How to calibrate dial indicators, gauges and scales
trainee
Hello Everybody!
this is my second post here, but i have been reading this forum for a while and it has been really really helpful.
I have several questions about calibrating measurement instruments.
How do you know how often you should calibrate your scales? specially if you haven't performed any traceability or r&r study...
And second, How do you calibrate a dial indicator? I'm not even sure if I should or not calibrate them.I have never dealt with this kind of gauges before...
Thank you very much!!!!
Don Palmer
Re: Dial indicators, gauges and scales calibration
Hi Trainee, Welcome to 'The Cove'.
First of all, I'm not a calibration expert. There are members here on the forum that are though. And hopefully one of the moderators will move your post to the 'Calibrations Section' where your questions will get more notice for response.
Some generalized answers to your questions. 1st, a good rule of thumb for setting calibration cycles (when do I recertify this thing) is to follow the manufactures (mfr of the scale) instructions. Additionally, you would want to keep mindful any specific requirements called out by the O.E.M. or your engineering people. If you have no previous history about reliability/certainty of the scales, then if the scale mfr did not set a schedule requirement, annual recertification would probably suffice.
As to dial indicators, we use a certified gauge block set of standards. They are locked up and used for certifying our calibrated tools only. We maintain another gauge block set for everyday shop use.
Hope this helps a little till the experts can answer your question in more detail.
Cheers,
Hello Everybody!
Welcome to the Cove. While this post does relate to ISO 9001, I moved your post to a more relevant forum frequented by metrology/calibration experts.
You should clarify what you mean by the term scale. Is this a scale used to measure length (e.g., six inch scale), or a scale used to weigh product?
BradM
Howdy! :bigwave: and Welcome!
Great thread by Don.
First, by no means am I an expert; there are experts on here, though. But, I'll give your question a shot. It is a good question, and it does come up a lot.
Here are some threads on this subject that may help you:
Weighing scales to be used at each machine on the shop floor
Calibration Frequency - Control of Monitoring and Measuring Devices - Clause 7.6
Specificity in Weigh Scale Calibration Frequencies
As far as calibration frequency, here are my suggestions:
1. You determine that the instrument shall/must be calibrated.
1. Determine any regulatory/specification requirement that you may have to comply with.
2. Determine any customer-specific requirements that have been established.
3. Say #1 applies, but 2 and 3 don't state anything specific, then you need to determine how often. Generally, the more critical the impact of the calibration failure, the shorter interval you want to keep. Calibration frequency is a trade-off, and you have to determine that. The shorter cycle the more costly. The longer the cycle, the more risk.
Also, maybe ask around with some friends in the business, vendors, suppliers, "friendly" competitors, and see what calibration frequency they have established.
Miner's question regarding scale also applies for me to the dial indicator. Don assumed (hopefully correctly) that you were referring to a depth gauge, or a caliper. Is that correct? Or, is this a temperature dial indicator, or pressure?
If you are looking at this device, and you are a bit unsure about calibrating it, it may be a wise move to find a competent calibration lab suitable for calibrating these devices. It may save you a lot of headache and trouble in the end.
However, if you would like to give some more specifics regarding the instrument, your general calibration competencies, and the like, we can probably provide some more detailed information.
Also, calibration of particular instruments have been covered extensively at the Cove. We (or if you like, experiment with the search function above) can find some threads already covering your instrument.
Look forward to seeing your re-posts!
trainee
Thank you guys!!!
I'm just reading all that information you provided me.
My dial indicator is a depth gauge, a hand-held device. The scales are weight scales, but i have no problem with them so far.
Have a great day!
crendfrey
Greetings and welcome Trainee:bigwave:
Brad is absolutely correct.
A little more info including the application of the devices will help us help you.
trainee
I'm using a dial indicator to check thickness in the samples of my product when I perform quality checks several times a day.
Do I have to do stability and R&R studies after i calibrate them?
I know we used to do for toquimeters (force gauges) stability and traceability studies once a year, and they were calibrated once a year too.
That's the way it should be? Or can I do those studies more often?
Thank you for your help.:tg:
Benjamin28
As stated in prior responses, if you have no regulatory requirements, or client requirements for calibration intervals, then it's a matter of cost vs risk in determining your cal interval. You should be cautioned, if you have no history data that establishes instrument performance you should start with a smaller calibration interval. Say you do monthly calibration for the first year, calibrations show no excessive variation, good instrument performance. You can then use these records to substantiate a change to a 3 month interval.
Gage R&R is a useful tool for determining whether the instrument you are using is suitable to the task it's performing, and what type of variation you have between users. It really isn't a required action, it's used more as a tool to evaluate your measuring systems performance.
As for your gage blocks, typically these are calibrated annually by an outside accredited laboratory.
The same will apply for your scale, and you will need master weights again calibrated by an outside lab.
Hershal
Metrologist-Auditor
The best answer is to have a lab accredited to ANS/ISO/IEC 17025 calibrate all your instruments.....then use the gage blocks and weights as check items for periodic checking.....
Scales in particular can be a challenge.....after all, uncertainty components like bouyancy, local acceleration of gravity, and barometric pressure all affect scale calibration, which is why they need to be calibrated where they are actually used.....and never moved by the users.....
Hope this helps.
Hershal
crendfrey
Greetings Trainee,
I think I am confused.
I have a few questions……
To what standard are you being held, assessed against, or complying with?
This makes a great deal of difference to the type of advice you receive.
Do you have 6 gages and 4 scales of 500 of each?
Do you do your own calibrations in house or send a few as masters out for comparisons, or sent them all out? What are the qualifications of whom ever is doing the calibrations?
Selecting an accredited lab may answer a lot of issues for you.
