Humidity Pre-Conditioning before leakage current and dielectric strength tests

Pads38

Moderator
Our (small) company does our own 60601-1 testing in-house. For the last couple of years our Notified Body has asked us to have our Test Report expert reviewed. This we have done by the test house arm of the NB. (It costs - but much less than an external test).

Up until now, when it comes to the requirement for humidity pre-conditioning (before leakage current and dielectric strength tests) I have marked it as N/A, stating that "there are no hygroscopic parts; no natural rubbers or sintered insulations and it is not for use in atmospheres where condensation may occur".
And this has been accepted.

However, our expert reviewer has now decided that that is not a strong enough argument. Has anyone got any useful "extra ammunition" in this argument?

He does seem to suggest a possibilty of using 'robust risk management' to justify the N/A status, but any suggestions would be most welcome.
 
N

NickM

There are already a few threads on this. If you search on humidity you will find them.
 
Last edited by a moderator:

Peter Selvey

Leader
Super Moderator
There are a few interesting issues here ...

First on the direct subject: For a long time I have used the rationale of "non-hygroscopic materials" to skip the RH test, without any problems, and other agencies do as well. The key point is that the dielectric strength test is a test of solid insulating materials (not along the surface, but through the insulation). As long as the materials (wiring, capacitors, optos, transformer insulation etc) are well established the humidity test is not important. The test is important for new or novel materials, impregnated paper and for powdered insulation such as sealed heaters. Some agencies assume the dielectric strength test is intended for the finished product, but this is a mistake, it is a material test.

While the humidity test may also be similarly skipped for leakage current, it may be important to check the leakage current when the equipment is cold. Materials, and in particular EMC capacitors can have a high temperature co-efficient. Changes in leakage of 10% between hot/cold conditions are not unusual, with the maximum often being in the cold condition.

On a technical point, I would use a verdict of "Pass" rather than N/A. The requirement is applicable, it is just the test itself has been waived. This would be the same as, for example, a stability test on a device shaped like a old VCR - the center of gravity is so low that the test is a waste of time, reasonably waived, but the verdict is still Pass.

On the subject of the "expert review" - I think this is not a bad idea, but some care is needed to define what it really means. Does the reviewer's report form part of the technical file? If so, does the reviewer take responsibility for the results in the report(s)? Or does the review form part of the audit (and hence end up in the audit report)?

Mostly likely the review is just a side service using an "all care, no responsibility" model, which is why it is cheaper. If so, the auditors should ignore this report, they should only refer to the report that is in the technical file.

It means you can take the expert review as feedback, but you don't have to follow all of their recommendations. The person that signs the report that ends up in the technical file has the final say.

If the review is part of the audit (which it could be, but there are procedural requirements the NB must follow to do this), then in effect the auditor is challenging your engineering judgement. This is also fine, but the rules state that the auditor must produce "objective evidence" for a non-conformity. In other words, if you have followed the procedures correctly, and written up a plausible rationale for waiving the test, then they need the ammunition, not you. "Objective evidence" is more than a feeling that it "... is not a strong enough argument", particular in a case like this where waiving a test is common practice.

:2cents:
 

Pads38

Moderator
Thanks for your reply Peter.

The reason for the report review is that the NB auditors have decided that they, as primarily quality system people, do not have the technical knowledge to pass judgement upon the adequacy of our type test reports. Instead, they ask an engineer from their test laboratory to review and make the decision.

As for what goes in the Tech File - at the moment we are limited to just adding a print out of the e-mails. I have wondered whether we should get the test lab. engineer to countersign the report. Although that has logistical challenges as the lab is in another country.

As for the need to do the conditioning - It is difficult to argue against the Test Lab's position as they state that they never waive the test. And they keep on bringing up condensation as being an issue, even though the test is clearly designed so that condensation cannot occur.

I expect that we will be stuck with doing some testing just to prove our point, and fortunately, I have identified a couple of local places that have the equipment to be able to do this. When we have proven that it makes no difference we may have more success with our arguments, provided we don't introduce any "unusual materials / components.

