Identification of "Essential Design Output"

Aphel

Involved In Discussions
#1
Hello all!

I'm really a little bit in trouble regarding one requirement in the QSR.

In the design control chapter "design output" you find the requirement that states you have to identfy those design outputs, that are essential for proper functioning of the device...[...]

Ok, it is clear what the understanding of the QSR of "essential design output" is. I have no problem with that.

But i at the moment i have no idea for a practical solution regarding the method of identifying "essential design output".
From the practical perspective - what are your experiences, what are your solutions for that topic? How do you perform this identification?

Thank you very much for your guesses and for your help!

Best regards,
Aphel
 
Elsmar Forum Sponsor

Aphel

Involved In Discussions
#3
hello all,

no more guesses on that topic?

how do you perform this exercise - identifying essential design output???


Once again - thanks a lot in advance...

Regards,
Aphel
 

Statistical Steven

Statistician
Staff member
Super Moderator
#4
hello all,

no more guesses on that topic?

how do you perform this exercise - identifying essential design output???


Once again - thanks a lot in advance...

Regards,
Aphel
It is documented in the DHR. Typically you flow from the URS to Design Inputs and then Design Outputs. During design review (also documented) the discussion and rationale are documented on how the output tests the input requirement based on the URS.
 
C

cturbo02

#5
Hey Aphel, did you get an answer on this? The answers posted did not seem to answer your question, in a way that I could understand it anyways. I am asking myself the same question on what makes the most sense here. IE - wondering if identification means being identified on a print somewhere, or that and differentiating the feature in the print as an essential item (or not). Please email me at cturbo02 (@gmail.com) reply (just registered and not sure if I will get a notification email if you reply).
 
#6
Hey Aphel, did you get an answer on this? The answers posted did not seem to answer your question, in a way that I could understand it anyways. I am asking myself the same question on what makes the most sense here. IE - wondering if identification means being identified on a print somewhere, or that and differentiating the feature in the print as an essential item (or not). Please email me at cturbo02 (@gmail.com) reply (just registered and not sure if I will get a notification email if you reply).

:bigwave:

This has been answered twice and both answers are absolutely correct.

If you don't understand; ask and that way you can help everybody.
 

Stijloor

Staff member
Super Moderator
#7
Hey Aphel, did you get an answer on this? The answers posted did not seem to answer your question, in a way that I could understand it anyways. I am asking myself the same question on what makes the most sense here. IE - wondering if identification means being identified on a print somewhere, or that and differentiating the feature in the print as an essential item (or not). Please email me at cturbo02 (@gmail.com) reply (just registered and not sure if I will get a notification email if you reply).
Welcome to The Cove Forums!! :bigwave: :bigwave:

This is a forum and we like the posts and information to be publicly shared and available for the benefit for all our Members. Unless absolutely necessary, we typically do not engage in private email conversations.

Thank you for your understanding.

Again, welcome to The Cove!! :agree1:

Stijloor. Forum Moderator.
 
C

cturbo02

#8
:bigwave:

This has been answered twice and both answers are absolutely correct.

If you don't understand; ask and that way you can help everybody.
Oh J0anne, here we go then, I appreciate any light you can shed on this for me.

Questions
Firstly, what does URS mean, I did not see that defined anywhere

The part of the regulation I am struggling with the most, just to be clear, is: "Design output procedures shall...ensure that those design outputs that are essential for the proper functioning of the device are identified.

What does identified mean? ie - does it mean that the essential outputs must just be on a design output (such as a drawing and/or procedure), or does it mean that the essential outputs must be on a design output and differentiated from the other non-essential design outputs.

Also what is the purpose of this essential design output identification. How is it used for the benefit of monitoring those essential outputs? I don't see that connected up anywhere else in the regs. If it doesn't effect anything else in a positive way, what is the purpose of it? I found a potential answer to that and I was gonna post a link to it, but I guess I can't do that, so I will just post the relevant info below and italicize it.

The Commonalities of Risk Management and Design Verification, Design Validation and Process Validation

Written by: Edwin Bills - Bilanx Consulting

OK, now the device is designed and you just have to get through that pesky Design Verification and
Design Validation process before you put the device on the market, and get some income after all the
design effort. But, really what is necessary to get through these steps?

