Hi All,
We have a device which is powered from mains 240V and secondary is 24V. Then the secondary 24V is stepped to 2kV thru switched converter. This 2kV required for the patient and it is delivered to patient thru hand held applicator.
From 24V there are other secondary circuits. One among them is control circuitry of applicator.
(not able to insert the image of the device block diagram). I will attach as soon as I find a way.
We can design 2 MOPP for the 2kV working voltage from primary to secondary. Here all secondary is included.
I would like to explore splitting the MOPP option.
Barrier A - Primary to secondary 2MOOP / 1MOPP (whichever is greater) for mains working voltage.
Barrier B - Secondary / HV circuit to applied part (consider full applicator as applied part) - 1 MOPP from secondary at working voltage of 2kV
Barrier C - Secondary / HV circuit to other secondary parts 2 MOOP (or 1 MOOP ??) at working voltage of 2kV.
Here Looking to make a separate module (box) for HV circuit which will be boxed and will meet the HV and CR / CP distances.
Query:
1. Is the above approach acceptable ?
2. If yes, how to test the full isolation at full device level ? Barrier A and Barrier B forms 2 MOPP. Do we need to test 1 MOPP separately ?
If the provided details are not enough, please ask queries.
Thanks,
Chandan N
We have a device which is powered from mains 240V and secondary is 24V. Then the secondary 24V is stepped to 2kV thru switched converter. This 2kV required for the patient and it is delivered to patient thru hand held applicator.
From 24V there are other secondary circuits. One among them is control circuitry of applicator.
(not able to insert the image of the device block diagram). I will attach as soon as I find a way.
We can design 2 MOPP for the 2kV working voltage from primary to secondary. Here all secondary is included.
I would like to explore splitting the MOPP option.
Barrier A - Primary to secondary 2MOOP / 1MOPP (whichever is greater) for mains working voltage.
Barrier B - Secondary / HV circuit to applied part (consider full applicator as applied part) - 1 MOPP from secondary at working voltage of 2kV
Barrier C - Secondary / HV circuit to other secondary parts 2 MOOP (or 1 MOOP ??) at working voltage of 2kV.
Here Looking to make a separate module (box) for HV circuit which will be boxed and will meet the HV and CR / CP distances.
Query:
1. Is the above approach acceptable ?
2. If yes, how to test the full isolation at full device level ? Barrier A and Barrier B forms 2 MOPP. Do we need to test 1 MOPP separately ?
If the provided details are not enough, please ask queries.
Thanks,
Chandan N