Hi Peter,

The original discussion was based on the linear transformer. But what working voltage for barrier after 60950 switching power supply should be?

60950 power supply can provide one MOPP only, the insulation, after such SPS, between secondary and applied part should therefore meet one MOPP at least. For SPS, the measured working voltage between pri and sec is 350Vrms/ 670Vpeak, in this case. For barrier after the SPS, I think 240V reference voltage is appropriate. Because if pri shorts to sec, the switching circuit will be destroyed. That means 350Vrms/ 670Vpeak will die.

If I got something wrong, please correct me. Thanks.
 

Peter Selvey

Staff member
Super Moderator
#13
Yes, basically correct.

The standard always requires 2 MOP.

For the purpose of the 350Vrms / 670Vp:

1st MOP is the isolation barrier
2nd MOP is the self desctruction of the 350Vrms in case of fault.

Of course both must be verified by inspection or test, and confirmed to be reliable MOP for regular production.
 

Peter Selvey

Staff member
Super Moderator
#15
It's kind of like the use of earthing.

1st MOP = basic insulation
2nd MOP = earthing + fuse (or circuit breaker)

Although we tend to think of just earthing being the MOP, actually the fuse is also important. In the case of 1st MOP fails, the fuse link is destroyed (or circuit breaker opens) isolating the supply.

So for your SMPS it is similar, in case of 1st MOP fails, switching MOSFET drive circuit is destroyed, stopping the creation of 350Vrms. It is not as neat as fuse/circuit breaker, but it is still the same effect.
 

Roland chung

Trusted Information Resource
#16
Peter, you did remind me of the self destruction of SMPS. But this seems that additional barrier after SMPS to protection against the stress of mains voltage is not needed anymore, because of the shutdown.
 

Peter Selvey

Staff member
Super Moderator
#17
Whether you need another MOPP (or MOOP) for mains voltage after the SMPS stops will depend on the test result and design.

I think it goes back to the original point: if you can design for double insulation at the highest voltage (e.g. 350Vrms), then life is simple.

The alternate is possible but messy, complicated and easy to make mistakes in the rationale.
 
R

raymond02302

#19
Hello peter,

In the post you gave as follow, there seems to be an inappropriate method for analysing the working voltage(W.V) that you use the SFC to determine the W.V.

Consider a transformer with input terminals P1 and P2, and output is S1 and S2, where P2 and S2 have the same polarity, and P1 is 0V (earthed), and the secondary is floating. The SFC analysis shows:

Short P1 to S1: S1 = 0V, S2 = 240V
Short P1 to S2: S1 = 240V, S1 = 0V
Short P2 to S1: S1 = 240V, S2 = 480V
Short P2 to S2: S1 = 0V, S2 = 240V
Working voltage is highest voltage to which the insulation or the component under consideration is, or can be, subjected when the equipment is operating under conditions of normal use (from 60950),
sometimes we should consider the open circuit condition to measure W.V.(from IEC 60664-1).
Therefore, When we measure the W.V, we should not consider the fault condition.

Actually, the key point to affect the W.V is the input and output referencing.


,
 
Last edited by a moderator:

Peter Selvey

Staff member
Super Moderator
#20
I agree, it's not per the definition in the standard.

Mainly I was just trying to show that 480V in SFC was possible. The original post indicates that summation of voltage cannot occur in SFC, but this is not true.

There may be other (rare) SFC where voltage increase well beyond NC, so that using the value measured in NC makes no sense (e.g. failures in the feedback loop of high voltage step up regulator).

We might see future standards modified to consider this.
 
