How to verify profile gages provided by customer
Hi Everyone,
This is my first post to Elsmar Cove.
I am a dimensional metrologist at a plastic injection molding company that makes quite a bit of automotive related plastic parts. Some of these parts are inspected by a production gage that checks the outer profile, or silhouette, all the way around the part. These "drop gages" are essentially go, no-go gages that verify that the parts haven't exceeded a maximum or least material boundary specified by a profile tolerance on the part drawing. Please review the inserted images. As you can see, these gages were machined out of aluminum. A quality inspector would simply attempt to place the part in each of the machined aluminum pockets. The part is required to fit into the larger oversized pockets, and resist fitting into the smaller undersized ones.
A large number of gages of this style were created and provided to us by our customer and I have been tasked with inspecting them. Unfortunately, there is no CAD information available that describes the shape of the various profiles that were used to CNC these shapes. Hence, there is no reference standard by which I can verify these gages. What is our company's responsibility in this context? I am not particularly well versed in ISO standards, but I have been told by my colleagues that we must inspect these gages to remain ISO compliant. However, without our client's assistance, I don't see what we could meaningfully check on these items. Would a simple periodic visual inspection for dents, nicks, etc suffice?
dwperron
From what I understand, you have numbers of these gauges that are being used in go-nogo testing of products.
You are also mentioning "ISO" requirements, and you are wondering what you need to do to comply with these.
First, there are many ISO standards that might apply. Since you did not state a particular ISO standards we'll look at the granddaddy of them all, ISO 9001 This is the section of ISO 9001 that most applies to your situation:
7.1.5.1 General
The organization shall determine and provide the resources needed to ensure valid and reliable results
when monitoring or measuring is used to verify the conformity of products and services to requirements.
The organization shall ensure that the resources provided:
a) are suitable for the specific type of monitoring and measurement activities being undertaken;
b) are maintained to ensure their continuing fitness for their purpose.
The organization shall retain appropriate documented information as evidence of fitness for purpose of
the monitoring and measurement resources.
Your case is using these gauges to measure parts to see if they are within acceptable size boundaries, so this applies.
a) states that these gauges must be suitable to the purpose, the company that owns them has determined that they are.
Your roll falls under b). You are trying to determine whether or not they are continuing to be fit for the job.
This is where you need guidance from your customer. They need to determine what makes this gauge fit for their purpose.
If it is dimensional accuracy, then they will need to provide critical measurement tolerances.
If it is only a visual inspection to determine fitness for use (dents, nicks, burrs, etc.) they will still need to tell you what is acceptable and what is not acceptable so you can make an independent determination of fitness.
Your customer is concerned about the last line of the quoted section above. They must retain your results that determined fitness for use.
At this point there is nothing you can do with these on your own. If your customer wants you to determine whether or not these are fit for use then they must give you the accept / reject parameters for your inspections.
Thank you for your response, dwperron!
If the customer is un-willing to provide guidence, how can I protect myself from an ISO non-conformance regarding this matter? We are a Tier 2 supplying a Tier 1 automotive supplier. My thought was to put something in writing stating that the customer is responsible for inspecting these particular gages and maintaining the associated records. Is that a reasonable course of action?
dwperron
You are not going to receive a non-conformance unless you "calibrate" these without guidance from the customer.
I would return them to the customer without calibration until they provide you with what they are expecting for results, and the pass / fail parameters for these. That is how you protect your organization.
If you are attempting to comply with ISO/TS 16949 you have additional issues. As an external lab you would need to be either ISO 17025 accredited for these calibrations, or you will require documentation from your customer that you are acceptable to them.
Then you will have a lot more specific requirements in your reporting of results. You would best look at ISO/TS 16949 section 7.1.5 before you proceed.
Ninja
Looking for Reality
If you don't have another measure of what the gages are supposed to be, you really don't have a choice but to have your customer vet them as fit for use. Do you have direct contact for the customer SQE? See what they think and how they suggest to proceed...if nothing else it may give you access to the drawings so you can have the gages verified....or even easier, an email saying "They are good, keep using those".
GRP
Your situation clearly would benefit from some communication and clarification with your client.
However, in terms of what would be meaningful to check, more than inspecting them I would ask: do the gages accept good parts and reject bad parts? In the automotive industry you would answer that with an MSA attribute gauge study. I would opt for signal detection.