Essential Design Outputs
The FDA has a requirement in Design Outputs (21 CFR 820.30 d) to “ensure that those design
outputs that are essential for the proper functioning of the device are identified”. In discussions with
Office of Compliance, personnel indicated that risk management processes are used to identify these
essential outputs. The outputs are those identified as Risk Control measures. So the requirement is
to identify in design output documents those risk controls for the purpose of communicating which are
essential to the functioning and safety of the device.

Why are the essential outputs important to this discussion? The identified Essential Outputs (EDO)
are those items that will be included in the Design Verification, Process Validation and Design
Validation activities. Each of the Risk Control measures (EDO) must have a record in the Risk
Management File that indicates the EDO has been verified for implementation, that the estimated risk
following implementation is verified as correct, and in the process to produce the device, the EDO has
been reviewed for Process Validation requirements.

Manufacturers need a mechanism to communicate the EDO to those internal and external
organizations that have an impact on the EDO so that the product produced is controlled within its
specifications. If the resources are focused on those truly important characteristics, then we should
have fewer recalls and Corrective Actions. Usually component specifications and drawings are
marked with information that indicates EDO and are accompanied by information communicated to
those responsible, in both internal and external organizations for those items to indicate that these
items have a major impact on the device safety.​

In this document the author instructs that this identification of essential outputs should be done using risk in the risk documentation, and the measures put in place to ensure the essential design outputs would also be managed within the risk process. However, the author of that article lost me a bit midway, i.e, didn't really understand how "the outputs are those identified as Risk Control measures". Seems like to me that the output is an output, and what you do to mitigate risk associated with it is the risk control measure.

Lastly, earlier in this thread, Statistical Steven stated that the DHR, which means Device History Record to me, can be used for this purpose. That didn't make sense to me either, as there could be hundreds or even thousands of essential design outputs, and not all would be captured on the DHR. For example, if the dielectric strength of an insulation is essential, the dielectric strength on incoming material may be guaranteed on the certificate of conformance from a supplier, but that dielectric strength wouldn't necessarily be tested on each unit and documented for each unit or batch on the DHR.

Thanks in advance for your help, sorry for the long post, and sorry to everyone and the forum operator for posting my email address, didn't know that was such a no-no but I understand now that could lead to private discussions that don't benefit anybody else.
 
C

cturbo02

#9
Also, after some googling I am guessing that URS means "User Requirements Specifications", which I am guessing is the document that covers the user needs, design inputs, etc. If that is the case, then I have the following questions on that method of documenting the identification of the essential design outputs.
1) Is that the only method to document the identification of the essential design outputs?
2) Does that necessarily mean that all the design outputs would need to be documented in the URS?
3) If it is done in the URS, does that mean it doesn't need to be done in the documentation actually used in the manufacture of the device (such as prints and manufacturing instructions). If so, I don't really see the point of doing the work in the URS, as it seems like the work going into the delineation of the essential design outputs would not have a positive effect on highlighting the importance of the essential elements in production, which I thought was maybe the point of all this.
 
C

cturbo02

#10
In retrospect, after reading the original post and comparing it to my posts, I think I have veered away a bit from the original question of how "the identification of essential outputs" should/can take place, and gone into how the "documentation of the identification of essential outputs" should/can take place. I am not as worried about distinguishing between essential and non-essential design outputs, I agree with earlier posts that that can be done via design review, risk assessment, etc. Sorry if I veered away to much in my posts, and I appreciate any help that can be given on my question.
 
