Hi Big Jim!
First, let me thank you publicly! Sounds as if you've literally written the book on this stuff!
Secondly, I should have explained myself more clearly when I started this ...sorry.
I am 25 yrs. into this business... know all about 5 Whys,
8D, Fishbone diagrams, Fault Tree, pareto,
PFMEA etc., etc.
So, let me re-explain my question/problem...
The Auditor has given us minor nc's for documentation issues only... for example:
NON-CONFORMITY
The process to ensure that devices are correctly classified is not effective.
REQUIREMENT [Annex IX, Section II]
Application of the classification rules shall be governed by the intended purpose of the devices.
If the device is intended to be used in combination with another device, the classification rules shall apply separately to each of the devices. Accessories are classified in their own right separately from the device with which they are used.
OBJECTIVE EVIDENCE
The Generator and associated accessories were not classified in their own right.
Annex IX, Section II
Omission:
.
Review and revise Essential Require :mg:ments (ER) documents for our products currently being sold in the EU and those pending approval for sale in the EU to reflect product and accessory classification (ER-1004; ER-1005; ER-1006; ER-1007; ER-1008).
Review and revise SOP-1010 titled ‘Essential Requirements and Technical Documentation’ procedure to include the requirements of Annex IX, Section II.
Review and revise SOP-1019 titled “Procedure for Compliance with the Regulatory Aspects of the MDD” to include the requirements of Annex IX, Section II.
Establish training record for appropriate personnel for document changes.
9/18/11
Regulatory/Quality
NON-CONFORMITY
The solutions adopted to fulfill the essential requirements are incomplete.
2.1 REQUIREMENT [Annex I, ER 4]
ER 4 states that the characteristics and performances are not affected during the lifetime of the device as indicated by the manufacturer. The product lifetime must be declared along with the rationale for its calculation.
OBJECTIVE EVIDENCE
The product lifetime was not defined, nor was any supporting documentation provided.
Article 3 Annex I
2.1 Contact contract manufacturer, and determine lifetime of RF Generator based on components/materials within the generator. Document lifetime and rationale within ‘product specification’ documents and reference lifetime in applicable ‘essential requirements’ (ER) documents.
9/18/11
Regulatory/Quality
2.2 REQUIREMENT [Annex I, ER 12.1a]
For devices which incorporate software…, the software must be validated according to the state of the art taking into account the principles of development lifecycle, risk management, validation and verification.
OBJECTIVE EVIDENCE
The manufacturer has not referenced any standards or procedures relating to software life-cycle processes.
Article 3 Annex I
Omission:
Did not reference an appropriate state of the art standard relating to software life-cycle processes established during the design process for the RF Generator.
2.2 Review standard ISO 62304:2006. Determine if existing RF Generator documentation requires additional action based on gap analysis. Review and revise as required the product specification PS-1002 titled ‘RF Generator Software Requirement’ specification to indicate reference to ISO 62304. Add reference that includes ISO 62304 to applicable ‘essential requirements’ (ER) documents. Review and revise existing document DOP-1024 titled ‘Software Development Life Cycle’ to indicate requirements of ISO 62304.
Establish training record for appropriate personnel as required.
9/18/11
Regulatory/Quality
The clinical data did not fully meet the requirements of Annex X in the following aspects:
NONCONFORMITY
The clinical evaluation process is inadequate
REQUIREMENT [Annex X, Section 1.1] ‘clinical evaluation’, where appropriate taking account of any relevant harmonized standards, must follow a defined and methodologically sound procedure based on:
Either a critical evaluation of the relevant scientific literature currently available relating to the safety, performance, design characteristics and intended purpose of the device, where: -there is demonstration of equivalence of the device to the device to which the data relates, and – the data adequately demonstrate compliance with the relevant essential requirements…
OBJECTIVE EVIDENCE: The manufacturer has not provided evidence to demonstrate equivalence of the device to the device to which the data relates.
Annex I, 6a Annex X
Omission: Comparative device information was not summarized within established clinical evaluations performed.
Clinical Evaluation has been performed and documented most recently through nContact documents EXD-1171 and EXD-1175. Action is to add a table to one of these evaluations summarizing the design comparison information between nContact devices and competitor’s devices.
9/20/11
Regulatory/Quality
NONCONFORMITY
The user documentation does not fully support the claimed compliance with EN 60601-1-2
REQUIREMENT [EN 60601-1-2:2001 (clause 6.8.3.201)]
For all EQUIPMENT and SYSTEMS, the ACCOMPANYING DOCUMENTS shall include the following information: Table 201, Table 202, the applicable tables 203/205 or 204/206.
OBJECTIVE EVIDENCE
The accompanying documents did not include the tables of guidance and manufacturer’s declaration regarding electromagnetic emissions and immunity.
Annex I, ER2
Omission:
Some of the 60601-1-2 requirements for statements of electromagnetic emissions and immunity were incomplete in the RF Generator’s Operator’s Manual.
Review the requirements of EN60601-1-2:2001, Clause 6.8.3.201. Revise RF Generator Operator’s Manual LBL-1095 as required to include table of guidance and statements regarding electromagnetic emissions and immunity.
9/20/11
R&D and Quality
NONCONFORMITY
The labeling requirements have not been fully met
5.1 REQUIREMENT [Annex I, ER 13.3(a)] The device label must bear the name and address of the manufacturer.
OBJECTIVE EVIDENCE
The RF Generator Label does not include the address of the manufacturer.
5.2 REQUIREMENT [Annex I, ER13.6(p)] The Instructions for Use must contain…(the) degree of accuracy claimed for devices with measuring function.
OBJECTIVE EVIDENCE
The RF Generator IFU does not include the degree of accuracy of the power.
Article 3
Annex I, ER 13
Omission:
A separate label with address was not applied directly to RF Generator.
Omission:
The degree of accuracy stated in the product definition was not included in the IFU.
5.1 Add a label to the RF Generator indicating the nContact address. Revise internal build documents as required to reflect new label.
5.2 Revise the RF Generator Operator’s Manual LBL-1095 to indicate the degree of accuracy of the power output.
9/20/11
Quality and Regulatory
So, this is my Corrective Action Plan...and as you can see, most of the "Root Causes" are "Omission", as in, I Missed It! What do they want me to say or do?? Root Cause... pay more attention?? or Make certain that I have read the latest Rev.? Not always reasonable for a very small company (<30, with all mfg. contracted)