How to calibrate Hardness Rockwell, Brinnell & Vickers Tester machine

J

jhoniegudel

#1
We have Hardness Rockwell, Brinnell & Vickers Tester machine.
We still confused how to calibrate this equipment.
Does anyone have Calibration procedure???? (base on ISO)

thanks

jhonie gudel
 
Elsmar Forum Sponsor
B

Benjamin28

#2
Typically we have these calibrated by an outside accreditted laboratory, who has the correct tools and expertise. We do however perform the routine operational checks (prior to use) which is perhaps what you're asking about?

Here is what we consider for operational checks:

Prior to test, verify that the test equipment is in good operating condition by performing and recording the prescribed Operational Checks. Verify that the test instruments carry a valid calibration sticker.

Operational check for the Brinell machine consists of:

4.3.1. Verify Brinell diameter measuring system
4.3.2. Verify condition of anvil
4.3.3. Verify condition of indenter
4.3.4. Verify hardness on certified test block for the applicable range to be used

Operational check for the Rockwell hardness testing machines consist of:

4.4.1. Balance depth-measuring device
4.4.2. Balance load cell
4.4.3. Verify condition of anvil
4.4.4. Verify condition of indenter
4.4.5. Verify hardness on a certified test block for the applicable range to be used


For the device calibration this is a guideline we follow to ensure the outside calibration company performs suitable calibrations.

BRINELL HARDNESS TESTING MACHINES

4.6. An NIST Traceable Load Cell or Proving Ring shall be used to verify Load applied over the range of the Testing Machine. A minimum of three readings shall be taken at each applicable load.

4.7. The Calibration Technician shall assist and supervise the Hardness Tester Calibration Technician and monitor all recorded data.

BRINELL HARDNESS OPTICAL SCANNING SYSTEM

4.8. The Brinell Hardness Optical Scanning system is verified by the Calibration Technician using calibrated WCG’s. These are calibrated using a stage micrometer with metric divisions of 0.01mm. Choose appropriate reference points in order to insure that the 0.01 mm divisions of the Stage Micrometer can be utilized for measurement. Record all measurements and deviations.

ROCKWELL HARDNESS TESTING MACHINES

4.9. The diamond indenter shall be free from surface defects (cracks, chips, pits, etc.) and polished to such an extent that no unpolished part of its surface makes contact with the test piece when the indenter penetrates to a depth of 0.3 mm for Rockwell hardness testing and 0.2 mm for Rockwell superficial hardness testing.

4.10. The Rockwell Testing Machine shall be verified by means of certified test blocks for multiple readings encompassing the ranges in which the machine is used. #### currently has two Rockwell Testing Machines; one is used for Standard Rockwell Testing and the other is used for Superficial Rockwell Testing.

4.11. For Standard Rockwell Testing, a minimum of six indentations shall be measured for each of the following ranges:
HRC: 60 - 68
HRC: 40 - 50
HRC: 23 - 30
HRA: 72 – 80
HRB: 80 – 90
4.12. For Superficial Rockwell Testing a minimum of six indentations shall be measured for each of the following ranges:
HR15N: 60 -70
HR30N: 70 - 80
HR45N: 65 - 75
HR30T: 60 - 70

Having an accreditted lab calibrate these testers will almost certainly be your best approach. For the operational checks you need to have an assortment of certified test blocks within the ranges of your testing activities.

Hope this helps.
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#3
I agree with getting an accredited calibration lab to calibrate the machine.....but checking it is also a good idea.....

You should get certified blocks, and obtain a copy of ASTM E 18 (available through ASTM in the U.S.), and implement a check methodology between calibrations.

Hope this helps.

Hershal
 
J

jhoniegudel

#4
Thanks to Benjamin & Hershal

In our company we have allready use Block refference for hardness tester verification. And we already verificated it. That is enough?
Our customer required for calibrating our hardness tester machine. Does anyone know where we can calibrated our hadness tester machine and can explain how this equipment calibrated?


jhoniegudel
 
M

micron master

#6
Thanks to Benjamin & Hershal

In our company we have allready use Block refference for hardness tester verification. And we already verificated it. That is enough?
Our customer required for calibrating our hardness tester machine. Does anyone know where we can calibrated our hadness tester machine and can explain how this equipment calibrated?


jhoniegudel
Since the testers are in use, you will be performing "indirect verification" by using test blocks. You can find the procedure for this in either ASTM E-18, or ISO 6508-2.

Basically, for the ASTM method, you make 5 readings on three test blocks per scale, one from the low, mid, and high hardness ranges. You only need to verify the scales that you actually use. You may have an outside source do this, or you may do it yourself.

Please feel free to contact our distributor in Singapore, [email protected]. He can either help you himself or direct you to someone in Indonesia who is familiar with the procedure.
 
