How to perform Verification for a Gage that is used in Destructive Testing

KnoKsuKaO

Involved In Discussions
#1
Hello all,
please can you help me.
We have gauge that performs destructive tests, such as tensile strength at silentblocks (metal with rubber) and I need some type of verification, because R & R ANOVA, they want majority customers. I think that this is not appropriate for the type of gauge.

Thank you very much for everything.
 
Elsmar Forum Sponsor

Bill McNeese

Involved In Discussions
#3
You may be asking two questions. One is to verify the tensile tester before running a test; the other is to determine the % GRR. In a heat treating facility I work with, we simply verify the hardness tester at the start of the shift by measuring the standard hardness block and plotting the result using an individuals control chart. As long as the X chart is in control and close to the value of the hardness block, the hardness tester can be used. We consider it verified.

The moving range chart can be used to estimate the tensile tester variance and compare that to the process variance to estimate the % of total variance due to the tensile tester. One method of getting at the %GRR. Another method of getting to the % GRR is to run a nested Gage R&R. This can be used when there is destructive testing.
 
Thread starter Similar threads Forum Replies Date
M How to perform Chelex DNA Method Verification in the lab General Measurement Device and Calibration Topics 2
R How do you perform verification of a NCR (Nonconformance)? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
C Do I need to perform design verification Design and Development of Products and Processes 3
J How & how often to perform verification of Test Software of Automatic Test Equipment? General Measurement Device and Calibration Topics 3
M MDR requires the NB to perform "tests in order to check that the quality management system" EU Medical Device Regulations 3
M Define voltage and frequency to perform tests 61010-1 and 61326-1 for CE certification CE Marking (Conformité Européene) / CB Scheme 4
M Informational How to perform a clinical evaluation of medical devices – Part 2 – Level of clinical evidence and what sufficient clinical evidence means Medical Device and FDA Regulations and Standards News 0
L Gage R&R studies for identical Devices - Need to confirm the requirement to perform them Reliability Analysis - Predictions, Testing and Standards 2
M Informational How to perform a clinical evaluation of medical devices – Part 1 – Overview and sample of activities Medical Device and FDA Regulations and Standards News 0
blackholequasar Internal Auditing Inspiration - Getting volunteers to perform internal audits. Internal Auditing 22
E Input to ANOVA 1-WAY - How can I perform this analysis in minitab? Using Minitab Software 2
B AS9102 FAI & Lower Level Drawings - How should we perform the FAI? AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 1
M Perform a coating weight test within 3 hours of processing the parts Manufacturing and Related Processes 2
D How to ensure independence and authority necessary to perform QM tasks ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
Q How to perform Process Capability for true position Statistical Analysis Tools, Techniques and SPC 10
H Do you have to perform an full internal prior to IATF 16949 audit? IATF 16949 - Automotive Quality Systems Standard 21
J Internal Audit clarification - How to perform the audits IATF 16949 - Automotive Quality Systems Standard 6
S When will notified bodies be ready to perform audit to the MDR? EU Medical Device Regulations 1
chris1price Is there a regulatory (MDD, MDR, FDA, ISO, etc) requirement to perform a mock recall? CE Marking (Conformité Européene) / CB Scheme 4
P Can CEO perform internal audit Internal Auditing 11
B How to perform a MSA - GRR (Gage R&R) A&R/ANOVA Vs GUM/VDA Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
W DCMA and AS9100C - Dedicating a week and 5 auditors to perform a AS9100C audit AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 7
W How to perform validation for an epoxy curing station Manufacturing and Related Processes 3
C How to perform Gauge R & R on Automated Test & Measurement Equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 8
E How do you perform a GR&R for system software generated values? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
B Do I need to perform Gage R&R on a Digital Temperature & Humidity Gage Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 18
G Plant Relocation - Is it necessary to perform FAI? AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 13
L Who can perform ISO 9001:2008 Internal Audit prior to our Registration Audit? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
R Registrars offering to perform Second Party Audits - Have they failed their role? Registrars and Notified Bodies 105
V Is it required to perform Process Validation for food or dietary supplements? Food Safety - ISO 22000, HACCP (21 CFR 120) 4
E Should companies that perform calibration services be ISO 17025 accredited? ISO 17025 related Discussions 13
W Hiring an External Auditor to perform our AS9100 Compliance Audits AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 5
E How to perform Discrimination check for Dial Indicator? General Measurement Device and Calibration Topics 2
A Do we need to perform Qualification and Validation? ISO 13485:2016 - Medical Device Quality Management Systems 3
A Does ISO/TS 16949:2009 require us to perform Final/Pre-dispatch Inspection? IATF 16949 - Automotive Quality Systems Standard 5
N Example of how to perform a Visual Attribute Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
J How do you perform a S3 Stability Study for a Non-Replicable Measurement System Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
K Do we need to perform Analytical Method Validation upon Relocation? General Measurement Device and Calibration Topics 5
R Looking for service provider to perform residual dirt test, GME7061 General Measurement Device and Calibration Topics 2
J Is MS Outlook a sufficient reminder to perform yearly supplier evaluations? Quality Manager and Management Related Issues 10
T How to perform an MSA on Cross Hatch Test Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
P Is it necessary to perform Root Cause Analysis for each failed part? Problem Solving, Root Cause Fault and Failure Analysis 16
R Can a Distributor perform Packaging Validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
Geoff Cotton Lego perform space jump Coffee Break and Water Cooler Discussions 0
E How to perform a DCC Audit (Document Control Processes) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
BradM Can Minitab version 16 perform Neural Network Analysis Using Minitab Software 1
M 7.6.3.1 Requirements for Capability to Perform Tests Correctly - Review of Records IATF 16949 - Automotive Quality Systems Standard 10
Sidney Vianna Accreditation Bodies to Perform Market Surveillance Visits to Certified Organizations Registrars and Notified Bodies 2
R Blood Pressure Transducers - Need to perform EMC test? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R Length Out of Specification - How do I perform Cause and Effect Diagram? Nonconformance and Corrective Action 6
Similar threads


















































Top Bottom