are using Robotic process automation for calibration lab management? ISO 17025 related Discussions 0
G Measuring customer parts on a CMM - How many decimals to report to the customer? ISO 17025 related Discussions 28
J Supplier choice limitation - Many of them dont have PSCR, VDA or even ISO 9001 VDA Standards - Germany's Automotive Standards 0
O Clean room performance qualification (pq) - How many times can I do each test? Manufacturing and Related Processes 8
G How many samples you require to test during PQ for validation of a equipment? ISO 13485:2016 - Medical Device Quality Management Systems 1
D IATF 16949: Do we have too many processes? Are they all the same? Registrars and Notified Bodies 7
D QMS Performance Based on how many NCRs were generated month to month Nonconformance and Corrective Action 8
S How many QMS processes and procedures? ISO 13485:2016 - Medical Device Quality Management Systems 2
L How many of you are Certified to IATF 16949 IATF 16949 - Automotive Quality Systems Standard 12
G How many Uncertainty Measurements for Micrometers and Calipers? Measurement Uncertainty (MU) 3
R How many Corrective Action Request can be Issued by DCMA for a single issue? Nonconformance and Corrective Action 2
A Training Program Help - Old docs, new docs, so many docs... ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
M How many Pieces to Test - Sterility Testing USP EU Medical Device Regulations 2
P Are we an outlier for so many variations of Change Management? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
M SPC with many data points from one dimension Statistical Analysis Tools, Techniques and SPC 16
S How many times do you have the term 'Requirements' in your Quality Manual? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
L How many working hours to required to implement ISO 14001 ? ISO 14001:2015 Specific Discussions 6
B ISO/TS16949 Internal Auditing - How many auditors? Internal Auditing 4
R Training Production Employees When We Have SOOO Many SOPs Training - Internal, External, Online and Distance Learning 9
P Scope Extension Audit - How many days? Do I audit all AS9100 requirements? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
J How can I manage Occurence Rates in FMEA efficiently when I have many processes? FMEA and Control Plans 9
V How many times the quality agreement has been invoked? ISO 13485:2016 - Medical Device Quality Management Systems 3
C How to document many document control systems with one Document Control Procedure? Document Control Systems, Procedures, Forms and Templates 6
A How many laboratories are accredited to ISO 17025 in the USA? ISO 17025 related Discussions 5
R Is it possible to randomly select samples from 1 column to many column Using Minitab Software 2
L How many ISO Standards are there and time allowed for Closing NCs? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
S How many mandays are required for QMS audit? Internal Auditing 5
Ajit Basrur FDA - "We Moved Forward on Many Fronts This Year" - Margaret A. Hamburg, M.D. US Food and Drug Administration (FDA) 1
F AS9100 - How many times can the same auditor audit your system AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1

Similar threads

Top Bottom