How to best prioritize Medical Device Label Contents

Mark Meer

Trusted Information Resource
#1
Hi Everyone!

A little bit of a dilemma here: we have a small, home-use, wireless medical device and are in the process of developing the device labels.

We'd like to have a single device label that complies with US, EEA and Canadian regs, and additionally get certified to IEC 60601-1, 60601-1-2 and 60601-1-11.

This makes for a lot of required info! Too much to fit to be sure (our allocated label-space is about the size of your thumb)...

So, I'm wondering if anyone has any input on what is "absolutely essential" and what can be relocated (e.g. to IFU)...

Here's the list of label requirements so far (please let me know if I'm misinterpreting regulation/standards, or if I've missed anything):

The ones I'm pretty sure are required:
  • General
    • IP Rating (required by IEC 60601-1-11 clause 7.2)
    • Consult Instructions symbol (recommended by IEC 60601-1 clause 7.2.3)
    • Applied Part symbol (required by IEC 60601-1 clause 7.2.10)
  • Identification
    • Name and address of manufacturer (required by MDD clause 13.3a)
    • Unique device identifier (required by FDA 21 CFR Part 801)
    • Year of manufacture (required by MDD clause 13.3l)
    • Lot Number
    • Serial Number
    • Model Number
    • GTIN (required because UDI must also include a human-readable format)
    • Label ID & Revision
  • Because it has a radio:
    • Non-Ionizing Radiation symbol (required by IEC 60601-1-2 clause 5.1.2)
    • FCC ID (required by US FCC)
    • IC ID (required by Industry Canada)

Shortlist: Ones that either would be nice to include, or I'm not sure about...
  • "Rx Only" (it is a prescription device for US....can't find and explicit regulation requiring this though...)
  • NRTL Mark
  • CE Mark
  • WEEE Symbol (WEEE Directive 2012-19-EU)

Anyway, obviously all of the above are next to impossible to cram into our label space (we can rule out laser-etching or other fancy stuff...we don't have the means).

So, seeking any advice on what (if anything) can be omitted from the device label (and moved to IFU and/or packaging).

Thanks in advance!
 
Elsmar Forum Sponsor

Marcelo

Inactive Registered Visitor
#3
Re: How to best prioritize device label contents

Anyway, ISO TC 210 has just decided to create a general labeling standard, maybe a problem like yours should be part of the document too as I see things like this questioned from time to time.
 

Mark Meer

Trusted Information Resource
#4
What risk control measure will you use this to refer to?
I'm not quite sure what you mean. There are many warnings, precautions and contraindications that obviously cannot be stated on the label itself. So it seems prudent to indicate that the operator should first consult with the IFU.

To be honest, I was hoping that maybe using the Mandatory Action Symbol, ISO 7010-M002, might be a way to avoid having to cram the label. :notme:

ISO TC 210 has just decided to create a general labeling standard
That's great...provided it aligns with all the aforementioned regulations and standards. Unless it has some exceptions that supersede the requirements I've listed, however, I don't see it being much help in terms of "label-overload" problem I'm describing.
 

Mark Meer

Trusted Information Resource
#5
Warning: Rant Time :mad:

I find this kind of label-cramming problem quite frustrating.

Like I say, we are dealing with label-space about the size of a thumb.

The GTIN is 14 characters, FCC ID up to 15 characters, IC ID up to 11 characters, date is 10 characters, lot is 8 characters, model is 5 characters, IP rating is 4 characters... That's 7 lines of text right there!

Now add to that all the required symbols: Applied Part, Non-Ionizing Radiation, UDI...

And THEN manufacturer information?!

For small labels, adhering to all these requirements quickly becomes ludicrous! :bonk:

And for what? In practice few people know what symbols mean without consulting the descriptions in the IFU. And seriously...does the presence of, say, a "BF Applied Part" symbol really affect how people use the device?
Also, the requirement that UDI also has human-readable interpretation (HRI) seems of little value...
 

Marcelo

Inactive Registered Visitor
#6
I'm not quite sure what you mean. There are many warnings, precautions and contraindications that obviously cannot be stated on the label itself. So it seems prudent to indicate that the operator should first consult with the IFU.

To be honest, I was hoping that maybe using the Mandatory Action Symbol, ISO 7010-M002, might be a way to avoid having to cram the label.
This symbol is to be used when there?s a link to a specific risk control measure for a specific risk, as defined in the rationale:

Subclause 7.2.3 ? Consult ACCOMPANYING DOCUMENTS
It is not intended in every case when the instructions for use contain warnings, that the ME EQUIPMENT be marked with ISO 7010-M002 (see Table D.2, safety sign 10). Too many warnings and unnecessary warnings are counterproductive. Only when the MANUFACTURER, as a RISK CONTROL measure for a specific RISK, decides to mark the ME EQUIPMENT to instruct the OPERATOR to read the instructions for use, should safety sign ISO 7010-M002 be used.
In previsous editions, this was not very clear so people understand that every time you had to put this symbol, which is stupid ("read the whole manual" is not a good reminder of risks). This also could cause even more risks.
 

Marcelo

Inactive Registered Visitor
#7
And for what? In practice few people know what symbols mean without consulting the descriptions in the IFU. And seriously...does the presence of, say, a "BF Applied Part" symbol really affect how people use the device?
It?s more related to liability than anything else.
 

pkost

Trusted Information Resource
#8
I'm not an expert on the US requirements (or electrical in general), however in the EU labelling the device does not always have to mean direct part marking; it can also be applied to the packaging.

