How to understand the IEC 60601 50 ms limit?
Roland chung
Trusted Information Resource
#1
Hello folks,
Either 2nd Ed or 3rd Ed, there is a statement saying that transient currents occurring during the first 50 ms following the short circuit shall be disregarded. This means, the currents shall not exceed the allowable value for SFC after 50 ms. It can also imply that the protection (e.g. fuse, regulating network) in place shall operate within 50 ms.
When we conducted a SFC (high voltage shorted to earthed enclosure plus PE interrupted because of the spacing is less than 1 MOP), the protection system took a little more time than 50 ms to operate and the touch current exceeded the limit before the operating. It seems to fail the standard clearly.
However, from the risk analysis, it is a very unlikely case (HV shorted to frame and PE interrupted à one touches the enclosure before the operation of protection system (millisecond field) à hurt or death). I would think the 50 ms limit can be skipped except for the application of heart.
What is your opinion? I would appreciate your thoughts.
Roland
Elsmar Forum Sponsor
Peter Selvey
Leader
Super Moderator
#2
The risk of electric shock, applied to the body externally, depends on the open circuit voltage, available current and time.
So to make a decision you would need to know all three factors.
I have an old Australian standard AS 3859 which has a diagrams which allow you to work out the probability of adverse effects based on these factors.
For example, if the open circuit voltage is high (>100V), the current is 50mA, it looks like about 300ms would be a "safe" limit for body current (not direct to the heart).
But if the current is 500mA, even 10ms would be dangerous.
#3
And IEC 61201, (Touch Voltage threshold values for protection against electric shock) has lots of data relating harms (fibrillation, muscular reaction, burns) to current path, time, current, AC/DC, skin condition (dry/wet) and many more.
It is a Technical Specification standard used to develop safety aspects of other standards.
Thread starter Similar threads Forum Replies Date
R How to understand the exception of IEC 60601 sub-clause 8.11.5 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R How to understand the "full scale peak-to-valley output" mentioned in IEC 60601-2-27 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
L What's the best way to understand IEC 60601-1 standards IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
C Need to understand difference between an actual Calibration vs. a Calibration check General Measurement Device and Calibration Topics 3
J Help to understand and response to API AAR during the re-certification audit Oil and Gas Industry Standards and Regulations 17
lanley liao How to understand and master the requirements of each section in API monogram product. Oil and Gas Industry Standards and Regulations 5
lanley liao How to understand correctly the section of 5.11 MOC in API Spec Q1? Oil and Gas Industry Standards and Regulations 3
lanley liao How to understand this words that the planning of internal audit shall take into consideration the results of previous audits? Oil and Gas Industry Standards and Regulations 10
lanley liao How to correctly understand the bullet list d) of 6.3 Analysis of Data for API Spec Q1 Oil and Gas Industry Standards and Regulations 7
lanley liao How to understand this bulleted list from API 600 Oil and Gas Industry Standards and Regulations 0
A Tips and Tricks to understand ISO 9001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
lanley liao How to understand the clause 6 Planning of ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
lanley liao How to understand the term 3.1.21 'Servicing' and the clause 5.7.1.2 'Servicing' of API Spec Q1? Oil and Gas Industry Standards and Regulations 19
sutie How to understand VDA 6.3 2016 P4.1 VDA Standards - Germany's Automotive Standards 6
Ajit Basrur Need help to understand HIPAA requirements ISO 13485:2016 - Medical Device Quality Management Systems 25
H Embedded Software - I don't understand that Calibration and Metrology Software and Hardware 2
J IATF 16949 Clause 9.3.2.1 - How to understand "measures" IATF 16949 - Automotive Quality Systems Standard 4
J Please help me understand what a process is ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
J Please Recommend books that explain how to understand IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
J How to understand the word "include" in IATF 16949 IATF 16949 - Automotive Quality Systems Standard 2
J How to understand IATF 16949 Clause 8.5.6.1.1 - Process Controls IATF 16949 - Automotive Quality Systems Standard 2
J How to understand IATF 16949 Clause 8.5.2.1 - Traceability IATF 16949 - Automotive Quality Systems Standard 5
S Trying to understand 7.2.1(d) of ISO 13485 - Medical Device User Training ISO 13485:2016 - Medical Device Quality Management Systems 8
M Help me understand how to conduct a Chi Square Test Statistical Analysis Tools, Techniques and SPC 3
S Informational Please help this newbie understand ISO 9001 Work Instructions ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 44
F Help me understand the ANOVA Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
M Kindly help me to understand MSA Acceptance Level Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
T Help me understand how to define Processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 40
N Help me understand the ISO section 7.5.2 from API Q1 perspective Oil and Gas Industry Standards and Regulations 4
A Help me understand the applicability of 21 CFR Part 820.198 (Customer Complaints) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
D Pls hlep me understand the "Total Anaerobic Plate Count" Food Safety - ISO 22000, HACCP (21 CFR 120) 1
A How to understand the source of variability and "stream-to-stream variation" Six Sigma 8
K Training the absolute beginner to understand AQL Training - Internal, External, Online and Distance Learning 1
B I want to understand the ISO certification process. ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
M Reference to understand ISO9001 Book, Video, Blog and Web Site Reviews and Recommendations 1
N Owner of Company Doesn't Understand Corrective Action.... Quality Manager and Management Related Issues 12
L Fourier Analysis for NVH - Reference info to understand the Fourier Analysis method General Measurement Device and Calibration Topics 4
A Help Engineers understand the differences between Bill of Materials and Routers Document Control Systems, Procedures, Forms and Templates 9
G Numbers that will help us understand what the Cost of Rework is Misc. Quality Assurance and Business Systems Related Topics 7
Z How to understand Processes, Procedures and Documented Procedures Misc. Quality Assurance and Business Systems Related Topics 7
G Newbie in IMDS - Please help me understand what I have to do RoHS, REACH, ELV, IMDS and Restricted Substances 4
J Knowledge of CAR/PAR Procedures - Should the entire organization know/understand? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
K General Term for Inspection Process of Sample Parts that everyone can understand Supplier Quality Assurance and other Supplier Issues 10
B ISO 9001:2008 - Can any one help me to understand this paragraph? Clause 8.3 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 14
Z Don't quite understand interpretation of ISO17025: 2005 clause 1.4 and 1.2 ISO 17025 related Discussions 13
J How to understand NOTE 2 in the ISO 9000 definition of defect? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
R Please help us to understand clause 5.5.2 of ISO 9001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
Q An Engineer that does not understand gage R&R? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 17
D Can Someone help me to understand....... Imported Legacy Blogs 3
B Books to Understand FMEA Book, Video, Blog and Web Site Reviews and Recommendations 5
Similar threads
Top Bottom