How to measure severity if my product is designed for emergency use and failure would result in death?

silentmonkey

Involved In Discussions
We are generating a design FMEA for a product intended for use in emergency situations, specifically when the patient is suffering from a life-threatening complication. The product is designed to intervene and ultimately save the patient's life.

If a failure were to occur and the device's function is compromised, the product would be unable to save the patient's life and therefore result in death. For the purposes of risk management and producing a sound design FMEA, would each risk item where the device's function is compromised automatically result in a maximum severity rating?

I find this logic a bit excessive as the patient is already dying whether the product is there or not. Am I thinking about this situation in the right way?

Thanks!
 

Ron Rompen

Trusted Information Resource
IMO, yes, if the device fails and the patient dies, then you assign the maximum severity rating. If the equipment HAD functioned as designed, there is a potential (don't know the details of your product so I can't comment on that) that they would have survived.
 

silentmonkey

Involved In Discussions
IMO, yes, if the device fails and the patient dies, then you assign the maximum severity rating. If the equipment HAD functioned as designed, there is a potential (don't know the details of your product so I can't comment on that) that they would have survived.

Thanks Ron! It made sense in my head but at the same time seemed a bit excessive, and it was cause for some debate at the office.
 

monoj mon

Trusted Information Resource
The product is designed to intervene and ultimately save the patient's life.
You have already answered your question by yourself. Your device is expected to prolong the survival time period of the patient and ultimately giving the physicians a fighting chance to revive the patient. That means, failure of any vital function of your device would eventually reduce that "survival time period" and would lead to death of the patient. For example, suppose you are dealing with a device to provide neurovascular embolization to treat the intracranial aneurysms. In such case, that said device is expected to intervene by occluding the blood flow into the aneurysm and thus providing a prolonged survival time period of the patient when the surgeon can provide timely intervention to treat the abnormalities. Failure of the device would cause severe harm to the patient and ultimately may lead to "death", for example,
(a) failure to occlude the blood flow may lead to rupture of the aneurysm causing hemorrhagic stroke and eventually may lead to death
(b) fragmentation of the device (or any part of it) may lead to ischemic stroke and eventually may lead to death etc.
So, in summary, any design failure of this device may lead to death as it is a very high risk device, such as yours.
It made sense in my head but at the same time seemed a bit excessive, and it was cause for some debate at the office.
My point is, while looking at the severity although it seems excessive, we also need to be in the shoes of the patient who would eventually pay the price if our device fails to perform as intended.
 

Tidge

Trusted Information Resource
Add my voice to the chorus: For patients who will suffer severe effects without the intervention of the device, it is presumably acceptable to tolerate higher levels of risk. The actual severity of a harm (e.g. death) is only one component of risk.

Keep in mind that medical device manufacturers are also required to assess the risk to users as well: it may be the case that a line of risk analysis which has death as an outcome for a patient is acceptable, but the death of an operator is far less likely to be acceptable.
 

Ed Panek

QA RA Small Med Dev Company
Leader
Super Moderator
Are there any stats on the problem your device is solving if left untreated? 90% mortality? etc? Or versus the standard of care?
 

silentmonkey

Involved In Discussions
Are there any stats on the problem your device is solving if left untreated? 90% mortality? etc? Or versus the standard of care?
Yes there plenty of stats which we have collected through our review of scientific/clinical literature; this has been included in our CER as well as our Overall Residual Risk Evaluation.
 
