IEC 18562-1 report

#1
Dear All,

we're manufacturers of mesh and compressor nebulizers and preparing for re-certification on MDR.
Our devices need to confirm with ISO 18562 standards.
We are wondering if we could prepare a report for IEC 18562-1 by ourselves, without laboratory services and how to do it. The other parts of the standards will be evaluated by labs.
Have you got any advice, what such a report should contain and how to prepare it? Are there any templates?

Thank you in advance!
 
Elsmar Forum Sponsor

planB

Super Moderator
#2
Yes, principally you could prepare a such a report on your own and only outsource the bare testing to qualified laboratories. Just be beware that ISO 18562-1 requires competence and experience per section 4.1:
The biocompatibility evaluation shall be planned, carried out and documented by knowledgeable and experienced professionals.
and per section 4.6:
All tests shall be conducted according to recognized current/valid best laboratory/quality practices and the data shall be evaluated by competent, informed professionals.
HTH,
 

Billy Milly

Involved In Discussions
#3
In general, when testing against a certain standard is done, testing lab should be accredited. If there is no accreditation, also testing procedures are scrutinized (validation, training...). It can be done but the NB review time can extend significantly due to additional reviews.
 
#4
Thank you!
Do you, or anyone here, know if there are any templates or guidance on how to prepare such a report on ISO 18562-1? What is mandatory information to be included?
 

planB

Super Moderator
#6
[...] any templates or guidance on how to prepare such a report on ISO 18562-1? [...]
You may want to have a look into ISO 18562-1, section 4.5, Biocompatibility evaluation plan, and this thread; the related FDA guidance on biocompatibility evaluations also contains guidance on Attachment E: Contents of a Test Report and Attachment F: Component and Device Documentation Examples.

I definitely do not want to curb your enthusiasm, but without previous exposure to biocompatibility evaluations, I would strongly recommend to seek professional support.
 
Thread starter Similar threads Forum Replies Date
D IEC 61000-3-2 applied at 120V and 100V IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
H IEC 62368-1 used in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Artificial hand (IEC 60601-1-2) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
L Differences between IEC 60601-1-2 and IEC TR 60601-4-2 US Food and Drug Administration (FDA) 1
A IEC 62366-1:2015+A1:2020 IEC 62366 - Medical Device Usability Engineering 3
L EN 62304:2006/AC:2008 vs IEC 62304:2006/Amd 1:2015 IEC 62304 - Medical Device Software Life Cycle Processes 1
A IEC 62366-1:2015+A1:2020 IEC 62366 - Medical Device Usability Engineering 2
H IEC 61010-1:2010/A1:2016 - Clause 10.5.3 - (b) - Interpretation CE Marking (Conformité Européene) / CB Scheme 6
H Regarding Double Insulation & Caution Symbol on Product according to IEC 61010-1 CE Marking (Conformité Européene) / CB Scheme 6
S PDF version of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P IEC 60601-2-25 requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Brexit IEC 61000-4-2 test point consideration IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
MaHoDie Is it possible to assign medical software to security class A (according to IEC 62304)? EU Medical Device Regulations 8
Richard Regalado Informational ISO/IEC 27001:2022 has been published IEC 27001 - Information Security Management Systems (ISMS) 0
F IEC 60601-2-54:2009 Accuracy of CURRENT TIME PRODUCT IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 62366 Hazard-related Use Scenario vs FDA Critical Task EU Medical Device Regulations 3
C Sample qualification to be IEC 60601-1 series tested IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H IEC 62368-1 Scope of Testing CE Marking (Conformité Européene) / CB Scheme 3
H IEC 61010-1 Clause No. 10.5 Resistance to Heat - Sub Clause 10.5.3 Insulating Material CE Marking (Conformité Européene) / CB Scheme 6
H Is testing according to IEC 60601-1 & IEC 60601-1-2 mandatory for a device being registered under MDR? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
I IEC 60812 or ISO 14971 for PFMEA? What should we use? ISO 14971 - Medical Device Risk Management 3
H Required explanation of sub clause 7.2.101.2 (Exception) of IEC 61010-2-020:2006 CE Marking (Conformité Européene) / CB Scheme 5
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
G History of IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
H Pollution Degree (IEC 60664-1) CE Marking (Conformité Européene) / CB Scheme 2
T FDA level of concern vs IEC 62304 safety classification - 510(k) exempt device Other US Medical Device Regulations 13
J IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Y Software to Manage IEC 62304 Traceability Requirement IEC 62304 - Medical Device Software Life Cycle Processes 3
J IEC 60601-1 Clause 7.9.3.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S Safety testing IEC 60601 series Medical Device and FDA Regulations and Standards News 2
H Regarding Usability Engineering File according to IEC 62366-1 CE Marking (Conformité Européene) / CB Scheme 2
F Battery overload in IEC 60601-1 vs 62133-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Mandatory tests for IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
D IEC 60601-1-2: Is EMC immunity testing required for a device without essential performance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 25
C Producing to Inactive ASTM and IEC Standards Manufacturing and Related Processes 11
K IEC Maximum Power Density for Muscle Stimulator IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E Clause 8.9.1.2 of AMD2:2020 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
H Regarding confusion of the validity of the European implementation of IEC standards Other Medical Device Related Standards 1
P IEC 60601-1-2 RF wireless technology IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Should I buy IEC 62304:2006, IEC 62304:2006/AMD 1:2015 or both? EU Medical Device Regulations 1
A IEC 60601-2-8 cl.201.10.1.2.101 and cl.201.10.1.2.104 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
G Risk Management for IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 15
J IEC 60601 Applied Part - Probes intended to be covered by sheaths IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M How to apply IEC 62304 Problem Resolution Process for bugs discovered prior to release i.e. during initial development? IEC 62304 - Medical Device Software Life Cycle Processes 2
S IEC 60601-1-11 Requirements related to Storage and Distribution IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
shimonv IEC 60601-1-2:2014+A1:2020 (Ed. 4.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
T IEC 62304 : Risk control for SaMD IEC 62304 - Medical Device Software Life Cycle Processes 8
G IEC 60601 Compliance IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W IEC 60601-1 clause 8.8.3 - Dielectric Strength IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
L ISO/IEC 20000-6 Technical Areas IT (Information Technology) Service Management 2

Similar threads

Top Bottom