How to Calibrate and Verify Torque Multiplier Gages
charlie79
Does anyone know about checking torque multipliers...???
There are only input and output....
We give input torque 10 and output must be torque 100...if the ratio 1 : 10...
So...what should i do....calibration or measure the actual ratio??
H
Harley11
That did not answer the question. We use a torque cell on the input and another on the output and determine an average power ratio for the multiplier.
charlie79
Thanks Harley...
I got your point
Then....
how to determine the uncertainty ?
do you have any reference for this calibration...
H
Harley11
Uncertainty budgeting is beyond my pay grade. I am sure you must account for both torque cells as standards used. Someone else here could assist you in uncertainty calculations.
As far as a reference, I do not have one. I am just telling you how we do it here by our procedure. Everything is traceable and it works well for us. It was done this way before I was hired here.
D How to Calibrate and Verify a Heating Oven General Measurement Device and Calibration Topics 3
W How to verify/calibrate a Neptune Water Meter General Measurement Device and Calibration Topics 4
S How to Calibrate or Verify Screw Pitch Gages General Measurement Device and Calibration Topics 16
J Work Instruction to Calibrate or Verify a Combination Square General Measurement Device and Calibration Topics 15
A Looking for a third party to calibrate our measure equipment - South Carolina General Measurement Device and Calibration Topics 6
I Can I Calibrate On Demand? General Measurement Device and Calibration Topics 9
S Calibrating our own equipment, can we? Micrometers to calibrate vernier calipers General Measurement Device and Calibration Topics 13
F Crimper Calibration - How do you Control & Calibrate your crimpers? Calibration Frequency (Interval) 13
P How to Calibrate a HP 3458A Digital Multimeter with Fluke 5080A calibrator General Measurement Device and Calibration Topics 1
B Choosing not to calibrate (IATF 16949) IATF 16949 - Automotive Quality Systems Standard 6
K New Kilogram - The process we use to calibrate the standards General Measurement Device and Calibration Topics 6
D How to Calibrate any Digital Sphygmomanometer under use for 2 years in a Hospital Forum News and General Information 1
L Re-calibrate after software update? Calibration Frequency (Interval) 3
J To Calibrate or not to calibrate electronics rack components Manufacturing and Related Processes 5
T How to Calibrate Taper Gages General Measurement Device and Calibration Topics 1
N Calibrate temperature on refrigerated Centrifuges? General Measurement Device and Calibration Topics 2
B Can we calibrate a standard pin with a gauge? General Measurement Device and Calibration Topics 2
D Lesser expensive to calibrate various test weights General Measurement Device and Calibration Topics 5
A How to Calibrate Laser Beam Tester & Radiation Monitor General Measurement Device and Calibration Topics 3
P How do you calibrate a VNA (Vector Network Analyzer) cal kit? General Measurement Device and Calibration Topics 4
V How to Calibrate Ring Force Gauges Service Industry Specific Topics 1
bobdoering Who says you can't calibrate a steel rule! Funny Stuff - Jokes and Humour 14
S How can I Calibrate a Biosafety Cabinet? General Measurement Device and Calibration Topics 1
C How to Calibrate a Viscosity Tester General Measurement Device and Calibration Topics 3
0 Is it possible to use a threaded ring gage to calibrate the M8 gage? Calibration and Metrology Software and Hardware 4
C How to Calibrate a Medical Infrared Thermometer General Measurement Device and Calibration Topics 3
T How do I calibrate NPT Thread Plug Gages? General Measurement Device and Calibration Topics 9
E Do you need a Certification to Calibrate Gauges? General Measurement Device and Calibration Topics 2
B Calibrate Hole Presence Detection Device? General Measurement Device and Calibration Topics 5
J What do you use to Calibrate 40" Calipers General Measurement Device and Calibration Topics 10
A Guideline to Calibrate Chamber Temp. & Pressure of a Medical Steam Sterilizer General Measurement Device and Calibration Topics 8
S How to calibrate AOI (Automated Optical Inspection) for IOPQ Qualification and Validation (including 21 CFR Part 11) 3
S Laboratory to Calibrate Objective Micrometer needed General Measurement Device and Calibration Topics 4
C Tolerance Estimation of a Load Cell I want to Calibrate General Measurement Device and Calibration Topics 9
D Procedure on How to Calibrate a External Micrometer General Measurement Device and Calibration Topics 1
B ISO/TS 16949 Section 7.6.3 - Use of Manufacturer to Calibrate IATF 16949 - Automotive Quality Systems Standard 21
N Laboratory that is able to Calibrate Dew Point Meter up to 60?Cdp - Needed General Measurement Device and Calibration Topics 1
C How to calibrate or control patterns for fiberglass General Measurement Device and Calibration Topics 4
N Scale with Internal Calibration - Do I need to calibrate? How to document this? Calibration Frequency (Interval) 8
I Need help finding an Accredited Lab that can Calibrate Brunson 76-RH Survey Transits General Measurement Device and Calibration Topics 5
M Controlling Temperature and Time - Do I need to calibrate the thermostat? IATF 16949 - Automotive Quality Systems Standard 4
D How much Tolerance do I use to Calibrate a 150kgs capacity Scale? General Measurement Device and Calibration Topics 5
M To Calibrate or Not to Calibrate various Weight Scales ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 24
N Thermocouple Inhomogeneity - Need to calibrate a roll of 50m thermocouple type K General Measurement Device and Calibration Topics 3
J Calibrate before use required? Inspection Gages and Measurement Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
R Traceability for Mercury Argon Calibration Source to Calibrate a Spectrometer? General Measurement Device and Calibration Topics 5
P How to Calibrate Laser Micrometer when no Calibration Company in the Country? General Measurement Device and Calibration Topics 3
S 3 Point Calibration or Calibrate to Extreme High and Low? General Measurement Device and Calibration Topics 9
C Seeking a holder/jig to calibrate large OD micrometer calipers (up to 23") General Measurement Device and Calibration Topics 8
S Is it OK to calibrate a measurement device tighter than manufacturer specifications? General Measurement Device and Calibration Topics 18