Thread starter Similar threads Forum Replies Date
V Uncertainty measure chemistry template Measurement Uncertainty (MU) 1
D How to measure Pitch Diameter of a spline Manufacturing and Related Processes 0
D How to properly measure serrations on hardware Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
S If 4 operate to measure, what is the value of k2?(MSA gage R&R) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
PQ Systems Cpk: Indispensible Index or Misleading Measure Using SQCpack Software 0
V MSA - Using a R&R study for a Tape Measure Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
K 10 parts, 4 operators, 3 replications and each part has 3 location to measure. Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
A How to measure the performance of ISMS? IEC 27001 - Information Security Management Systems (ISMS) 3
M How to measure a plastic cap General Measurement Device and Calibration Topics 19
C First Article Inspections - 'how to measure' General Measurement Device and Calibration Topics 11
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
M Gage R&R and right way to measure Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 16
P How to measure DPPM in Chemical batch manufacturing Quality Tools, Improvement and Analysis 1
A Looking for a third party to calibrate our measure equipment - South Carolina General Measurement Device and Calibration Topics 6
J Validity / outcomes measure for custom made medical device ISO 13485:2016 - Medical Device Quality Management Systems 2
D Performance specification as a Risk Control Measure, EN 14971 ISO 14971 - Medical Device Risk Management 7
J KPIs or Metrics to Measure a New Complaint Handling Process 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
D Torque fractional unit of measure; cN-m or N-cm? General Measurement Device and Calibration Topics 2
qualprod ISO 9001 5.1.1 - How to measure System effectiveness ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
J How to measure a Critical Characteristic that is a Basic Dimension Misc. Quality Assurance and Business Systems Related Topics 6
I Can we use pin gauges to measure an accurate Cpk ? Capability, Accuracy and Stability - Processes, Machines, etc. 9
qualprod Best criteria to measure Corrective Action effectiveness - Poor Maintenance ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
1 How to measure straightness on a HEX Long Bar Manufacturing and Related Processes 7
Ron Rompen Measurement Method Challenge - Measure feature #91 Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
G Measure 3 times, for calibration report? Customer Owned Reference Standard General Measurement Device and Calibration Topics 4
M Informational The ANSM decides, as a precautionary measure, to withdraw from the market macrotextured breast implants and breast implants with polyurethane-coated s Medical Device and FDA Regulations and Standards News 0
M Informational Health Canada advises Allergan of its intent to suspend its licences for Biocell breast implants as a precautionary measure Medical Device and FDA Regulations and Standards News 0
U SI 13 - Not Every Process Requires an Efficiency Measure IATF 16949 - Automotive Quality Systems Standard 1
P NPR Metric - What if Customer has relaxed measure? TL 9000 Telecommunications Standard and QuEST 1
T Tape Measure, the right device for 2 or 3 decimals accuracy on prints General Measurement Device and Calibration Topics 8
M How do we document and measure a service we do not provide? IATF 16949 - Automotive Quality Systems Standard 1
M For a Type 1 Gage R&R I want to use 1 unit and measure it 30 times Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
M How to Measure Paper considering Variation and Humidity General Measurement Device and Calibration Topics 10
rob73 Advice/recommendations for equipment to measure a soft pvc tube - Optical Measuring? General Measurement Device and Calibration Topics 4
R Height gauge to measure short hard to access heights General Measurement Device and Calibration Topics 3
D Proper Tool to Measure Wall Thickness (ISO 3611 - Micrometers) Manufacturing and Related Processes 7
S Quality Measure for Engineering Services Misc. Quality Assurance and Business Systems Related Topics 1
A What are methods and tools to measure Workload analysis ? Human Factors and Ergonomics in Engineering 1
M How do I measure variation in my process ? Capability, Accuracy and Stability - Processes, Machines, etc. 3
J How to measure process quality? Is DPPM the correct metric? Capability, Accuracy and Stability - Processes, Machines, etc. 5
M Micrometer Anvil Placement to measure the diameter of Steel Wires Measurement Uncertainty (MU) 6
S Where to Get SPC Training to Measure our Processes? Training - Internal, External, Online and Distance Learning 12
M A question on Measurement System Analysis - Stability to measure got Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
Jim Wynne Measure Twice, Cut Once: The French Railroad Snafu World News 18
S Measure water activity for fresh raw chicken meat and giblets Food Safety - ISO 22000, HACCP (21 CFR 120) 1
O Metrology Devices to Measure the Basic Dimensions of Plastic Specimens General Measurement Device and Calibration Topics 6
L How to Measure Paint Thickness for an Enclosure Manufacturing and Related Processes 5
C Chamfer Measurements - Do you measure the height and width or the hypotenuse? Manufacturing and Related Processes 15
R Repeatability in MSA - Is repeatability a measure of Location Error Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2

Similar threads

Top Bottom