Re: How to handle Out of Tolerance during Calibrations and exceeded Alert Action Limi
Hello Phoebe!!
OK, this is just me kind of talking about the processes and such. Duke and Jim have provided some excellent advice. So this is more food for thought more than anything.
1. Have your acceptance tolerances clearly defined up front, and in writing. So for example, say your adjusment criteria is 2, but the process tolerance which would impact the process is 4, then only initiate a CAPA (if that is your use) for what would impact the process. So, only involve others, when the error found could impact their process. Again, this should be identified up front.
2. If you have an out of tolerance that might affect the process, then enter that information with the CAPA. I would include the magnitude of error, and some previous history on the device. The calibration entity should provide the sufficient information surrounding the instance, so people can make better informed decisions.
Your CAPA should follow a general process:
CORRECTION: The unit was adjusted for maximum accuracy, replaced, taken out of service, sent to mfg. etc.
CORRECTIVE ACTION: Maybe replace the unit, investigate the calibration process to see where the error may have come from during calibration, etc. May be some training issues. Did you identify the cause for the failure? Do you have a solution for the cause?
PREVENTIVE ACTION: Whatever your organization deems to assure this does not happen again. A fair amount of time there is too tight of a tolerance somewhere, or it's a random event (assuming that the calibration process is stable and somewhat mature).
However you decide to assess these occurences (and yes, they will happen), have things identified up-front to minimize knee-jerk reactions and such (and should be a risk-based approach), and complete it; from start to finish. Too... If you find you are having failures on the same instruments, address your corrective/preventive actions, as they are most probably not sufficient.
Oh...
try to avoid taking the easy paths- shortening calibration interval, sending to mfg. for service, or replacing the instrument. They have their place, but so many times are used as Band-aids for a problem, without figuring out what is really going on.