How to determine the values of k1, k2, k3 in Gage R&R

F

fl350z

#1
hi all.. have smtn to ask u all..

how to determine the values k1, k2,k3?
is there a standard value for them?

how to determine which one to use in a gr&r study? based on number of operators? number of trials?

Thanks..
 
Elsmar Forum Sponsor

antoine.dias

Quite Involved in Discussions
#2
k1,k2,k3

You can take the examples given in the MSA manual.

k1 is the constant in relation to the number of trials in the study.
2 trials 0,8862
3 trials 0,5908

k2 is the constant in relation to the number of appraisers in the study.
2 operators 0,7071
3 operators 0,5231

k3 is the constant in relation to the number of parts in the study.
Parts K3
2 0,7071
3 0,5231
4 0,4467
5 0,4030
6 0,3742
7 0,3534
8 0,3375
9 0,3249
10 0,3146

For further details, See MSA manual.

Best regards,

Antoine
 
C

cplusplus - 2009

#3
Re: k1,k2,k3

Hi,
When I review some of the excel sheets, softwares, and AIAG MSA, i found 2 kind of K1, K2, K3 values.
Eg:
According to several softwares and excel files (minitab, spc for excel) k1 is:

Trials K1
2 4.5600
3 3.0500


According to the MSA handbook (3rd edition, page 114), and some softwares K1 is

Trials K1
2 0,8862
3 0,5908

Do you happen to know the reason of that???

Thank you,
 

antoine.dias

Quite Involved in Discussions
#4
Welcome to the cove.

The difference you have found is due to the different approach in the MSA manual ( difference between version 2 and version 3 of the MSA manual )

Hope this helps,

Antoine
 
A

Alex.G

#5
I'm ok with the answer but what is the difference between the old system and the new? because the constant value are really differente and change considerably the result?

Anyone have an answer to this?
 
M

martin elliott

#6
I'm ok with the answer but what is the difference between the old system and the new? because the constant value are really differente and change considerably the result?

Anyone have an answer to this?
The answer lies in the different equations used in Version 2 and Version 3.

You cannot just simply substitute the Version 3 K values in an Version 2 worksheet.
 
A

Alex.G

#7
I checked the equation and they are exactly the same in the 2 cases?? :confused:
Can u tell me which equation exactly is different ?
:thanx:
 
A

Alex.G

#9
I still don't clearly see where the problem is from, PV or TV?
I review my spreadsheet and I tried to insert the values from the MSA reference manual 3rd edition p 113, with my spreadsheet I get exactly the same value as p114, my excel is wrong or not? did they made a mistake in the book? :confused:

My excel used the old constant values before and I only change it with the new value in 3rd edition, so i didn't change the way to calculate.

below is my excel :
 

Attachments

Y

Yew Jin

#10
I came to this question few years ago and had done some further study on them.......We need to use the Appendix C - d2 table for K constants calculation.

K1 - Used for Repeatability, Equipment Variation (EV) calculation
Find the d2 from Appendix C - d2 table,
focus on the subgroup size(m) for number of trials we used in GR&R study, number of subgroups (g) for parts x number of operators we used in GR&R study. If the parts x number of operator greater then 20, find the d2 at the bottom of the table.
K1 = 1/d2

K2 - Used for Reproducibility, Appraiser Variation (AV) calculation
Find the d2 from Appendix C - d2 table,
focus on the subgroup size(m) for number of operators we used in GR&R study, number of subgroups (g) is set as 1 as there is only one range calculation in this study.
K2 = 1/d2

K3 - Used for Part Variation (PV) calculation
Find the d2 from Appendix C - d2 table,
focus on the subgroup size(m) for number of samples we used in GR&R study, number of subgroups (g) is set as 1 as there is only one range calculation in this study.
K3 = 1/d2

AIAG MSA guideline used formula K = 1/d2, all the sigma (EV, AV and PV) calculated used the original formula with K value replaced and %PT Ratio = (100x 5.15 x sigma R&R)/product specification.

Larry B. Barrentine used formula K = 5.15/d2, all the sigma (EV, AV and PV) calculated used the original formula with K value replaced BUT divided by 5.15 backand %PT Ratio = (100 x 5.15 x sigma R&R)/product specification.

