IATF 7.1.5.2.1 Calibration/verification records :Program/software verification

#1
Hello,

During IATF audit, the auditor asked about the moisture content equipment (verification of humidity percentage of plastic row material after drying), in this equipment every row material has its program, here he asked how we validate/verify the program: here we responded that we validate the program by the verification of the result of the test and compare it with the material datasheet, but also we mentioned that during calibration of equipment, to verify the equipment we start one program and perfrom the calibration , but there is no prove: no in calibration certificat, no any document,

so my question : how to verifiy the software or program of an equipment? how to respond to how did you validate the program ?

thank you guys :)
 
Elsmar Forum Sponsor

John C. Abnet

Teacher, sensei, kennari
Leader
Super Moderator
#3
Good day @ghanfari
Could you please share the exact verbatim nonconformance as written by the CB?

This will better help us reply to your situation. Thank you.

Be well
 
#4
good day,

here the nonconformance as written by the CB:

Statement of nonconformity
Calibration record control not fully effective.
Objective evidence
Evidences for the verification that the software or program version used for the control of the residual moisture applicable for the Mettler-Toledo
could not be submitted during the audit.
Justification for classification
List of used program is related to limited number of material, verification of the program with the applicable supplier recommendation led to
positve result.
All program parameters included in the soft war of the equipment (Temperature, time weight are conform to the specification).
No risk for the customer, limited risk for the quality system management of the client.
 

dwperron

Trusted Information Resource
#5
good day,

here the nonconformance as written by the CB:

Statement of nonconformity
Calibration record control not fully effective.
Objective evidence
Evidences for the verification that the software or program version used for the control of the residual moisture applicable for the Mettler-Toledo
could not be submitted during the audit.
Justification for classification
List of used program is related to limited number of material, verification of the program with the applicable supplier recommendation led to
positve result.
All program parameters included in the soft war of the equipment (Temperature, time weight are conform to the specification).
No risk for the customer, limited risk for the quality system management of the client.

I am seeing this more and more in ISO 17025 audits, where assessors are writing up labs for not having complete calibration records:

"6.4.13 Records shall be retained for equipment which can influence laboratory activities. The records
shall include the following, where applicable:
a) the identity of equipment, including software and firmware version...."

This sort of applies to your situation, and may be what the auditor was looking at, because there is the 16949 requirement for calibration records:
g) verification that the software version used for product and process control is as specified

I am guessing that he wants to see software / firmware revisions in your calibration records for the moisture analyzer, and that the revision matches what your process calls for it to be.
 

Sebastian

Trusted Information Resource
#6
It is IATF 16949 section 7.1.2.5.1 g), but i) is also interesting.

There is equipment software prepared by equipment manufacturer, which runs according to programs designed by you.
Equipment manufacturer sets a mask and program names shall comply with this mask.
You have to take this mask into account and create program naming "procedure" to control future program revisions.
Program name could look like "123456.000" and revision "123456.001"
Also it could be named "123456A" and revision "123456B".
Calibration records shall refer full name, including ".001, or "B".
This is about controlling program revisions in a same manner as document revisions are controlled, e.g. product drawings.
 
Last edited:
#7
hello, this is the NC:

Evidences for the verification that the software or program version used for the control of the residual moisture applicable for the Mettler-Toledo
could not be submitted during the audit.

it was IATF audit
 

John C. Abnet

Teacher, sensei, kennari
Leader
Super Moderator
#8
hello, this is the NC:

Evidences for the verification that the software or program version used for the control of the residual moisture applicable for the Mettler-Toledo
could not be submitted during the audit.

it was IATF audit
Although the NC is written oddly (poorly?)...i.e. "Calibration record control not fully effective."...from what i can glean from your posts, yes...this does indeed sound like a valid NC.

From your original post..." here he asked how we validate/verify the program ?"

Is that not a valid question? Your organization has a program that delivers as result. How do we know the program is giving the correct result unless there is some type of program validation?


If you are "validating" by simply comparing to the results of the material data sheet...
1- Do you have evidence that the equipment generating the material data sheet has been validated?
2- Assuming that your internal results don't always match the material data sheet EXACTLY, what is the allowable variance from the results the material data sheet provided? Is this documented?/justified?
3- Can your organization work with the provider of the material data sheet to have their equipment calibrator assist with collaborative testing ?
4- Can your organization work with the provider of the material data sheet to establish an interlaboratory collaborative testing plan?
5- Does the manufacture provide or recommend a collaborative testing service ?

Food for thought.

Hope this helps.
Be well.
 
