How to determine Preventive Maintenance and Calibration Grace Periods

SGquality

Quite Involved in Discussions
#1
Any thoughts on determining the grace period for Preventive Maintenance or Calibration schedules for daily, monthly, semi-annually and yearly activities ?

Is there any ISO or regulatory documents to support this claim ?

Thank you !
 
Elsmar Forum Sponsor

Jerry Eldred

Forum Moderator
Super Moderator
#2
Re: How to determine the PM and Calibration grace periods ?

I think the general answer is none. I ran a high quality calibration lab a few years back where they were so strict that anyone who did not turn in their instruments on time got into trouble. They had about zero tolerance for late turn ins for calibration.

The philosophical answer is that the interval is the defined period of time to a given confidence that an instrument can be expected to remain in its specifications.

It is often written into quality systems to allow for extensions to calibration interval based on historical performance (if there is data to support that the instrument has a documented probability to remain within specifications).

If you don't have to comply with any ISO or other document, some companies to write into policy the ability to includle a grace period. But, if you are audited by an ISO auditor or other such auditor, you might have to defend such actions.

One way I can think of to build in a legitimate (in my opinion) "grace period" would be to make the due date 15 days prior to the end of the interval. For a one year calibration, if you calibrated on Jan 15 2013, the due date could be Dec 31 2013, but you have to define in your quality system that the due date will occur 15 days prior to the defined interval. The subsequent 15 day period could be the "grace period" (but you might not want to use that term) - perhaps say you allow that 15 day window for the instrument to be delivered for calibration. You could get your grace period and still abide by the defined intervals. Just my thoughts. Perhaps others may find a problem with the idea.
 

dgriffith

Quite Involved in Discussions
#3
Re: How to determine the PM and Calibration grace periods ?

One way I can think of to build in a legitimate (in my opinion) "grace period" would be to make the due date 15 days prior to the end of the interval. For a one year calibration, if you calibrated on Jan 15 2013, the due date could be Dec 31 2013, but you have to define in your quality system that the due date will occur 15 days prior to the defined interval. The subsequent 15 day period could be the "grace period" (but you might not want to use that term) - perhaps say you allow that 15 day window for the instrument to be delivered for calibration. You could get your grace period and still abide by the defined intervals. Just my thoughts. Perhaps others may find a problem with the idea.
The 15-day period would have to be company policy; the quality document could say 'all MTE ar required to be turned in 15 days prior to the due date on the cal label'. If you put the 15-day-early due date on the asset, and they don't remove it from service as they should, it would be overdue anyway. I would leave the 1-yr due date on it and require them to turn it in early as a matter of policy.
 

Michael_M

Trusted Information Resource
#4
Re: How to determine the PM and Calibration grace periods ?

We have, per procedures, the ability to extend calibration 90 days when justified (in use for example) so long as the reason is reasonably justified and documented.

On thread gages, I have the ability to extend calibration 1 year if the gage has not been used since last calibration. Per our procedure, if the gage is not used after the second year, it gets pulled from service and placed in bond. If it is determined that the gage is needed later, it can be put back into active service and will require calibration after completing the current job. We have some thread gages that get used maybe once every 3-4 years.

All this is specified in our internal procedures. We got to determine what we wanted to do.
 
