How to determine the Lot Size for the OQ Sampling Plan
E
edwardkwan
Hi guys,
I need your expertise for the issue I have with respect to getting the justification on the lot size of the below:
Operational Qualification using with the below conditions:
  • Single sampling plan for attribute or pass/fail data with c=0.
  • Sample selected is representative
  • For 95% Confidence:
    LTPD, Reliability, Single Sampling Plan Parameters, AQL α=0.05
    5 95 n = 59, c = 0 0.09%
    3 97 n = 99, c = 0 0.05%
    1 99 n = 299, c = 0 0.02%
  • LTPD β=0.05
After the risk analysis done by our customer, the plan 95% Confidence and 99% Reliability, sample size out to be 299 with c=0 (n=ln(1-Confidence)/ln(Reliability) has been selected for the attribute plan.
As indicated by our customer's specification, the batch size,N, in this case is assumed to be 2,990 pcs (N=10xn).
The issue is, I have trouble trying to find out the rationale for 10xn for justification of the statistical rationale. Any statistical reference and explanation would be of great help and much appreciated. By the way we are Contract manufacturer for plastic injection molding for Medical components.
Statistical Steven
Statistician
Hi guys,
  • LTPD β=0.05
Keep looking for the reference to 10xn...because it does not exist. There is not relationship between the sample size and lot size using the ln(1-confidence)/ln(reliability). It assumes a "large" population of parts to be sampled from. I believe they just "made up" the 10xn.
E
edwardkwan
Hi Steven,
Thanks for your reply. It may be from their assumption as their requirement indicated that "This sampling plan is based on Type A OC Curve, which describes protection for individual batches (batches not from a continuously running process) assuming approximately 3000 piece batches (Batch size N approximately 10X sample size n)". With that being said, will it make any difference as far as statistical validity is concerned if the lot size never factored in the first place as I foresee the question of proportion ( n sample size out of Lot size of N) will come into play. And also if the Lot size does not matter or valid for the statistical rationale, I am questioning economical and practicality of their requirement.
Appreciate everyone's help on this.
Rgds,
Ed
Lot size does not matter (unless it is a very small lot adn you are no where near that size)
The ancient Mil-Std/ANSI standards on AQL sampling plans do take lot size into account but they were the result of negotitions more than statistics.
There is no statistically valid reason to take lot size into account. The ratio you are being quoted was pulled out of thin air.
Of course they may be using this 'reasoning' to ensure a sufficiently scaled batch for OQ - but this is a very convoluted way of doing that.
I might question the use of AQL based sample plans for OQ but that is another thread.
COECON
Having some similar sampling number rational problems. Can you supply me with the
formula with exact result of = 59 please
:bigwave:hi COECON: can you provide more context for your question?
I can think of about a billion equations that will result in an answer = 59, including the year I was born in :notme:
COECON
Hi Bev D,
thanks for you quick reply:
Ok I know there is no relationship between sample size and lot size. But there are a lot of recommendations to use at least a sample size of 59 because of statistic evidence.
I am preparing a rational who the recommendation of a attributive sample test is recommended on the base of 95% reliance and 95% confidence using a binomial distribution.
There should be no failure in the samples!
Im struggling wit the formula giving the 59 as a result!
Would be happy for your help
Statistical Steven
Statistician
Hi Bev D,
For c=o plans, the general formual is n=ln(1-Confidence)/ln(Reliability)
So for 95% confidence/90% reliability it is 28
Here are some other examples
Conf Rel. n
0.95 0.9 28
0.95 0.95 58
0.95 0.99 298
0.99 0.9 44
0.99 0.95 90
0.99 0.99 458
mdurivage
Steven's method is what I prefer to use. The only thing is make sure the numbers you use are risk based and spelled out in a procedure.
We modify our sample size based on lot size only when the LOT size is less than ~about 50. This choice is rather arbitrary as it will depend on how accurate we wnat our estimate to be.
When lot sizes get 'small', attribute (categorical, pass/fail) sampling results will follow a hypergeometric distribution and not the general approximation of the Binomial which applies to 'large' lot sizes (theoretically infinite, in practice greater than ~50). The other rule of thumb for adjusting yoru sampel size for lot size is when the sample size is ~ 5% of the lot size or greater...I rarely use this approach as I've found in practice that it really make no practical difference but I do use the lot size <50 rule...
Technically speaking the use of the hypergeometric applies when we sample without replacement from a finite lot. This is because the probability of finding something in a lot changes as the lot size decreases when we pull a sample and don't replace it. (this is the probability that we all despised in high scholl or college: you are getting dressed in the dark and you 20 pairs of socks randomly scattered in your sock drawer. 10 pairs are black, 5 pairs are blue, 4 pairs are brown and 1 pair is pink. what are the odds that you pick at least 1 pink sock? my answer was always: none becuase I am going to turn the lights on)
the Binomial applies perfectly when we are sampling from an infinite population: if you flip a coin...this also works when you sample with replacement.