Why do a R & R after calibration when calibration should give you the performance as found and after correction (if necessary to correct) and the uncertainty of the results?
Do you just need some system of interim checks between calibrations (like GMP)
Do the operators question the performance of said devices? They use them every day, they know if something is not right (or should)
Are any of the devices legal for trade?
What do the manufacturers of your devices suggest for calibration?
Are you using said devices according to the stated manufacturer’s specifications?
I’m sorry, could go on……..just the way my mind works ……apologies.
D Calibration vs. Verification - Can we calibrate dial indicators in-house? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J How to Calibrate a Digimatic Dial Indicator? General Measurement Device and Calibration Topics 5
B How to Calibrate a Dial Indicator General Measurement Device and Calibration Topics 13
A Looking for a third party to calibrate our measure equipment - South Carolina General Measurement Device and Calibration Topics 6
I Can I Calibrate On Demand? General Measurement Device and Calibration Topics 9
S Calibrating our own equipment, can we? Micrometers to calibrate vernier calipers General Measurement Device and Calibration Topics 13
F Crimper Calibration - How do you Control & Calibrate your crimpers? Calibration Frequency (Interval) 13
P How to Calibrate a HP 3458A Digital Multimeter with Fluke 5080A calibrator General Measurement Device and Calibration Topics 1
B Choosing not to calibrate (IATF 16949) IATF 16949 - Automotive Quality Systems Standard 6
K New Kilogram - The process we use to calibrate the standards General Measurement Device and Calibration Topics 6
D How to Calibrate any Digital Sphygmomanometer under use for 2 years in a Hospital Forum News and General Information 1
L Re-calibrate after software update? Calibration Frequency (Interval) 3
J To Calibrate or not to calibrate electronics rack components Manufacturing and Related Processes 5
T How to Calibrate Taper Gages General Measurement Device and Calibration Topics 1
N Calibrate temperature on refrigerated Centrifuges? General Measurement Device and Calibration Topics 2
B Can we calibrate a standard pin with a gauge? General Measurement Device and Calibration Topics 2
D Lesser expensive to calibrate various test weights General Measurement Device and Calibration Topics 5
A How to Calibrate Laser Beam Tester & Radiation Monitor General Measurement Device and Calibration Topics 3
D How to Calibrate and Verify a Heating Oven General Measurement Device and Calibration Topics 3
P How do you calibrate a VNA (Vector Network Analyzer) cal kit? General Measurement Device and Calibration Topics 4
V How to Calibrate Ring Force Gauges Service Industry Specific Topics 1
bobdoering Who says you can't calibrate a steel rule! Funny Stuff - Jokes and Humour 14
S How can I Calibrate a Biosafety Cabinet? General Measurement Device and Calibration Topics 1
C How to Calibrate a Viscosity Tester General Measurement Device and Calibration Topics 3
0 Is it possible to use a threaded ring gage to calibrate the M8 gage? Calibration and Metrology Software and Hardware 4
C How to Calibrate a Medical Infrared Thermometer General Measurement Device and Calibration Topics 3
T How do I calibrate NPT Thread Plug Gages? General Measurement Device and Calibration Topics 9
E Do you need a Certification to Calibrate Gauges? General Measurement Device and Calibration Topics 2
B Calibrate Hole Presence Detection Device? General Measurement Device and Calibration Topics 5
J What do you use to Calibrate 40" Calipers General Measurement Device and Calibration Topics 10
A Guideline to Calibrate Chamber Temp. & Pressure of a Medical Steam Sterilizer General Measurement Device and Calibration Topics 8
S How to calibrate AOI (Automated Optical Inspection) for IOPQ Qualification and Validation (including 21 CFR Part 11) 3
S Laboratory to Calibrate Objective Micrometer needed General Measurement Device and Calibration Topics 4
C Tolerance Estimation of a Load Cell I want to Calibrate General Measurement Device and Calibration Topics 9
D Procedure on How to Calibrate a External Micrometer General Measurement Device and Calibration Topics 1
B ISO/TS 16949 Section 7.6.3 - Use of Manufacturer to Calibrate IATF 16949 - Automotive Quality Systems Standard 21
W How to verify/calibrate a Neptune Water Meter General Measurement Device and Calibration Topics 4
N Laboratory that is able to Calibrate Dew Point Meter up to 60?Cdp - Needed General Measurement Device and Calibration Topics 1
C How to Calibrate and Verify Torque Multiplier Gages General Measurement Device and Calibration Topics 6
C How to calibrate or control patterns for fiberglass General Measurement Device and Calibration Topics 4
N Scale with Internal Calibration - Do I need to calibrate? How to document this? Calibration Frequency (Interval) 8
I Need help finding an Accredited Lab that can Calibrate Brunson 76-RH Survey Transits General Measurement Device and Calibration Topics 5
M Controlling Temperature and Time - Do I need to calibrate the thermostat? IATF 16949 - Automotive Quality Systems Standard 4
D How much Tolerance do I use to Calibrate a 150kgs capacity Scale? General Measurement Device and Calibration Topics 5
M To Calibrate or Not to Calibrate various Weight Scales ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 24
N Thermocouple Inhomogeneity - Need to calibrate a roll of 50m thermocouple type K General Measurement Device and Calibration Topics 3
J Calibrate before use required? Inspection Gages and Measurement Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
R Traceability for Mercury Argon Calibration Source to Calibrate a Spectrometer? General Measurement Device and Calibration Topics 5
P How to Calibrate Laser Micrometer when no Calibration Company in the Country? General Measurement Device and Calibration Topics 3
S 3 Point Calibration or Calibrate to Extreme High and Low? General Measurement Device and Calibration Topics 9