How to determine Failure Tolerance Time of Laser Equipment?

Roland chung

Trusted Information Resource
#1
Hello,

Anyone in the forum has experience with laser equipment?

I am wondering how to determine the failure tolerance time of laser equipment which required by the clause 51.2 in 60601-2-22.

The laser equipment has a laser power monitoring system (by sensing current). 60601-2-22 requires the monitoring shall be carried out at intervals shorter than the failure tolerance time. Our monitoring system is real time. Is it necessary to evaluate the failure tolerance time?

Thanks and regards,
Roland
 
Elsmar Forum Sponsor
P

plantengineer

#2
Why type of laser equipment? Is it a laser interferometer? If so, their should probably be a failure tolerance associated with it since measurements with those types of lasers are continuous, and based on pulses. Might need to verify the measurement rate is what the manufacturer says it is more than the failure tolerance time.

I believe that laser autocollimators have a calibration responses measured in arc-seconds. It measures rotation about the axis, so in this case, IU would assume their would not be a failure time tolerance associated with this type.
 

Peter Selvey

Staff member
Super Moderator
#3
This is an old requirement which has some historical basis. Older lasers were not very stable, so the output could fluctuate significantly even in normal condition. The feedback system, whether used for closed loop or open loop control should be fast enough to respond in time to such fluctuations.

In a modern lasers the cavity designs are more stable. Feedback loops can be tuned such that there is no significant overshoot/settling time to a step change. In such a case, the response time of the feedback loop is not critical.

In the event of a genuine fault, the standard specifies a 100ms response time for the protection system.
 

Roland chung

Trusted Information Resource
#4
Hello both, thank you for your input.

Why type of laser equipment?
This is Nd:YAG laser equipment.

This is an old requirement which has some historical basis.
The requirement remain unchanged in 60601-2-22:2007. So I think it is not outmoded.

Any thoughts on the failure tolerance time?
 
P

plantengineer

#5
I would recommend contacting the laser manufacturer. They should know what the capabilities and requirements of their laser should be. Too difficult to set the parameter yourself.
 

Peter Selvey

Staff member
Super Moderator
#6
Yes, I saw it is word for word copied in the 2007 edition.

But I still think the requirement is clear in a historical background.

I have worked with many types of lasers since around 1995 in Australia and Japan (both countries seem to be hot spots for medical lasers). Some were very unstable, others perfectly stable. Some used a lot of tricks to get a stable output.

The impression I had was that with the invention of new types of lasers (e.g. diode lasers), improvement in cavity design for existing types (e.g. CO2 lasers), new ways to get to useful wavelengths (e.g. frequency doubling) and improvements in feedback loops (e.g. speed of sensors, op-amps, digital processing etc), stability in normal condition is becoming less of an issue.

But there are lots of laser types and some types may still be unstable in normal condition, thus the requirement should remain.

If you want to measure the "monitoring time" I recommend investigate laser output response to step change (i.e. step response). Even "real time" control is not perfect, and takes time to settle and there may be overshoot.

The settling time and overshoot energy (if applicable) should be negligible with respect to the clinical application ("failure tolerance time"). Some applications like ophthalmic work have very short tolerance time, others for general surgical procedures may have long tolerance time.
 

Marcelo

Inactive Registered Visitor
#7
The requirement remain unchanged in 60601-2-22:2007. So I think it is not outmoded.
It's not unusual to standards to keep requirements with can be easily fulfilled just per se... for example, the third edition was to have removed the requirements for flamable anaesthetic gases because no one in the world uses them anymore. However, some people said that they could use it someday, so it remained, as an Annex, in the standard.

Another example were the old requirements for CRTs...the usual problems with CRTs were solved a lot of time ago, but the standard still have requirements.

Those in a way might be there just to keep things in check.

Just a generic comment, anyway...not totally sure this is the case for the laser requirements.
 

Roland chung

Trusted Information Resource
#8
Actually, I am not concerned the "monitoring time" since this time is up to the speed of feedback loop. It is not difficult to obtain.

The question of failure tolerance time is still open. I have worked with infusion pump. It is clear to determine the failure tolerance time (FTT) for air infusion. That is the volume of the ADMINISTRATION SET between the air detector and the venous cannula connected to it divided by the maximum flow rate of the pump.

For the laser equipment, the FTT is indeed short (mS). But how to calculate it?
 

Peter Selvey

Staff member
Super Moderator
#9
It's different than an infusion pump: in that case it is assumed any air is bad (death), so fault tolerance time is based on time it takes the air bubble to reach the patient. This is true of any situation where it takes time for something "bad" to travel to the patient.

For a laser, the "bad" thing (e.g. overpower) will for all intentional purposes reach the patient instantaneously.

But, it takes time for the tissue to respond, increase in temperature, and be damaged. And, even if damage occurs, the severity may not be the same (in many locations the body can deal with tissue damage). So the fault tolerance time is really the time in which significant or serious unintended damage to the tissue occurs.

This depends very much on the clinical application, type of laser etc.
 

Roland chung

Trusted Information Resource
#10
Do you think it is necessary for us to simulate the clinical application on mimic tissue and record the tissue damage time (i.e. fault tolerance time)?
 