Neither of these situations fit the practical industrial world. Fortunately in practice once the lot gets over~50 the differences in teh resutlsbetweent he two distribtuions is so small as to have no practical meaning; the differences will have no effect on the ultimate decision we make.
The hypergeometric is complicated to calculate so the 'rule of thumb' was established years ago and is mostly accepted without the mathematical curves since they're a bit of work to generate. Not that I'm against the propagation of knowledge I just would rather re-derive other more impactful rules than this one...perhaps one of our other statisticians has this derivation handy for posting?
The formula used to modify the sample size (calculated the 'standard way' without regard to lot size) is generally taken to be: N/(N+n) where N = lot size and n = sample size.
C Statistics equation to determine the number of samples in a lot Statistical Analysis Tools, Techniques and SPC 8
R How to determine LTPD (lot tolerance percent defective) or RQL Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
B 105E/Z1.4 - How do you determine acceptance of an inspected lot? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
J UDI-DI how should we interpret Device version or model to determine if a new UDI-DI is needed? EU Medical Device Regulations 0
L How to determine / validate Medical Device Storage Conditions ISO 13485:2016 - Medical Device Quality Management Systems 1
John C. Abnet ISO 9001 4.4.1 "...shall determine the processes needed..." ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 72
E How do you determine if hipot is necessary for a cable product? Manufacturing and Related Processes 8
S How to determine & document Organizational Knowledge of a company Document Control Systems, Procedures, Forms and Templates 4
O Monitoring performance - How do I determine performance measurement basis within my organization? Misc. Quality Assurance and Business Systems Related Topics 4
T IEC 60601-1 - Risk assessment to determine the liquid - 11.6.3 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
Ashok sunder How to determine the internal and external issues in ISO 45001 standard? Occupational Health & Safety Management Standards 2
C ICP versus AA: How to determine appropriate specification for assay method verification Qualification and Validation (including 21 CFR Part 11) 1
samer Do we have to determine Educational level for all Staff as pr 7.2 b clause? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
S Supplier Qualification - How do we determine which suppliers to qualify? ISO 13485:2016 - Medical Device Quality Management Systems 9
K ISO 13485:2016 Cl. 4.2.3 - Determine QMS Processes, Monitoring, Measuring, etc. ISO 13485:2016 - Medical Device Quality Management Systems 4
A Compliance Obligations - Implementing 6.1.1 and 6.1.3 NOTE - Determine risks and... ISO 14001:2015 Specific Discussions 1
B Auditing Senior Management to determine the Effectiveness of Management Processes IATF 16949 - Automotive Quality Systems Standard 6
R What testing would you run to determine if a device is "Clean"? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
B Maximum Material Condition with Bonus - Determine Cpk for this one kpc Measurement Uncertainty (MU) 2
Q Tissue Adherence Test to determine how Tissue sticks to Gauze Other US Medical Device Regulations 2
C DOE - How to determine optimal settings and transposing data Statistical Analysis Tools, Techniques and SPC 2
insect warfare Cannot determine what type of medical device this product is.... 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
Buckyb Are KPIs a requirement to determine Process Effectiveness? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 14
M Is a Real Time Stability Study required to determine Expiration Date? EU Medical Device Regulations 9
K How to determine rated input power for the markings? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
I How to determine the thread wire size needed - Buttress Thread Gauge General Measurement Device and Calibration Topics 3
M How to determine Process Capability using Minitab 17 Capability, Accuracy and Stability - Processes, Machines, etc. 6
V How does ASQ determine Exam Cutoff Marks Professional Certifications and Degrees 2
R How to Determine Working Voltage & Tolerance for Isolation Barriers? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
apestate How to determine all regulatory and statutory requirements to be met? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
J Gage R&R Study (Crossed) - How to determine Process Tolerance Using Minitab Software 4
D How can I determine a process metrics? Manufacturing and Related Processes 1
D Can I use the % Tolerance to determine the actual process tolerance? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
R How to determine the Thickness tolerance of the Cold Rolled Steel Manufacturing and Related Processes 6
T Technical regulations that determine the acceptable limits of cigarettes General Measurement Device and Calibration Topics 4
J How does Layout Inspection determine Reference Value? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
Q PPAP - How to determine the "Design Record Change Level" APQP and PPAP 2
K 7.6 c) c) Have Identification in order to determine Calibration Status Calibration Frequency (Interval) 3
S How to determine part that is most difficult to sterilize Other Medical Device Related Standards 10
G How to determine if a Standard is Auditable General Auditing Discussions 15
L Who and when to determine AQL for our product inspection AQL - Acceptable Quality Level 1
R 4 Factors 3 Level DOE to determine which factors are significant Using Minitab Software 2
C How to determine a sample size for comparison purposes ? Quality Tools, Improvement and Analysis 2
S How to determine Preventive Maintenance and Calibration Grace Periods General Measurement Device and Calibration Topics 4
J How to determine a Capability Study for Low-Volume Suppliers Capability, Accuracy and Stability - Processes, Machines, etc. 3
S How to determine the Air Clearance for Floated Secondary Circuit IEC 60601 - Medical Electrical Equipment Safety Standards Series 3