Thread starter Similar threads Forum Replies Date
L How to determine / validate Medical Device Storage Conditions ISO 13485:2016 - Medical Device Quality Management Systems 1
John C. Abnet ISO 9001 4.4.1 "...shall determine the processes needed..." ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 72
S Is using ANSI/ASQ Z1.4-2008 the correct sampling plan to determine Pass/Fail of Apparel measurements? AQL - Acceptable Quality Level 4
E How do you determine if hipot is necessary for a cable product? Manufacturing and Related Processes 8
J Standard used to determine if check fixture gauge is acceptable Reliability Analysis - Predictions, Testing and Standards 2
M Measurement Error - How to determine what is acceptable? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
S How to determine & document Organizational Knowledge of a company Document Control Systems, Procedures, Forms and Templates 4
O Monitoring performance - How do I determine performance measurement basis within my organization? Misc. Quality Assurance and Business Systems Related Topics 4
T IEC 60601-1 - Risk assessment to determine the liquid - 11.6.3 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
Ashok sunder How to determine the internal and external issues in ISO 45001 standard? Occupational Health & Safety Management Standards 2
C ICP versus AA: How to determine appropriate specification for assay method verification Qualification and Validation (including 21 CFR Part 11) 1
samer Do we have to determine Educational level for all Staff as pr 7.2 b clause? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
A Calibration of Pressure Gauge, How to Determine Calibration Tolerance General Measurement Device and Calibration Topics 1
S Supplier Qualification - How do we determine which suppliers to qualify? ISO 13485:2016 - Medical Device Quality Management Systems 6
K ISO 13485:2016 Cl. 4.2.3 - Determine QMS Processes, Monitoring, Measuring, etc. ISO 13485:2016 - Medical Device Quality Management Systems 4
A Compliance Obligations - Implementing 6.1.1 and 6.1.3 NOTE - Determine risks and... ISO 14001:2015 Specific Discussions 1
B Auditing Senior Management to determine the Effectiveness of Management Processes IATF 16949 - Automotive Quality Systems Standard 6
R What testing would you run to determine if a device is "Clean"? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
B Maximum Material Condition with Bonus - Determine Cpk for this one kpc Measurement Uncertainty (MU) 2
Q Tissue Adherence Test to determine how Tissue sticks to Gauze Other US Medical Device Regulations 2
C DOE - How to determine optimal settings and transposing data Statistical Analysis Tools, Techniques and SPC 2
insect warfare Cannot determine what type of medical device this product is.... 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
Buckyb Are KPIs a requirement to determine Process Effectiveness? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 14
M Is a Real Time Stability Study required to determine Expiration Date? EU Medical Device Regulations 9
K How to determine rated input power for the markings? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
I How to determine the thread wire size needed - Buttress Thread Gauge General Measurement Device and Calibration Topics 3
M How to determine Process Capability using Minitab 17 Capability, Accuracy and Stability - Processes, Machines, etc. 6
V How does ASQ determine Exam Cutoff Marks Professional Certifications and Degrees 2
R How to Determine Working Voltage & Tolerance for Isolation Barriers? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
apestate How to determine all regulatory and statutory requirements to be met? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
J Gage R&R Study (Crossed) - How to determine Process Tolerance Using Minitab Software 4
D How can I determine a process metrics? Manufacturing and Related Processes 1
D Can I use the % Tolerance to determine the actual process tolerance? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
R How to determine the Thickness tolerance of the Cold Rolled Steel Manufacturing and Related Processes 6
T Technical regulations that determine the acceptable limits of cigarettes General Measurement Device and Calibration Topics 4
J How does Layout Inspection determine Reference Value? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
Q PPAP - How to determine the "Design Record Change Level" APQP and PPAP 2
K 7.6 c) c) Have Identification in order to determine Calibration Status Calibration Frequency (Interval) 3
S How to determine part that is most difficult to sterilize Other Medical Device Related Standards 10
G How to determine if a Standard is Auditable General Auditing Discussions 15
L Who and when to determine AQL for our product inspection AQL - Acceptable Quality Level 1
R 4 Factors 3 Level DOE to determine which factors are significant Using Minitab Software 2
C How to determine a sample size for comparison purposes ? Quality Tools, Improvement and Analysis 2
J How to determine a Capability Study for Low-Volume Suppliers Capability, Accuracy and Stability - Processes, Machines, etc. 3
C Statistics equation to determine the number of samples in a lot Statistical Analysis Tools, Techniques and SPC 8
S How to determine the Air Clearance for Floated Secondary Circuit IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
somashekar How to Determine the Necessary Employee Competence ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
F ISO 22000 Clause 7.4.4 to determine whether a control is a CCP or and OPRP Food Safety - ISO 22000, HACCP (21 CFR 120) 2
D Best Minitab Tool to determine a Statistically Significant Trend Using Minitab Software 5
E How to determine the Lot Size for the OQ Sampling Plan Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11

Similar threads

Top Bottom