How to verify SOP training effectivness - Section 6.2.2 in ISO 13485
TomRadiak
:bigwave:
In section 6.2.2 in ISO13485 it is stated that training (of a SOP) has to be evaluated for effectivness.
- Does this mean that emplyees has to be tested (confirmed by a written test) for every SOP they read?
Best regards from a sunny Stockholm,
6.2.2 c) requires you to "evaluate the effectiveness of the actions taken" (from 6.2.2b).
The question is: "How do you know (when you train on an SOP) that the trainee "got" the message?" A written test may be an indicator, but how many people do you know that are good with tests, but lousy with performance. Remember one of the major changes in 13485 is the change from "qualified" to "competent". A written test probably will not meet the competent requirement.
The answer lies in how you determine that the trainee is competent to perform to the SOP. Probably the best way is to monitor performance. If the supervisor is satisfied that the new performance is acceptable, then you've met 6.2.2 c).
The next logical question would be concerning records of the above. 6.2.2 e) requires you to "maintain appropriate records". We've battled this greatly in the ISO 9K threads, however, my belief is that it is up to you (your customer, or national agency) to determine what type of records are to be kept.
Hope that helps.
Oh, by the way, I just noticed... Welcome to the Cove :bigwave:
K
Kevin H
Training effectiveness
Hi Tom,
I'm not working with 13485, but ISO 9001:2000 & TS 16949, but the requirements are the same. Several ways in the US that are used for training effectiveness include results of a test after training and a sign-off sheet maintained by a supervisor indicating that the employee was observed performing the job/task after training and that they performed it at an acceptable level. For professional employees, such as engineers, a typical way to verify competency is to verify it in an annual review.
For training by an independent organization, the certificate issued by the organization is normally recognized as verifying competency. A lot of these training organizations are now using tests at the end of their training to verify competency before they issue a certificate.
Hope this helps.
Wes Bucey
Is it effective?
There are two factors to your question:
  1. Is the training effective? (Can the trainee perform the SOP correctly and consistently? Is the trainee "competent" in that task?)
  2. Is the fact of the determination of competency recorded?
The organization is the sole arbiter of what constitutes effectiveness of the training, whether it is a formal written test, a standard "practical" test, or a supervisor's judgment.
Once the competency is determined, the organization should record the fact. Where the data is recorded is immaterial for the purpose of the Standard. Some organizations keep files of each SOP, for example, with a list of everyone who has qualified as "competent." Others stick the competency data in an employee's personnel file.
An added feature that I consider "good practice" is periodic re-evaluation of the competency of the employee to perform the task (results of re-evaluation ALSO recorded.)
Note it is also good practice to re-evaluate the PROCESS of determining competency to determine if it continues to meet the organization's needs.
Bwana.Jo
Important thing in "Training" requirement !
Hi, folks
I'd like to introduce my point on 'Training Requirement' between FDA QSR & ISO 13485.
ISO 13485, 6.2.2
a) Competence <-- Job Specification --> (goes to make) Job Description
b) Provide training needs, <-- Identity what needs to be trained
c) Evaluate*
d) Concentrating on awareness --> see FDA QSR 820.25 (1)
e) Records --> clue for evaluate above 'd)' or effectiveness of 'c)'
* Evaluate: refer to HR formulations.
Sc = Sa / Sb Skill
Kc = Ka / Kb Knowledge
Pc = Pa / Pb Performance
a= after training, b= before training, c= change
FDA QSR 820.25 describes
(a) General. => Determine what is necessity
(b) Training. => Identify what should be trained
(1) Personnel shall be made aware of device defects which may occur from ..their..job. => Defect device training (kind of foreseeable training; screening by operators)
(2) Personnel shall be made aware of defects and errors that may be encountered as..their job.. => Preventive training (aspect of "Risk" view to 'Reduce' training for engineerings)
Comment me ! :bigwave:
Alex Jo
liuyy
short term and long term effectives
you must evaluate the short term effectives and long term eddectives of what actions you have taken.
rtstewart1
Evaluating the training and/or the trainer
In addition to ensuring that the person being trained was competent, you should also verify that training and/or the trainer is doing an effective job. At the company that I previously worked for, we had a questionnaire on the back of our training record form and the person being trained was required to provide feedback on the training. The trainer was then required to review the comments and incorporate the feedback into future training sessions. We found it to be quite helpful in obtaining objective input on the effectiveness of the training that was being provided.
Bwana.Jo
I love your inputs.. thanks !
Thanks for your inputs for considering effectiveness and measurable requisitions.
Our company doing now two kind of training certification.
One is simple training certification as general thing.
The other is for 'special operating trainee' like a special process operator/technician. They will be certified by "Certification Check Off (CCO)" for certain trained work instruction, FYR.
Measure is all the time headache for me cause that function should not be subjective. But our CCO is also nonobjective for qualifying operator, unfortunately.
Any helping concept for me... will be highly appreciated.