Thread starter Similar threads Forum Replies Date
I Hazard Identification and Naming ISO 14971 - Medical Device Risk Management 10
C ISO 45001 6.1.2.1 Hazard Identification Occupational Health & Safety Management Standards 1
J 1.11 Preliminary Identification of Special Product and Process Characteristics APQP and PPAP 4
T ISO 9001 8.5.2. - Identification and traceability to Identify Outputs - Services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
K Identification of hazards and Risk file IEC 62366 - Medical Device Usability Engineering 7
M Medical Device Identification & Codes - Article 27 Requirements questions EU Medical Device Regulations 1
T Non conformance product identification and traceability 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
qualprod Monitoring of lead time - Good KPI identification? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 14
qualprod Controlled sticker for product identification? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 15
Watchcat Identification of Test Sample in Test Reports? Design and Development of Products and Processes 22
B Marking of Medical Electrical equipment and accessories - Cl. 7.2.2 "Identification" and Cl. 7.2.4 "Accessories" IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M Informational EU – Unique Device Identification (UDI) System – FAQs Medical Device and FDA Regulations and Standards News 0
S ISO 14971 Risk Management - Questions for Hazard identification ISO 14971 - Medical Device Risk Management 2
Z Two Payment Identification Number (PIN) for the same order in DFUF website 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
K Identification and Traceability with an ERP system - Barcode Labels? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
M MDR Annex IX Chapter I, 2.2 (c) - Device identification procedures during manufacture. EU Medical Device Regulations 1
M Informational USFDA final guidance – Unique Device Identification: Convenience Kits Medical Device and FDA Regulations and Standards News 0
Stefan Mundt ISO 9001:2015 - 8.5.2 Identification and Traceability Manufacturing and Related Processes 14
S Looking for procedure on UDI (Unique Device Identification) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
S UDI (Unique Device Identification) Requirements for Remanufactured devices 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
B Quality Management System documentation identification Document Control Systems, Procedures, Forms and Templates 11
K Document Numbering (Identification) System Document Control Systems, Procedures, Forms and Templates 10
N Requirements for the identification and traceability of demo product for sales force US Food and Drug Administration (FDA) 1
M RFID (Radio Frequency Identification) Registration in Europe and in MENA countries EU Medical Device Regulations 1
qualprod Identification of Training Needs = People Performance? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
H European Pharmacopoeia First Identification Requirements Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
Q Identification of gage blocks General Measurement Device and Calibration Topics 8
DeeDeeM IATF16949, clause 8.5.2.1 Identification and traceability-supplemental IATF 16949 - Automotive Quality Systems Standard 1
DeeDeeM IATF 16949 - Clause 8.5.2 Identification and Traceability IATF 16949 - Automotive Quality Systems Standard 7
Q ISO 9001 Cl. 8.5.2 and 8.5.4 - Identification in Products ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Measurement Equipment - Identification of Calibration Status General Measurement Device and Calibration Topics 25
J Customer Identification and Traceability in Manufacturing Plans Manufacturing and Related Processes 5
M Risk Identification and Risk Assessment for any Process - Is it necessary? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 22
Edward Reesor UDI (Unique Device Identification): HIBCC or GS1? ISO 13485:2016 - Medical Device Quality Management Systems 39
R Identification of Medical Devices in MDD 93/42 Certificate EU Medical Device Regulations 2
L Managing Finance Processes - Identification of Sub Processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
M Informational Is Identification of Risks and Opportunities required for QMS Processes? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 96
dubrizo Initial Supplier Identification, Review and Controls ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
H UDI (Unique Device Identification) Requirements for IVD Software EU Medical Device Regulations 2
A Receiving Goods Inwards - Identification Records and Data - Quality, Legal and Other Evidence 8
Pmarszal UDI (Unique Device Identification) Transition Period - Packaging Labeling Other US Medical Device Regulations 5
Q RFID (radio frequency identification) registration for Medical Device Other Medical Device Regulations World-Wide 6
B Class II Medical Device UDI (Unique Device Identification) Question(s) Other US Medical Device Regulations 8
A Is Risk Identification and Treatment a Process? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 25
D 820.120 UDI (Unique Device Identification) Labeling Verification Requirements Other US Medical Device Regulations 11
M Identification of Glass Instruments and Measurement Devices General Measurement Device and Calibration Topics 2
A Identification of Customer Property: Customer-Supplied Thumb Drives & Ext Hard Drives ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
Z Failure Mode Identification in PFMEA according to AIAG FMEA Rev.4 FMEA and Control Plans 6
M Reagent Status Identification - 7.4.3 Verification of Purchased Product ISO 13485:2016 - Medical Device Quality Management Systems 6
Gman2 Identification of Raw Material being used In-Process ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5