Thread starter Similar threads Forum Replies Date
J UDI-DI how should we interpret Device version or model to determine if a new UDI-DI is needed? EU Medical Device Regulations 0
L How to determine / validate Medical Device Storage Conditions ISO 13485:2016 - Medical Device Quality Management Systems 1
John C. Abnet ISO 9001 4.4.1 "...shall determine the processes needed..." ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 72
S Is using ANSI/ASQ Z1.4-2008 the correct sampling plan to determine Pass/Fail of Apparel measurements? AQL - Acceptable Quality Level 4
E How do you determine if hipot is necessary for a cable product? Manufacturing and Related Processes 8
J Standard used to determine if check fixture gauge is acceptable Reliability Analysis - Predictions, Testing and Standards 2
M Measurement Error - How to determine what is acceptable? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
S How to determine & document Organizational Knowledge of a company Document Control Systems, Procedures, Forms and Templates 4
O Monitoring performance - How do I determine performance measurement basis within my organization? Misc. Quality Assurance and Business Systems Related Topics 4
T IEC 60601-1 - Risk assessment to determine the liquid - 11.6.3 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
Ashok sunder How to determine the internal and external issues in ISO 45001 standard? Occupational Health & Safety Management Standards 2
C ICP versus AA: How to determine appropriate specification for assay method verification Qualification and Validation (including 21 CFR Part 11) 1
samer Do we have to determine Educational level for all Staff as pr 7.2 b clause? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
A Calibration of Pressure Gauge, How to Determine Calibration Tolerance General Measurement Device and Calibration Topics 1
S Supplier Qualification - How do we determine which suppliers to qualify? ISO 13485:2016 - Medical Device Quality Management Systems 9
K ISO 13485:2016 Cl. 4.2.3 - Determine QMS Processes, Monitoring, Measuring, etc. ISO 13485:2016 - Medical Device Quality Management Systems 4
A Compliance Obligations - Implementing 6.1.1 and 6.1.3 NOTE - Determine risks and... ISO 14001:2015 Specific Discussions 1
B Auditing Senior Management to determine the Effectiveness of Management Processes IATF 16949 - Automotive Quality Systems Standard 6
R What testing would you run to determine if a device is "Clean"? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
B Maximum Material Condition with Bonus - Determine Cpk for this one kpc Measurement Uncertainty (MU) 2
Q Tissue Adherence Test to determine how Tissue sticks to Gauze Other US Medical Device Regulations 2
C DOE - How to determine optimal settings and transposing data Statistical Analysis Tools, Techniques and SPC 2
insect warfare Cannot determine what type of medical device this product is.... 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
Buckyb Are KPIs a requirement to determine Process Effectiveness? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 14
M Is a Real Time Stability Study required to determine Expiration Date? EU Medical Device Regulations 9
K How to determine rated input power for the markings? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
I How to determine the thread wire size needed - Buttress Thread Gauge General Measurement Device and Calibration Topics 3
M How to determine Process Capability using Minitab 17 Capability, Accuracy and Stability - Processes, Machines, etc. 6
V How does ASQ determine Exam Cutoff Marks Professional Certifications and Degrees 2
R How to Determine Working Voltage & Tolerance for Isolation Barriers? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
apestate How to determine all regulatory and statutory requirements to be met? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
J Gage R&R Study (Crossed) - How to determine Process Tolerance Using Minitab Software 4
D How can I determine a process metrics? Manufacturing and Related Processes 1
D Can I use the % Tolerance to determine the actual process tolerance? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
R How to determine the Thickness tolerance of the Cold Rolled Steel Manufacturing and Related Processes 6
T Technical regulations that determine the acceptable limits of cigarettes General Measurement Device and Calibration Topics 4
J How does Layout Inspection determine Reference Value? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
Q PPAP - How to determine the "Design Record Change Level" APQP and PPAP 2
K 7.6 c) c) Have Identification in order to determine Calibration Status Calibration Frequency (Interval) 3
S How to determine part that is most difficult to sterilize Other Medical Device Related Standards 10
G How to determine if a Standard is Auditable General Auditing Discussions 15
L Who and when to determine AQL for our product inspection AQL - Acceptable Quality Level 1
R 4 Factors 3 Level DOE to determine which factors are significant Using Minitab Software 2
C How to determine a sample size for comparison purposes ? Quality Tools, Improvement and Analysis 2
S How to determine Preventive Maintenance and Calibration Grace Periods General Measurement Device and Calibration Topics 4
J How to determine a Capability Study for Low-Volume Suppliers Capability, Accuracy and Stability - Processes, Machines, etc. 3
C Statistics equation to determine the number of samples in a lot Statistical Analysis Tools, Techniques and SPC 8
S How to determine the Air Clearance for Floated Secondary Circuit IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
somashekar How to Determine the Necessary Employee Competence ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
F ISO 22000 Clause 7.4.4 to determine whether a control is a CCP or and OPRP Food Safety - ISO 22000, HACCP (21 CFR 120) 2

Similar threads

Top Bottom