Thanks...Alex
qualprod D5 of 8D clarification, how to verify root cause Problem Solving, Root Cause Fault and Failure Analysis 24
C Where to draw the line for "sufficient evidence" to verify safety/performance of a device? CE Marking (Conformité Européene) / CB Scheme 2
K Verify Software Architecture - supporting interfaces between items IEC 62304 - Medical Device Software Life Cycle Processes 2
M How to verify profile gages provided by customer General Measurement Device and Calibration Topics 5
E Layouts : Can you "verify" things that are not visible (such as loctite present, internal dimensions) in a layout ? APQP and PPAP 2
K How to verify the Storage and Transport environmental condition for Medical Device Miscellaneous Environmental Standards and EMS Related Discussions 3
J How to verify the actual capacity of suppliers Supplier Quality Assurance and other Supplier Issues 3
V Regrinding Material - Verify whether a product contains any regrinding material Manufacturing and Related Processes 13
Q Using Caliper A to verify Caliper B - Verification of Calibration ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
V Tie-wrap pull test to verify contact retention Manufacturing and Related Processes 6
D How to Calibrate and Verify a Heating Oven General Measurement Device and Calibration Topics 3
A How to Verify Supplier I-9 Forms for ITAR Requirements Other ISO and International Standards and European Regulations 2
O Creating a Tool to Track & Verify Mistake Proofing Devices Document Control Systems, Procedures, Forms and Templates 5
N What is the easiest way to verify we have the correct customer supplied materials? Misc. Quality Assurance and Business Systems Related Topics 5
R IEC 62304 - 5.2.9 Verify Software Requirement IEC 62304 - Medical Device Software Life Cycle Processes 5
B Verify GB/T 16886 is the same as ISO 10993 US Food and Drug Administration (FDA) 5
A How to verify and validate the effectiveness of CAPA with regards to employees ? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
R IEC 62304 Section 5.3.6 Verify Software Architecture IEC 62304 - Medical Device Software Life Cycle Processes 4
R LKV Value: Can not verify Reliasoft Case Study Reliability Analysis - Predictions, Testing and Standards 2
S Verify Software System Testing: Verify that results meet the Pass/Fail Criteria IEC 62304 - Medical Device Software Life Cycle Processes 1
L Gauge to test and verify a Set Square Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
D What to verify on Test Setup for Environmental Testing General Measurement Device and Calibration Topics 7
A When to verify and when to validate Design Changes 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
D Since NIST # can not be used to verify Traceability - What an Auditor should do AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
C I need to verify the surface profile of a part - Sine plates General Measurement Device and Calibration Topics 4
W How to verify/calibrate a Neptune Water Meter General Measurement Device and Calibration Topics 4
optomist1 CMM Shortcomings - GD&T - Inability of CMMs to verify the MMC situation Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
C How to Calibrate and Verify Torque Multiplier Gages General Measurement Device and Calibration Topics 6
J How do I continually Verify Compliance to my Processes? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
R A way to verify 7075 T0 Aluminum has been successfully Heat Treated to T6 Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
Ron Rompen Recommended Rockwell 30T Scale Hardness Tester to verify Incoming Product General Measurement Device and Calibration Topics 1
D How to verify a process that uses software to accomplish certain tasks? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
X Cannot Verify until Customer uses Product - Clause 7.5.2 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
C Surface Plate Calibration - How to verify the flatness of a granite surface plate General Measurement Device and Calibration Topics 1
E How Do You VERIFY Customer Property? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
R Verify Accuracy for Validation of CNC Starrett Vision System Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
L Verify, Inc. Source Inspection experience? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
H Definition Inspect, Review, Verify and Witness - What are the definitions and differences? Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 5
B How to inspect Mylars to verify dimensions, grid lines and grid points Inspection, Prints (Drawings), Testing, Sampling and Related Topics 17
D How do I verify the Cold Room temperature in Restaurant Kitchen Food Safety - ISO 22000, HACCP (21 CFR 120) 7
C Bridgeport Mill as a Gage - Verify gage dimensions using a Bridgeport Mill Measurement Uncertainty (MU) 4
A Test Method - DIN ISO 1133 for melt flow needed to verify supplier results Manufacturing and Related Processes 1
F How would you Verify Repeatability and Reproducibility of Measurement Systems? Quality Tools, Improvement and Analysis 5
S How to Calibrate or Verify Screw Pitch Gages General Measurement Device and Calibration Topics 16
J Work Instruction to Calibrate or Verify a Combination Square General Measurement Device and Calibration Topics 15
H Hygrometer Calibration - Best way to verify accuracy of a 'reference' Hygrometer? General Measurement Device and Calibration Topics 11
C How to verify the SAL (sterilie acceptable level) is 10-6 for ETO Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 6
K Could someone verify the levels of quality management system standards Misc. Quality Assurance and Business Systems Related Topics 2
K Prohibited metal verification - I am not sure of a PHYSICAL way to verify this Various Other Specifications, Standards, and related Requirements 9
P CMM Programming verification - How would you verify that a CMM program is correct? Quality Tools, Improvement and Analysis 5