In the past I have had to make similar decisions and did it on the basis of risk - what information is critical for safety (primarily), post market surveillance (secondary)

I ended up with all labelling on the primary packgaing and the LOT number on the device itself
 

Marcelo

Inactive Registered Visitor
#10
Is there more information available on this? Who from US FDA is participating?
No, this was decided in the last plenary and initial work is being discussed.

No one is participating yet, but it was proposed by WG 2 - which is chaired by Carol Hermann from AAMI, with input from Dave Osbourne from Philips.
 
Thread starter Similar threads Forum Replies Date
earl62 What is the best way to control special characteristics in Control plan? Is it Mandatory to have SPC for IATF 16949? IATF 16949 - Automotive Quality Systems Standard 5
J ISO 13485 System 'soft start' - How to best reflect this in initial audits, management review minutes and other records? ISO 13485:2016 - Medical Device Quality Management Systems 3
S What is the best method to bond urethane foam to urethane foam? Manufacturing and Related Processes 4
Watchcat Best Regulatory Oversight for Off-Shore Device Manufacturing? Other Medical Device Regulations World-Wide 1
E Best way to keep up with pharmaceutical quality knowledge and current affairs Service Industry Specific Topics 1
K Best Measurement Systems Demos in California? General Measurement Device and Calibration Topics 1
qualprod Best approach to get a real value as average? Statistical Analysis Tools, Techniques and SPC 6
CPhelan Best metrics for monitoring low volume (<1000) parts per vendors Supply Chain Security Management Systems 1
DuncanGibbons Best practice for identifying "items" of parts for DFMEA analysis AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
I What kind of wine best complements the Friday that you close out your external audit findings? Opinions are welcome. Coffee Break and Water Cooler Discussions 12
I Non-Conformance vs OFI -- your best descriptions ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 61
A What are Practical data center best practices IEC 27001 - Information Security Management Systems (ISMS) 0
G Best Practices for IT auditing - Is a session-id necessary for a complete audit trail? IEC 27001 - Information Security Management Systems (ISMS) 0
S Best software for customer support/complaints? Customer Complaints 0
R Bill of Materials for a complex product - Industry best practice ISO 13485:2016 - Medical Device Quality Management Systems 2
G Oversized Threads - What sized stab pin would be best to quickly check for oversized thread holes General Measurement Device and Calibration Topics 5
D What is the best software used for the pharma compliance management? Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
M Informational MHRA launches consultation on how to best engage patients and the public Medical Device and FDA Regulations and Standards News 0
optomist1 Mobility Field Manufacturing Best - Worst Practices Manufacturing and Related Processes 5
qualprod Best criteria to measure Corrective Action effectiveness - Poor Maintenance ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
N Best practices for capturing audit objective evidence in a practical manner? Internal Auditing 3
S How to Learn all aspects of ISO 27001:2013 | The best way to grab the knowledge on 27001:2013 (Step by Step) IEC 27001 - Information Security Management Systems (ISMS) 7
I Best visual description for SOPs? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 18
Z Best Practices - small volume medical device tube cutting (0.020" to 0.080" OD)? Manufacturing and Related Processes 3
M Over-labelling - Any requirements or best-practices? Other Medical Device Regulations World-Wide 0
M Informational DITTA White Paper on Cybersecurity: Best Practices in the Medical Technology Manufacturing Environment Medical Device and FDA Regulations and Standards News 0
M Informational BSI – MDR Documentation Submissions Best Practices Guidelines Medical Device and FDA Regulations and Standards News 0
K Which clause would best fit this Nonconformance? (Supplier Related) Internal Auditing 2
P How to have employees buy-in of quality procedures. Best practices? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
S Expiry date vs. Best Before Date - Reagents ISO 13485:2016 - Medical Device Quality Management Systems 9
C Will anyone please share training material for ISO:13485:2016 for best practices Training - Internal, External, Online and Distance Learning 0
E Best GD&T training - Your suggestions and recommendations please Training - Internal, External, Online and Distance Learning 3
M Best Practice for setting tolerances on a Drawing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
M PCBA Drawing Best Practices IPC Class 3 ISO 13485:2016 - Medical Device Quality Management Systems 1
Ron Rompen Best Practices in CMM Correlation Studies General Measurement Device and Calibration Topics 3
B Best place is to purchase IEC 60417 labels that will pass markings test IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P Best approach to tackle difficult certifications Career and Occupation Discussions 2
E Best Ways to Close a Project APQP and PPAP 3
S Opinions on the "Best" CMM for High Accuracy Machining General Measurement Device and Calibration Topics 9
C What is the best ISO 22000 template pack to purchase? Food Safety - ISO 22000, HACCP (21 CFR 120) 2
M Best Type of Grease for Vex Plastic Gears Manufacturing and Related Processes 5
W Which is best IAS academy in Chennai Career and Occupation Discussions 1
S Referencing procedures on Control Plans - Best Practices FMEA and Control Plans 3
somashekar Which Medical Device Regulatory Body you rate as the best Other Medical Device and Orthopedic Related Topics 6
P Best Method of Cleaning of Thread Ring Gages Calibration Frequency (Interval) 5
A Best Method of Auditing an Installation Process at a client job-site ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
J How to manufacture best Roll Forming machine? Manufacturing and Related Processes 1