We test the HF Surgical Unit . If suppose working voltage is 890Vrms Between live and applied part then as per the table 6 of 60601-1 we have applied test voltage 2 x (√2U+ 1 500) (approx : 6.3 KV). But EUT is not pass on 6.3 KV , so as per the particular standard 201.8.8.3
lf, during dielectric strength testing of solid insulation forming MOPP, a breakdown or flashover occurs at the Creepage Distance specified in
8.9 of the general standard and 201.8.5.1.2 of this standard, then the test shall be carried out on the components which provide MOPP, such as
transformers, relays, optocouplers or Creepage Distance on printed circuit boards.
So can we test on transformers, relays, optocouplers components? And how much KV we have applied.
Regards
Roshan
We test the HF Surgical Unit . If suppose working voltage is 890Vrms Between live and applied part then as per the table 6 of 60601-1 we have applied test voltage 2 x (√2U+ 1 500) (approx : 6.3 KV). But EUT is not pass on 6.3 KV , so as per the particular standard 201.8.8.3
lf, during dielectric strength testing of solid insulation forming MOPP, a breakdown or flashover occurs at the Creepage Distance specified in
8.9 of the general standard and 201.8.5.1.2 of this standard, then the test shall be carried out on the components which provide MOPP, such as
transformers, relays, optocouplers or Creepage Distance on printed circuit boards.
So can we test on transformers, relays, optocouplers components? And how much KV we have applied.