Our (small) company does our own 60601-1 testing in-house. For the last couple of years our Notified Body has asked us to have our Test Report expert reviewed. This we have done by the test house arm of the NB. (It costs - but much less than an external test).
Up until now, when it comes to the requirement for humidity pre-conditioning (before leakage current and dielectric strength tests) I have marked it as N/A, stating that "there are no hygroscopic parts; no natural rubbers or sintered insulations and it is not for use in atmospheres where condensation may occur".
And this has been accepted.
However, our expert reviewer has now decided that that is not a strong enough argument. Has anyone got any useful "extra ammunition" in this argument?
He does seem to suggest a possibilty of using 'robust risk management' to justify the N/A status, but any suggestions would be most welcome.
Up until now, when it comes to the requirement for humidity pre-conditioning (before leakage current and dielectric strength tests) I have marked it as N/A, stating that "there are no hygroscopic parts; no natural rubbers or sintered insulations and it is not for use in atmospheres where condensation may occur".