Thread starter Similar threads Forum Replies Date
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 6
J Calibration/Verification Records (IATF 16949 7.1.5.2.1) IATF 16949 - Automotive Quality Systems Standard 5
E IATF 16949 Cl. 7.1.5.2.1 - Calibration and Verification Records Requirements IATF 16949 - Automotive Quality Systems Standard 1
J IATF 16949 Cl. 7.1.5.2.1 - Gauges (Measuring Jigs) Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
M Clarification on Calibration/Verification Records 7.1.5.2.1d (IATF 16949) General Measurement Device and Calibration Topics 11
B Calibration/Verification Records - IATF 16949 section 7.1.5.2.1 f) IATF 16949 - Automotive Quality Systems Standard 6
P IATF 16949 Cl. 7.1.5.2.1 "Calibration/Verification Records" Interpretation IATF 16949 - Automotive Quality Systems Standard 12
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
N IATF Calibration Lean in Manufacturing and Service Industries 5
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
Casana IATF 16949 7.1.5.3.2 External Laboratory - On Site Calibration IATF 16949 - Automotive Quality Systems Standard 8
N IATF 16949 7.1.5.3.2 External Laboratory - Calibration of Nikon machine in factory IATF 16949 - Automotive Quality Systems Standard 24
I IATF 16949, accredited logos, calibration capabilities, etc. IATF 16949 - Automotive Quality Systems Standard 17
B IATF 16949 7.1.5.3.2 FAQ #7 Audit Finding - External Calibration Laboratory IATF 16949 - Automotive Quality Systems Standard 7
sswaim IATF 16949 7.1.5.3.2 External Calibration Laboratory Requirelents IATF 16949 - Automotive Quality Systems Standard 15
J IATF 16949 Calibration Laboratory Scope IATF 16949 - Automotive Quality Systems Standard 22
M Producing IATF 16949 compliant calibration certificates IATF 16949 - Automotive Quality Systems Standard 24
B IATF 16949 - 7.1.5.3.1 Internal calibration laboratory requirements IATF 16949 - Automotive Quality Systems Standard 13
D IATF 16949 Clause 7.2.2 - On the job training IATF 16949 - Automotive Quality Systems Standard 2
G IATF Remote Location audit timing IATF 16949 - Automotive Quality Systems Standard 3
Sidney Vianna IATF 16949 News IATF Performance Complaint Management System (IATF CMS) within the IATF Database IATF 16949 - Automotive Quality Systems Standard 4
malasuerte Resourcing for ISO and or IATF? IATF 16949 - Automotive Quality Systems Standard 0
L IATF 16949 8.3 Exclusions Manufacturing and Related Processes 3
M IATF external audit NC closure IATF 16949 - Automotive Quality Systems Standard 4
D IATF 16949 7.3.1 Awareness- Supplemental IATF 16949 - Automotive Quality Systems Standard 2
W IATF 9.2.2.1 Internal Audit how to determine risk IATF 16949 - Automotive Quality Systems Standard 12
Q IATF 16949 8.4.2.4 Supplier Monitoring IATF 16949 - Automotive Quality Systems Standard 2
J Implementation, establishment and certification of IATF 16949. IATF 16949 - Automotive Quality Systems Standard 6
D IATF mandatory KPI? IATF 16949 - Automotive Quality Systems Standard 15
xfngrs 3 year audit cycle IATF 16949 IATF 16949 - Automotive Quality Systems Standard 10
D IATF 16949 Process Map IATF 16949 - Automotive Quality Systems Standard 19
Vader22 IATF 16949 extended manufacturing site help IATF 16949 - Automotive Quality Systems Standard 4
J PROBLEM SOLVING - IATF and GM CSR IATF 16949 - Automotive Quality Systems Standard 0
D IATF 16949 Certification and Letter of Conformance expiry (Help!!!) IATF 16949 - Automotive Quality Systems Standard 8
D AS9100 VS IATF 16949 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
Ashland78 Need IATF 16949 ISO Gap Analysis Excel File Internal Auditing 3
J Internal Audit Schedule IATF Internal Auditing 4
D IATF Requirements if remote plant has no interface with OEMs who are part of IATF? IATF 16949 - Automotive Quality Systems Standard 6
G IATF NC - Contingency Planning IATF 16949 - Automotive Quality Systems Standard 2
T Justification of "Design Excluded" in IATF 16949 Design and Development of Products and Processes 5
Sidney Vianna Job Opening in the USA - IATF 16949 Lead Auditor Position Job Openings, Consulting and Employment Opportunities 0
Ron Rompen IATF 16949 8.5.1.4 IATF 16949 - Automotive Quality Systems Standard 7
I Work Instruction Update- IATF Impact IATF 16949 - Automotive Quality Systems Standard 6
P IATF 16949:2016 Letter of conformity IATF 16949 - Automotive Quality Systems Standard 5
D IATF PROCESSES IATF 16949 - Automotive Quality Systems Standard 2
D IATF 16949 SI 10, External non-accredited lab IATF 16949 - Automotive Quality Systems Standard 4
A Control plan for IATF - example? FMEA and Control Plans 1
J IATF CB Auditor questions Internal Auditor Competence IATF 16949 - Automotive Quality Systems Standard 19
A IATF MSA study Type 1 General Measurement Device and Calibration Topics 7