How to set equipment calibration/verification criteria
Dongzkie
Hi, need your expertise here.
When can you say that the equipment is need to be scrap or it is totally defective. Just for example I have some non-contact diameter gauge with 0.001mm readability while product tolerance is +/-0.020mm. But during my verification with pin gauge 1 to 5mm the reading was 1.005mm 2.005mm and so on... the error was 0.005mm and I tried to adjust the gauge but it didnt. In your opinion the equipment should be scrap?
thanks in advance.
Ninja
Looking for Reality
The decision for what to do with equipment (use or scrap) should be made considering:
1. What do you need in order to trust the results on an ongoing basis?
vs.
2. What can the machine actually do?
Readability, error, linearity, Manufacturer's claims, whatever.......does the equipment give you what you need to tell "good" from "bad" or not?
In your example, the direct gage error you state uses up 1/8 of your spec range...if you are comfortable with that (and if your customers are comfortable with that), keep it.
If that crosses the line you set for acceptability (or the line your customers drew), repair/adjust/scrap/sell.
L How to set up a basic electronic test equipment calibration laboratory General Measurement Device and Calibration Topics 3
M New Calibration Laboratory Set Up Measurement Parameters and Equipment General Measurement Device and Calibration Topics 4
P Should Equipment Set-Up be a PFMEA Line Item? FMEA and Control Plans 6
E Machine Set-Up Sheets for More Than One Kind of Equipment FMEA and Control Plans 15
R MSA on set of parts with low variation Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 20
A Need to calculate tolerance Intervals with a set of non-normal data and 3-Parameter Weibull distribution Using Minitab Software 0
P Best european location to set up for a virtual medical device manufacturer? EU Medical Device Regulations 4
S Is QMS like a set of rules and regulations that a company follows? ISO 13485:2016 - Medical Device Quality Management Systems 10
W Regarding Requirement for Hi pot & Leakage Current test Station Set-up IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
G CE marking of MD and IVD set EU Medical Device Regulations 3
J How to Set Microbial Limits in Food Supplements Food Safety - ISO 22000, HACCP (21 CFR 120) 3
qualprod To set frequency to review documents in ISO 9001 7.5? Document Control Systems, Procedures, Forms and Templates 13
M IATF 16949 8.5.1.3 Verification of job set-ups - Do we need secondary check? IATF 16949 - Automotive Quality Systems Standard 7
B Set RH spec from dewpoint spec General Measurement Device and Calibration Topics 2
S Is Optical Parallel Set required calibration? ISO 17025 related Discussions 1
R Gauge R&R on pre-set torque wrench Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
A How to set up Continuous CpK monitoring of an injection mold process Reliability Analysis - Predictions, Testing and Standards 7
N Internal Audit Schedule – Who gets to set the schedule? Internal Auditing 16
G Comparing Gauge Blocks - Translating the calibration between the master set and the QA set General Measurement Device and Calibration Topics 2
L How to evaluate the process capability of a data set that is non-normal (cannot be transformed and does not fit any known distribution)? Capability, Accuracy and Stability - Processes, Machines, etc. 12
N When to record values if a spec for a mfg process is set Manufacturing and Related Processes 3
Marc Time Zone - Set your Time Zone Elsmar Xenforo Forum Software Instructions and Help 1
D As a newcomer to QMS, I really appreciate how ISO is set up. ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
C IATF 16949 Cl 8.5.1.3 - Verification of job set up IATF 16949 - Automotive Quality Systems Standard 4
S GRR on ID with set of pin gauges Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
G What to Check - Set Up vs. In Process Inspections Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
N The known set of all root causes Nonconformance and Corrective Action 11
G What degrees of freedom means in uncertainty budget and how do I set the value Measurement Uncertainty (MU) 2
D What is the exact Temperature Coefficient Value to be set on a Conductivity Meter? General Measurement Device and Calibration Topics 2
G Training Employees about Quality - Philosophy / Mind set Misc. Quality Assurance and Business Systems Related Topics 1
renenatasha MDD Classification for Pressure Monitoring Set CE Marking (Conformité Européene) / CB Scheme 1
U Are all manufacturers required to set up eMDR system? Other US Medical Device Regulations 4
S Gage Block Verification or buy a new set (cost aspect) General Measurement Device and Calibration Topics 1
E How to use the Standard Error and how to set the error limits? Reliability Analysis - Predictions, Testing and Standards 2
C Implementing a set of bore gages at a machine Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
S How to set up the Accupoint 2 ATP sanitation monitoring system Quality Assurance and Compliance Software Tools and Solutions 1
L How do I set a limits to cfu in an ISO class 7 Clean Room? ISO 13485:2016 - Medical Device Quality Management Systems 2
4 Quality vs Production on CMM Set Up Requirements Manufacturing and Related Processes 1
A What is the definition of Installation that requires Record Keeping vs Set-up 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
G AS9120 - What should we be expecting once we set forth into becoming certified? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
C ISO 9001:2008 Audit Major Finding - Dial Calipers not set to zero ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
A IEC 60601-2-24 Question: Free Flow Clamp on Set? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C Lowest Acceptable Grade for Master Gage Set Calibration and Metrology Software and Hardware 1
Q Inquire TS 16949 standard - Clause 7.5.1.3 Verification of 'Job Set-Up' IATF 16949 - Automotive Quality Systems Standard 8
M AS9100C Document Control for Machine Set Up Work Instructions AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
J Appropriate In-Process Monitoring Frequency for 2X per day Change Set-Up. Statistical Analysis Tools, Techniques and SPC 3
S Insufflation Tubing & Filter Set Product Classification question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
B How to set Cleanroom Alert and Action Limits as part of Microbiological Evaluation Quality Assurance and Compliance Software Tools and Solutions 8
F How to set-up a Dual Company QMS (Quality Management System) Quality Manager and Management Related Issues 2
R Can the same Biocompatibility Report for Infusion Set be used for its Components? ISO 13485:2016 - Medical Device Quality Management Systems 7