C I need to verify the surface profile of a part - Sine plates General Measurement Device and Calibration Topics 4
E Layouts : Can you "verify" things that are not visible (such as loctite present, internal dimensions) in a layout ? APQP and PPAP 2
K How to verify the Storage and Transport environmental condition for Medical Device Miscellaneous Environmental Standards and EMS Related Discussions 3
J How to verify the actual capacity of suppliers Supplier Quality Assurance and other Supplier Issues 3
V Regrinding Material - Verify whether a product contains any regrinding material Manufacturing and Related Processes 13
Q Using Caliper A to verify Caliper B - Verification of Calibration ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
V Tie-wrap pull test to verify contact retention Manufacturing and Related Processes 6
D How to Calibrate and Verify a Heating Oven General Measurement Device and Calibration Topics 3
A How to Verify Supplier I-9 Forms for ITAR Requirements Other ISO and International Standards and European Regulations 2
O Creating a Tool to Track & Verify Mistake Proofing Devices Document Control Systems, Procedures, Forms and Templates 5
N What is the easiest way to verify we have the correct customer supplied materials? Misc. Quality Assurance and Business Systems Related Topics 5
R IEC 62304 - 5.2.9 Verify Software Requirement IEC 62304 - Medical Device Software Life Cycle Processes 5
B Verify GB/T 16886 is the same as ISO 10993 US Food and Drug Administration (FDA) 5
A How to verify and validate the effectiveness of CAPA with regards to employees ? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
R IEC 62304 Section 5.3.6 Verify Software Architecture IEC 62304 - Medical Device Software Life Cycle Processes 4
R LKV Value: Can not verify Reliasoft Case Study Reliability Analysis - Predictions, Testing and Standards 2
S Verify Software System Testing: Verify that results meet the Pass/Fail Criteria IEC 62304 - Medical Device Software Life Cycle Processes 1
L Gauge to test and verify a Set Square Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
D What to verify on Test Setup for Environmental Testing General Measurement Device and Calibration Topics 7
A When to verify and when to validate Design Changes 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
D Since NIST # can not be used to verify Traceability - What an Auditor should do AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 8
W How to verify/calibrate a Neptune Water Meter General Measurement Device and Calibration Topics 4
optomist1 CMM Shortcomings - GD&T - Inability of CMMs to verify the MMC situation Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
C How to Calibrate and Verify Torque Multiplier Gages General Measurement Device and Calibration Topics 6
J How do I continually Verify Compliance to my Processes? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
R A way to verify 7075 T0 Aluminum has been successfully Heat Treated to T6 Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
Ron Rompen Recommended Rockwell 30T Scale Hardness Tester to verify Incoming Product General Measurement Device and Calibration Topics 1
D How to verify a process that uses software to accomplish certain tasks? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
X Cannot Verify until Customer uses Product - Clause 7.5.2 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
C Surface Plate Calibration - How to verify the flatness of a granite surface plate General Measurement Device and Calibration Topics 1
E How Do You VERIFY Customer Property? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
R Verify Accuracy for Validation of CNC Starrett Vision System Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
L Verify, Inc. Source Inspection experience? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
H Definition Inspect, Review, Verify and Witness - What are the definitions and differences? Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 5
B How to inspect Mylars to verify dimensions, grid lines and grid points Inspection, Prints (Drawings), Testing, Sampling and Related Topics 17
D How do I verify the Cold Room temperature in Restaurant Kitchen Food Safety - ISO 22000, HACCP (21 CFR 120) 7
C Bridgeport Mill as a Gage - Verify gage dimensions using a Bridgeport Mill Measurement Uncertainty (MU) 4
A Test Method - DIN ISO 1133 for melt flow needed to verify supplier results Manufacturing and Related Processes 1
F How would you Verify Repeatability and Reproducibility of Measurement Systems? Quality Tools, Improvement and Analysis 5
S How to Calibrate or Verify Screw Pitch Gages General Measurement Device and Calibration Topics 16
J Work Instruction to Calibrate or Verify a Combination Square General Measurement Device and Calibration Topics 15
Howard Lee Hygrometer Calibration - Best way to verify accuracy of a 'reference' Hygrometer? General Measurement Device and Calibration Topics 11
C How to verify the SAL (sterilie acceptable level) is 10-6 for ETO Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 6
K Could someone verify the levels of quality management system standards Misc. Quality Assurance and Business Systems Related Topics 2
K Prohibited metal verification - I am not sure of a PHYSICAL way to verify this Various Other Specifications, Standards, and related Requirements 9
P CMM Programming verification - How would you verify that a CMM program is correct? Quality Tools, Improvement and Analysis 5
Douglas E. Purdy To Verify Concentricity of Flats to a Centerline General Measurement Device and Calibration Topics 9
Adriane Need Help to Verify an Infrared Cure Oven General Measurement Device and Calibration Topics 9
K Verify paper print to CAD file so we can use the CAD models for inspection purposes Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
T How to verify SOP training effectivness - Section 6.2.2 in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 7