How would you address a Validation Gap/Escape (Product is in the Field)

Q

QE

Hello All

I am presently working on a issue/exception where we are finding out that the validation was not sufficient and an important requirement was missed. now there is product in the field.

Can some one advise what can be done to resolve this issue ?
 

Bev D

Heretical Statistician
Leader
Super Moderator
not sure exactly what you might be looking for.
if the problem is severe, the product may need recall
you may need to revise the design
you may need to put in point of service rework
you definitely should address the business process gap in your validation process.

aer any of thses suggestions appropriate?
 
M

MIREGMGR

Retroactive validation is sometimes an acceptable process...particularly in the case of legacy products that were originally documented under lesser requirements than we would understand now to be appropriate, but I've also heard of it being done in an instance such as yours.

You could perform a retroactive validation ASAP, both to resolve the documentation gap and also as a mechanism for establishing that the product hasn't been dangerous during the time between first marketing and the completion of that retroactive validation.

Properly, you should halt the use of the product while that retroactive validation is done. That of course might not be practical unless the number of units in the field is relatively small and you have contact information for all of them.

If the number of distributed units is large and/or you otherwise don't have effective control, and/or you don't have the ability to do the rretroactive validation in a hurry, you might not have any (proper) choice other than to call your FDA regional office's Recall Coordinator and discuss the situation with them.

In any case, I'd develop that validation plan as quickly as possible and get going. The FDA might regard a combination of forthrightness and urgency in fixing the gap as sufficient to justify waiting on a field action, if you can convince them that the risk is low.
 
Thread starter Similar threads Forum Replies Date
K Validation of Processes - How do I address TS 16949 Sec. 7.5.2.1 IATF 16949 - Automotive Quality Systems Standard 30
E Change in manufacturer address US Medical Device Regulations 2
Richard Regalado 6.1 Actions to address risks and opportunities to the ISMS IEC 27001 - Information Security Management Systems (ISMS) 0
dgrainger Informational Proposal to address MDR challenges EU Medical Device Regulations 23
D How to address the content deviation of 'cannot apply criteria of risk acceptability prior to...' ISO 14971 - Medical Device Risk Management 1
F Change in address for critical supplier - does NB need to be informed? CE Marking (Conformité Européene) / CB Scheme 20
K Change in address on ERLM US FDA: Implications to import US Medical Device Regulations 0
Ed Panek How to address a QMS Entry Error ISO 13485:2016 - Medical Device Quality Management Systems 9
P Update on NBOG 2014-3 to address MDR/IVDR, or any plan to do that? EU Medical Device Regulations 0
M Address change for a company with CE/ISO13485 EU Medical Device Regulations 2
J FDA notification of address change US Food and Drug Administration (FDA) 2
R Where does IATF 16949 address Process mapping? IATF 16949 - Automotive Quality Systems Standard 3
N Address of Legal Manufacturing Site in ISO 13485 certificates? ISO 13485:2016 - Medical Device Quality Management Systems 1
Ajit Basrur Physical Address on Product Labels (EU MDR perspective) EU Medical Device Regulations 2
A How to address Environmental & OHS Opportunities? ISO 14001:2015 Specific Discussions 6
Q Legal Manufacturing Address Change – multiple registrations at same address 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
K Changing the Address in 510(K) Approved Product 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
E Received a Major finding during IATF Surveillance audit for loss of BIQS Level 3 (more than 6 SPPS in 6 months)...how should we address SYSTEMIC CA? IATF 16949 - Automotive Quality Systems Standard 11
M Manufacturer address on the label Other US Medical Device Regulations 18
M GDPR - Is anonymizing sufficient to address right to erasure? Medical Information Technology, Medical Software and Health Informatics 3
K Implant Card - Full Manufacturer's Address - Space issues EU Medical Device Regulations 4
I Importer Address in addition to Manufacturer details on label EU Medical Device Regulations 13
S Medical Device Initial Importer (DII) Requirements - Delivery Address 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
B Manufacturing Company Address Change - implications for FDA Other US Medical Device Regulations 1
M Informational USFDA – Efforts to evaluate materials in medical devices to address potential safety questions Medical Device and FDA Regulations and Standards News 0
M Putting AR's name and address on label - Authorized Representative Labeling Requirements CE Marking (Conformité Européene) / CB Scheme 0
S Risk Approach doesn't address External Issues (Auditor's Comment) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 30
M Medical Device News FDA's Policies and Procedures Should Better Address Postmarket Cybersecurity Risk to Medical Devices Other US Medical Device Regulations 0
Marc Email Addresses - Changing your email address Elsmar Xenforo Forum Software Instructions and Help 0
B ISO 17025 8.5 Actions to address risks and opportunities ISO 17025 related Discussions 7
TheMightyWife Address Change - Medical Device Packaging Labelling Requirements EU Medical Device Regulations 7
M "Minor" Address Change Implications on Product Labeling Requirements Misc. Quality Assurance and Business Systems Related Topics 2
K Labeling - New company address Canada Medical Device Regulations 6
S How to address and implement OH&S opportunities in ISO 45001:2018? Occupational Health & Safety Management Standards 12
qualprod ISO 9001:2015 - Right method to address clause 6.1? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 14
Q ISO 14001:2015 Clause 6.1: Actions to address Risks & Opportunities ISO 14001:2015 Specific Discussions 2
J How to address IATF 16949 clause 5.1.1.1 in my Quality Manual IATF 16949 - Automotive Quality Systems Standard 6
Q ISO 9001:2015 - Clarification in 6.1.2 Note 1 (Options to Address Risks) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
P Requirements/guidance on what constitutes the Manufacturers "Address" EU Medical Device Regulations 7
M FDA Email Address needed for CECATS 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
R Actions To Address Risks and Opportunities IATF 16949 - Automotive Quality Systems Standard 1
V 21 CFR 820 Compliant way to use Kaizen to address Nonconformances Nonconformance and Corrective Action 9
R Is PFMEA (Process FMEA) for OEM sufficient to address Risk? ISO 13485:2016 - Medical Device Quality Management Systems 7
S Change of Medical Device Company Registered Address - Implications Quality Manager and Management Related Issues 3
R How to address unresolved Medical Device Software Anomalies? IEC 62304 - Medical Device Software Life Cycle Processes 2
B Different Label Address Required for EU? EU Medical Device Regulations 7
M How to communicate the change in company address to CFDA ? China Medical Device Regulations 1
C Appropriate ways to address Competency? Generic vs Specific ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
K Customer Address on ISO 17025 Test Report - Is email address okay? ISO 17025 related Discussions 9
R How to address an audit of another facility that is not ISO/TS 16949 ? IATF 16949 - Automotive Quality Systems Standard 9

Similar threads

Top Bottom