Thread starter Similar threads Forum Replies Date
A Looking for a third party to calibrate our measure equipment - South Carolina General Measurement Device and Calibration Topics 6
I Can I Calibrate On Demand? General Measurement Device and Calibration Topics 9
S Calibrating our own equipment, can we? Micrometers to calibrate vernier calipers General Measurement Device and Calibration Topics 13
F Crimper Calibration - How do you Control & Calibrate your crimpers? Calibration Frequency (Interval) 13
P How to Calibrate a HP 3458A Digital Multimeter with Fluke 5080A calibrator General Measurement Device and Calibration Topics 1
B Choosing not to calibrate (IATF 16949) IATF 16949 - Automotive Quality Systems Standard 6
K New Kilogram - The process we use to calibrate the standards General Measurement Device and Calibration Topics 6
D How to Calibrate any Digital Sphygmomanometer under use for 2 years in a Hospital Forum News and General Information 1
L Re-calibrate after software update? Calibration Frequency (Interval) 3
J To Calibrate or not to calibrate electronics rack components Manufacturing and Related Processes 5
T How to Calibrate Taper Gages General Measurement Device and Calibration Topics 1
N Calibrate temperature on refrigerated Centrifuges? General Measurement Device and Calibration Topics 2
B Can we calibrate a standard pin with a gauge? General Measurement Device and Calibration Topics 2
D Lesser expensive to calibrate various test weights General Measurement Device and Calibration Topics 5
A How to Calibrate Laser Beam Tester & Radiation Monitor General Measurement Device and Calibration Topics 3
D How to Calibrate and Verify a Heating Oven General Measurement Device and Calibration Topics 3
P How do you calibrate a VNA (Vector Network Analyzer) cal kit? General Measurement Device and Calibration Topics 4
V How to Calibrate Ring Force Gauges Service Industry Specific Topics 1
bobdoering Who says you can't calibrate a steel rule! Funny Stuff - Jokes and Humour 14
S How can I Calibrate a Biosafety Cabinet? General Measurement Device and Calibration Topics 1
C How to Calibrate a Viscosity Tester General Measurement Device and Calibration Topics 3
0 Is it possible to use a threaded ring gage to calibrate the M8 gage? Calibration and Metrology Software and Hardware 4
C How to Calibrate a Medical Infrared Thermometer General Measurement Device and Calibration Topics 3
T How do I calibrate NPT Thread Plug Gages? General Measurement Device and Calibration Topics 9
E Do you need a Certification to Calibrate Gauges? General Measurement Device and Calibration Topics 2
B Calibrate Hole Presence Detection Device? General Measurement Device and Calibration Topics 5
J What do you use to Calibrate 40" Calipers General Measurement Device and Calibration Topics 10
A Guideline to Calibrate Chamber Temp. & Pressure of a Medical Steam Sterilizer General Measurement Device and Calibration Topics 8
S How to calibrate AOI (Automated Optical Inspection) for IOPQ Qualification and Validation (including 21 CFR Part 11) 3
S Laboratory to Calibrate Objective Micrometer needed General Measurement Device and Calibration Topics 4
C Tolerance Estimation of a Load Cell I want to Calibrate General Measurement Device and Calibration Topics 9
D Procedure on How to Calibrate a External Micrometer General Measurement Device and Calibration Topics 1
B ISO/TS 16949 Section 7.6.3 - Use of Manufacturer to Calibrate IATF 16949 - Automotive Quality Systems Standard 21
W How to verify/calibrate a Neptune Water Meter General Measurement Device and Calibration Topics 4
N Laboratory that is able to Calibrate Dew Point Meter up to 60?Cdp - Needed General Measurement Device and Calibration Topics 1
C How to Calibrate and Verify Torque Multiplier Gages General Measurement Device and Calibration Topics 6
C How to calibrate or control patterns for fiberglass General Measurement Device and Calibration Topics 4
N Scale with Internal Calibration - Do I need to calibrate? How to document this? Calibration Frequency (Interval) 8
I Need help finding an Accredited Lab that can Calibrate Brunson 76-RH Survey Transits General Measurement Device and Calibration Topics 5
M Controlling Temperature and Time - Do I need to calibrate the thermostat? IATF 16949 - Automotive Quality Systems Standard 4
D How much Tolerance do I use to Calibrate a 150kgs capacity Scale? General Measurement Device and Calibration Topics 5
M To Calibrate or Not to Calibrate various Weight Scales ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 24
N Thermocouple Inhomogeneity - Need to calibrate a roll of 50m thermocouple type K General Measurement Device and Calibration Topics 3
J Calibrate before use required? Inspection Gages and Measurement Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
R Traceability for Mercury Argon Calibration Source to Calibrate a Spectrometer? General Measurement Device and Calibration Topics 5
P How to Calibrate Laser Micrometer when no Calibration Company in the Country? General Measurement Device and Calibration Topics 3
S 3 Point Calibration or Calibrate to Extreme High and Low? General Measurement Device and Calibration Topics 9
C Seeking a holder/jig to calibrate large OD micrometer calipers (up to 23") General Measurement Device and Calibration Topics 8
S Is it OK to calibrate a measurement device tighter than manufacturer specifications? General Measurement Device and Calibration Topics 18
B Difference between Calibration & Verification and How to calibrate Go/No Go gauge? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4

Similar threads