I take your point about marking it as "Pass" rather than "N/A". Perhaps with my own testing in place that will make the most sense, such that only other contrary "objective evidence" could force us to repeat the tests for other devices.
 

Kevin Mader

One of THE Original Covers!
Leader
Admin
Nice job Peter. I agree it's a Pass using mitigating rationale. I'd say that would settle a number of auditors who might see N/A as objectionable.

And any of us should require objective evidence for any allegation made by an auditor or technical reviewer. Gut feel is never enough. "In God we trust...all others bring data!" WED
 

swathi.sp

Involved In Discussions
Dear folks,

We are in to radiotherapy device manufacturing and as part of mains isolation, we have been using Isolation Transformer at mains input and also it was tested as part of end system as per IEC 60601-1, 3.1 ed. All the electrical safety testing like Dielectric strength , Leakage Current , Transformer Overload and short circuit testing after Humidity Preconditioning treatment have been performed. Now we are in plan to change the transformer rating, however dielectric strength remains the same. In this scenario, do we need to perform all the electrical safety testing related to Isolation Transformer from the test lab? Or can we test internally. If in that case, is that mandate to get tranformer subject to humidity preconditioning.

Can any one , please throw some light on this.
 

Peter Selvey

Leader
Super Moderator
There's no strict rule for a design change (such as this), it's just engineering judgement. In principle, if a test is not done, you need to find a qualified person willing to sign their life away (= take responsibility) and say that the tests on the old design are representative the new design. So it's not a case of a mandatory rule to test or not, but rather finding the person who will be held responsible and tying them down to a decision.

Personally, I would expect to be OK with a spot check dielectric and skip the humidity preconditioning, but I would want to formally review the detailed documents such as drawings, specifications, material list, description of changes etc (which is not what this forum is for ;)).

Another engineer might say better just to do all the tests again. Which is fine - it just depends on what the individual is willing to take responsibility for.
 

Benjamin Weber

Trusted Information Resource
Dear folks,

We are in to radiotherapy device manufacturing and as part of mains isolation, we have been using Isolation Transformer at mains input and also it was tested as part of end system as per IEC 60601-1, 3.1 ed. All the electrical safety testing like Dielectric strength , Leakage Current , Transformer Overload and short circuit testing after Humidity Preconditioning treatment have been performed. Now we are in plan to change the transformer rating, however dielectric strength remains the same. In this scenario, do we need to perform all the electrical safety testing related to Isolation Transformer from the test lab? Or can we test internally. If in that case, is that mandate to get tranformer subject to humidity preconditioning.

Can any one , please throw some light on this.

I as a tester highly recommend to retest the transformer incl. humidity preconditioning. But in my opinion it is sufficient to test the transformer separately, not the whole system.

In the unlikely case of an incident, you don't want to explain yourself based on a pure engineering judgement.
 
