How to determine rated input power for the markings?

K

Kleibeuker

#1
Within our organization we have some discussion wrt rated input power.
Approach A:
Sum up all the maximum power of the supplies that are in the MEE. This is about 200W.
Rationale: IEC60601-1 3ed Clause 4.11 last sentence: A supplier certification may be used in place of the above measurement as the basis for steady state current or input power specification.

Approach B:
Measure the input power (for 100/240Vac 50/60Hz) and either use single value or a range?
And include some realistic design margin (eg 5%) This is about 80W.
Rationale: IEC60601-1 3ed Clause 4.11: The steady state MEASURED input of the MEE or MES at RATED voltage.

The result in both approaches is imho too big 200W vs 80W.

I think that rated value on the marking should make sense and therefor e not be too much over rated?

What do you prefer?
 
Elsmar Forum Sponsor

Peter Selvey

Staff member
Moderator
#3
I think the last sentence in Clause 4.11 is intended as a get out clause for labs testing very large equipment (permanently installed, 3 phase, linear accelerators etc running at 50kVA type of equipment ...).

There is no upper limit on the rating. The standard allows a 10% tolerance so you can select a typical value (e.g. 75W), and it is no problem if the actual value is 80W.

That said, most manufacturers usually include a moderate margin, e.g. a device with a real input of 75W would normally use a 100W rating. The reason I guess is to allow for upgrades, new features etc that can be added to the secondary circuits without needing the marking to be updated.

I agree that it does make sense not to go too far with the overrating, but it is not as critical for small low power equipment (<150VA). In practice a user worried about overloading the circuit (causing a breaker to trip) would be looking out for equipment around 500VA or higher. Excessive overrating in high power equipment might cause a user to unnecessarily worry.

Note that a marking in W is unusual for low power equipment. Usually the power factor is less than 0.9 so a marking in VA is required. Smaller power supplies may not have power factor correction circuits, or they may not be effective at small loads.
 
Thread starter Similar threads Forum Replies Date
S Is using ANSI/ASQ Z1.4-2008 the correct sampling plan to determine Pass/Fail of Apparel measurements? AQL - Acceptable Quality Level 4
E How do you determine if hipot is necessary for a cable product? Manufacturing and Related Processes 8
J Standard used to determine if check fixture gauge is acceptable Reliability Analysis - Predictions, Testing and Standards 2
M Measurement Error - How to determine what is acceptable? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
S How to determine & document Organizational Knowledge of a company Document Control Systems, Procedures, Forms and Templates 4
O Monitoring performance - How do I determine performance measurement basis within my organization? Misc. Quality Assurance and Business Systems Related Topics 4
T IEC 60601-1 - Risk assessment to determine the liquid - 11.6.3 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
Ashok sunder How to determine the internal and external issues in ISO 45001 standard? Occupational Health & Safety Management Standards 2
C ICP versus AA: How to determine appropriate specification for assay method verification Qualification and Validation (including 21 CFR Part 11) 1
samer Do we have to determine Educational level for all Staff as pr 7.2 b clause? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
A Calibration of Pressure Gauge, How to Determine Calibration Tolerance General Measurement Device and Calibration Topics 1
S Supplier Qualification - How do we determine which suppliers to qualify? ISO 13485:2016 - Medical Device Quality Management Systems 6
K ISO 13485:2016 Cl. 4.2.3 - Determine QMS Processes, Monitoring, Measuring, etc. ISO 13485:2016 - Medical Device Quality Management Systems 4
A Compliance Obligations - Implementing 6.1.1 and 6.1.3 NOTE - Determine risks and... ISO 14001:2015 Specific Discussions 1
B Auditing Senior Management to determine the Effectiveness of Management Processes IATF 16949 - Automotive Quality Systems Standard 6
R What testing would you run to determine if a device is "Clean"? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
B Maximum Material Condition with Bonus - Determine Cpk for this one kpc Measurement Uncertainty (MU) 2
Q Tissue Adherence Test to determine how Tissue sticks to Gauze Other US Medical Device Regulations 2
C DOE - How to determine optimal settings and transposing data Statistical Analysis Tools, Techniques and SPC 2
insect warfare Cannot determine what type of medical device this product is.... 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
Buckyb Are KPIs a requirement to determine Process Effectiveness? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 14
M Is a Real Time Stability Study required to determine Expiration Date? EU Medical Device Regulations 9
I How to determine the thread wire size needed - Buttress Thread Gauge General Measurement Device and Calibration Topics 3
M How to determine Process Capability using Minitab 17 Capability, Accuracy and Stability - Processes, Machines, etc. 6
V How does ASQ determine Exam Cutoff Marks Professional Certifications and Degrees 2
R How to Determine Working Voltage & Tolerance for Isolation Barriers? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
apestate How to determine all regulatory and statutory requirements to be met? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
J Gage R&R Study (Crossed) - How to determine Process Tolerance Using Minitab Software 4
D How can I determine a process metrics? Manufacturing and Related Processes 1
D Can I use the % Tolerance to determine the actual process tolerance? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
R How to determine the Thickness tolerance of the Cold Rolled Steel Manufacturing and Related Processes 6
T Technical regulations that determine the acceptable limits of cigarettes General Measurement Device and Calibration Topics 4
J How does Layout Inspection determine Reference Value? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
Q PPAP - How to determine the "Design Record Change Level" APQP and PPAP 2
K 7.6 c) c) Have Identification in order to determine Calibration Status Calibration Frequency (Interval) 3
S How to determine part that is most difficult to sterilize Other Medical Device Related Standards 10
G How to determine if a Standard is Auditable General Auditing Discussions 15
L Who and when to determine AQL for our product inspection AQL - Acceptable Quality Level 1
R 4 Factors 3 Level DOE to determine which factors are significant Using Minitab Software 2
C How to determine a sample size for comparison purposes ? Quality Tools, Improvement and Analysis 2
S How to determine Preventive Maintenance and Calibration Grace Periods General Measurement Device and Calibration Topics 4
J How to determine a Capability Study for Low-Volume Suppliers Capability, Accuracy and Stability - Processes, Machines, etc. 3
C Statistics equation to determine the number of samples in a lot Statistical Analysis Tools, Techniques and SPC 8
S How to determine the Air Clearance for Floated Secondary Circuit IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
somashekar How to Determine the Necessary Employee Competence ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
F ISO 22000 Clause 7.4.4 to determine whether a control is a CCP or and OPRP Food Safety - ISO 22000, HACCP (21 CFR 120) 2
D Best Minitab Tool to determine a Statistically Significant Trend Using Minitab Software 5
E How to determine the Lot Size for the OQ Sampling Plan Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
A How to determine if product is a Medical Device in Australia Other Medical Device Regulations World-Wide 2
R How to determine the ratings of components in primary circuit IEC 60601 - Medical Electrical Equipment Safety Standards Series 8

Similar threads

Top Bottom