Thread starter Similar threads Forum Replies Date
J Mislabeling - Consider this an FDA notified recall? CFR 806.10 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
R When does the FDA consider a component a medical device? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 17
S How to consider the relevant standards during development of ISO13482:2016 for IVD manufacturing Blood grouping Other Medical Device Related Standards 6
qualprod Criteria for print shop - How do you consider the cycle time in a print shop? Lean in Manufacturing and Service Industries 21
M Medical Device News FDA News - 14-09-18 - Benefit-Risk Factors to Consider for Substantial Equivalence Other US Medical Device Regulations 0
T Question: Do we consider 60601-1-2 (4th ed) EMC testing a mode for a diagnostic port? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
Q What to consider in 6.4 work environment? ISO 9001:2008 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
A Implantable Medical Device Consignment Stock - What to consider? ISO 13485:2016 - Medical Device Quality Management Systems 3
Ajit Basrur FDA issues Guidance Document - Benefits-Risks Factors to consider for 510(K) US Food and Drug Administration (FDA) 1
K Can we consider Operating System as OTS (Off The Shelf) Software? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
K Can I consider My medical Application as MDDS ? Other US Medical Device Regulations 1
P Is it necessary to consider ISO 22002-1 for the implementation of ISO 22000 Food Safety - ISO 22000, HACCP (21 CFR 120) 3
G Points to consider while defining the Quality Policy AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
D What Factors to consider to determine the Number of Auditors Internal Auditing 3
M Machine Feature: Is it possible to consider it Error Proofing? FMEA and Control Plans 6
T Pilot's Checklist for Six Sigma - Steps to Consider when Initiating Projects Six Sigma 3
T Does FDA consider Flow Charts sufficient instruction for SOPs? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
K If auditor fails to consider something which is material to the outcome of the audit? Internal Auditing 19
N Note interpretation - Ensure that the preservation of the product has been consider ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
A How to write Quality Manual? Point to consider Quality Management System (QMS) Manuals 2
S Do you consider 'unintended' process output(s) while mapping a process? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
Sidney Vianna Tough talk on Detroit Three doesn't consider suppliers World News 0
E Do you consider this a Top Management Nonconformity? Nonconformance and Corrective Action 19
S Should we consider Precision of Machine Reading when setting Inspection Specs? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
T Can I consider Reliability Test Plan part of the DVP&R, or the other way round Reliability Analysis - Predictions, Testing and Standards 3
D Should low volume supplier PPM comparisons consider level of opportunity for error? Quality Tools, Improvement and Analysis 2
S Painting on Floor of Machine Shop - What Aspects/Impacts and Hazards to consider? Miscellaneous Environmental Standards and EMS Related Discussions 10
S CSQE : Do they consider the sw development ?Decision making? experience? Professional Certifications and Degrees 1
P What factors should I consider so as to check how "lean" a company is? Lean in Manufacturing and Service Industries 12
Marc Things to consider in a Beer FMEA - Beer Troubleshooting FMEA and Control Plans 7
Marc Where would you retire? Would you consider becoming an 'Expat'? Immigration and Expatriate Topics 29
Casana How to select a consultant? What criteria to consider and look for when looking Consultants and Consulting 5
Govind Metrology Laboratory 'Classes' - Atmospheric, Ergonomic and Other Factors to consider General Measurement Device and Calibration Topics 6
A Gage R&R analysis - Do I have to consider the sample specifications? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
RoxaneB Several Key Indicators a Company Should Consider Using Misc. Quality Assurance and Business Systems Related Topics 40
T Do You Consider Your Quality Policy Valuable to your Organization? (Poll) Other ISO and International Standards and European Regulations 44
I FMEA Issue - Consider all inspections that we have like a processes? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
M How can we encourage our supplier base to consider ISO14001 Supplier Quality Assurance and other Supplier Issues 2
B 3-phase supply, overvoltage catogory, line-to-neutral and working voltage (IEC 61010-1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
John Broomfield Guidance on Safe Working During Covid-19 Pandemic Misc. Quality Assurance and Business Systems Related Topics 5
S Experience working with TUV SUD or Rheinland, and/or BSI Registrars and Notified Bodies 5
G Working on a Root Cause Problem Solving, Root Cause Fault and Failure Analysis 11
M Informational TGA-led IMDRF Personalised Medical Devices working group meets in Canberra Medical Device and FDA Regulations and Standards News 0
M Informational EU – Minutes of the 24 July 2019 SCHEER Working Group on safety of breast implants in relation to anaplastic large cell lymphoma (BIA-ALCL) meeting Medical Device and FDA Regulations and Standards News 0
Watchcat MDD, MDR/IVDR, Working Groups, Expert Panels, MEDDEVs and Other EU Guidances EU Medical Device Regulations 32
T Is anyone working with N299.1 (Supply/service to nuclear power plants)? Various Other Specifications, Standards, and related Requirements 0
M Informational EU – 12th Meeting of the Working Group on Guidelines on benefit – risk assessment of Phthalates in Medical Devices Medical Device and FDA Regulations and Standards News 0
G Anyone working with or planning to do business in the CBD (cannabidiol) industry? US Food and Drug Administration (FDA) 1
M Applicability of Means of Protection, working voltage in an Automated External Defibrillator IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
Sidney Vianna ISO 9001 News Tirelessly Improving the Brand Integrity of ISO 9001 - Working Group under ISO TC 176 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 42

Similar threads

Top Bottom