Thread starter Similar threads Forum Replies Date
E Humidity Pre-Conditioning IEC 60601-1 (2005) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
S Why does GAGEpack drop the temperature and humidity readings out of the verification event card? Quality Assurance and Compliance Software Tools and Solutions 0
CaliperJim Temperature and Humidity Effects on Measurement Devices General Measurement Device and Calibration Topics 6
J Warehouse temperature and humidity control EU Medical Device Regulations 7
C One Time Service Supplier - Temperature and Humidity Testing Service ISO 13485:2016 - Medical Device Quality Management Systems 5
J ISO 15223-1: Keep dry vs. humidity limitation label symbol Other Medical Device Related Standards 2
Q Weigh Scale Electronics - Suggested Storage Humidity and Temperature Manufacturing and Related Processes 7
S IEC 60601-1 Humidity Precondition Test IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S Methods to reduce relative humidity in Datacenter IEC 27001 - Information Security Management Systems (ISMS) 3
M Saturated salts issue - Humidity Chamber General Measurement Device and Calibration Topics 5
V Temperature/humidity requirements in internal metrology laboratory General Measurement Device and Calibration Topics 3
N ESD and Humidity Monitoring in Work Areas Manufacturing and Related Processes 1
Ajit Basrur MS Excel to study Correlation between Humidity & Differential Pressure Excel .xls Spreadsheet Templates and Tools 7
M How to Measure Paper considering Variation and Humidity General Measurement Device and Calibration Topics 10
D Sterile Medical Device Temperature and Humidity (Storage) Labeling Requirements EU Medical Device Regulations 5
Ron Rompen Help with Corrosion (Humidity) Testing Procedure - Sintered Metal Steel Parts Manufacturing and Related Processes 9
J Transformer Compliance Testing, overheat then humidity treat, then dielectric test? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Do I need to perform Gage R&R on a Digital Temperature & Humidity Gage Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 18
J Relative Humidity and Plastic Raw Material Moisture Manufacturing and Related Processes 4
G IEC 61010 Humidity Preconditioning and Condensation CE Marking (Conformité Européene) / CB Scheme 4
M Daily Temperature and Humidity Graph template Document Control Systems, Procedures, Forms and Templates 2
A MSA (Measurement Systems Analysis) for Humidity Machine Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
B Temperature/Humidity inside a store Human Factors and Ergonomics in Engineering 7
M Temperature and Humidity requirements for Raw Material Storage and Quality Labs IATF 16949 - Automotive Quality Systems Standard 13
R 5% to 95% Non-Condensing Humidity Storage Test Conditions and Failure Modes APQP and PPAP 2
D PCBA Lab Temperature & Humidity Requirements - Advice needed Manufacturing and Related Processes 7
M Cl 5.7 Humidity preconditioning required for all Medical Electrical Equipment? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H What is the ISO Class 7 cleanroom requirement on Temperature and Humidity ? Manufacturing and Related Processes 4
S ETO ISO 11135-1 & ISO 17665-2 - Purity and Quality of Steam (Humidity) Qualification and Validation (including 21 CFR Part 11) 3
T ISO 2230 (Guidelines for Storage of Rubber Products) - Temperature, Humidity & Light Other ISO and International Standards and European Regulations 8
K Desiccants & Humidity Control inside ocean going Shipping Containers Supplier Quality Assurance and other Supplier Issues 5
T Calibration Laboratory Temperature/Humidity goes out of tolerance due to Power Loss General Measurement Device and Calibration Topics 9
D Buying a Temperature and Humidity Device, and Calibration AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
D Will just Heat and Humidity be Adequate to Simulate EtO Sterilization? Design and Development of Products and Processes 2
R Inspection Humidity Control - Range for Mechanical Inspection Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
N Chilled Mirror Dew Point Meter for Humidity Measurement Measurement Uncertainty (MU) 2
R Does Humidity affect Electronic Instrument Calibration? General Measurement Device and Calibration Topics 16
D How to use temperature and humidity environmental control data? Statistical Analysis Tools, Techniques and SPC 18
B Analytical Lab Environmental Condition (Temperature & Humidity Control Limit) General Measurement Device and Calibration Topics 5
G Humidity Problem in our Laboratory General Measurement Device and Calibration Topics 12
S Temperature and Humidity Chamber Verification General Measurement Device and Calibration Topics 5
C Need source for temperature and humidity gage General Measurement Device and Calibration Topics 8
M Temperature/humidity monitoring process procedure Manufacturing and Related Processes 2
N Uncertainty calculations for RH (Relative Humidity) sensors Measurement Uncertainty (MU) 5
N Uncertainty for Humidity Probes - Temperature Effect from 0.5?C dew point = 1.3%rh Measurement Uncertainty (MU) 2
I Temperature and humidity storage requirements? US Food and Drug Administration (FDA) 6
J Controlling Temperature and Humidity in the CMM Room and does anyone use Artifacts? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
K Relative Humidity Requirement on Calibration Certificates General Measurement Device and Calibration Topics 5
E Humidity Requirements for an ISO 7 or ISO 8 Medical Device Cleanroom? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
A Relative Humidity + ISO 9001:2000 Audit of Maintenance Department ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 30

Similar threads

Top Bottom