Noted: We do not use the factor 5.15 or 6 in GR&R% calculation (compare to process variation), it is only apply in PT Ratio calculation (compare to product specification):cool:
 
Thread starter Similar threads Forum Replies Date
D Is it possible to determine the Cpk value from Cp and Cr values? Capability, Accuracy and Stability - Processes, Machines, etc. 4
L How to determine / validate Medical Device Storage Conditions ISO 13485:2016 - Medical Device Quality Management Systems 1
John C. Abnet ISO 9001 4.4.1 "...shall determine the processes needed..." ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 72
S Is using ANSI/ASQ Z1.4-2008 the correct sampling plan to determine Pass/Fail of Apparel measurements? AQL - Acceptable Quality Level 4
E How do you determine if hipot is necessary for a cable product? Manufacturing and Related Processes 8
J Standard used to determine if check fixture gauge is acceptable Reliability Analysis - Predictions, Testing and Standards 2
M Measurement Error - How to determine what is acceptable? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
S How to determine & document Organizational Knowledge of a company Document Control Systems, Procedures, Forms and Templates 4
O Monitoring performance - How do I determine performance measurement basis within my organization? Misc. Quality Assurance and Business Systems Related Topics 4
T IEC 60601-1 - Risk assessment to determine the liquid - 11.6.3 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
Ashok sunder How to determine the internal and external issues in ISO 45001 standard? Occupational Health & Safety Management Standards 2
C ICP versus AA: How to determine appropriate specification for assay method verification Qualification and Validation (including 21 CFR Part 11) 1
samer Do we have to determine Educational level for all Staff as pr 7.2 b clause? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
A Calibration of Pressure Gauge, How to Determine Calibration Tolerance General Measurement Device and Calibration Topics 1
S Supplier Qualification - How do we determine which suppliers to qualify? ISO 13485:2016 - Medical Device Quality Management Systems 9
K ISO 13485:2016 Cl. 4.2.3 - Determine QMS Processes, Monitoring, Measuring, etc. ISO 13485:2016 - Medical Device Quality Management Systems 4
A Compliance Obligations - Implementing 6.1.1 and 6.1.3 NOTE - Determine risks and... ISO 14001:2015 Specific Discussions 1
B Auditing Senior Management to determine the Effectiveness of Management Processes IATF 16949 - Automotive Quality Systems Standard 6
R What testing would you run to determine if a device is "Clean"? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
B Maximum Material Condition with Bonus - Determine Cpk for this one kpc Measurement Uncertainty (MU) 2
Q Tissue Adherence Test to determine how Tissue sticks to Gauze Other US Medical Device Regulations 2
C DOE - How to determine optimal settings and transposing data Statistical Analysis Tools, Techniques and SPC 2
insect warfare Cannot determine what type of medical device this product is.... 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
Buckyb Are KPIs a requirement to determine Process Effectiveness? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 14
M Is a Real Time Stability Study required to determine Expiration Date? EU Medical Device Regulations 9
K How to determine rated input power for the markings? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
I How to determine the thread wire size needed - Buttress Thread Gauge General Measurement Device and Calibration Topics 3
M How to determine Process Capability using Minitab 17 Capability, Accuracy and Stability - Processes, Machines, etc. 6
V How does ASQ determine Exam Cutoff Marks Professional Certifications and Degrees 2
R How to Determine Working Voltage & Tolerance for Isolation Barriers? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
apestate How to determine all regulatory and statutory requirements to be met? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
J Gage R&R Study (Crossed) - How to determine Process Tolerance Using Minitab Software 4
D How can I determine a process metrics? Manufacturing and Related Processes 1
D Can I use the % Tolerance to determine the actual process tolerance? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
R How to determine the Thickness tolerance of the Cold Rolled Steel Manufacturing and Related Processes 6
T Technical regulations that determine the acceptable limits of cigarettes General Measurement Device and Calibration Topics 4
J How does Layout Inspection determine Reference Value? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
Q PPAP - How to determine the "Design Record Change Level" APQP and PPAP 2
K 7.6 c) c) Have Identification in order to determine Calibration Status Calibration Frequency (Interval) 3
S How to determine part that is most difficult to sterilize Other Medical Device Related Standards 10
G How to determine if a Standard is Auditable General Auditing Discussions 15
L Who and when to determine AQL for our product inspection AQL - Acceptable Quality Level 1
R 4 Factors 3 Level DOE to determine which factors are significant Using Minitab Software 2
C How to determine a sample size for comparison purposes ? Quality Tools, Improvement and Analysis 2
S How to determine Preventive Maintenance and Calibration Grace Periods General Measurement Device and Calibration Topics 4
J How to determine a Capability Study for Low-Volume Suppliers Capability, Accuracy and Stability - Processes, Machines, etc. 3
C Statistics equation to determine the number of samples in a lot Statistical Analysis Tools, Techniques and SPC 8
S How to determine the Air Clearance for Floated Secondary Circuit IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
somashekar How to Determine the Necessary Employee Competence ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
F ISO 22000 Clause 7.4.4 to determine whether a control is a CCP or and OPRP Food Safety - ISO 22000, HACCP (21 CFR 120) 2

Similar threads

Top Bottom