I am frustrated with our ISO 13485 auditor relating to the storage/distribution requirements for a FDA Class 2 Hearing Aids.

We have identified the values for temperature, humidity, and pressure in the Instructions for Use derived from the section 4.2.2 of the IEC 60601-1-11 standard and tested per the V&V testing to demonstrate that the product complies with the requirements of the standard. The instructions for use indicate the permissible environmental conditions of storage and transport of the medical device after the device has been removed from its protective packaging and subsequently between uses and does not apply for the device PRIOR TO SALE.

The Hearing Aids do not have any storage/transport requirements prior to sale and the auditor not familiar with the IEC standards is applying these "between uses" scenario to our distribution centers and transportation.

We explained him that the product has no requirements and is not stated in the label and labeling - and these artefacts have been cleared by FDA. But this individual is steadfast and uncompromising.

Could you please provide me guidance on how to address this issue?
 
Elsmar Forum Sponsor

Peter Selvey

Leader
Super Moderator
#2
Not sure I understand the issue completely.

Since the device passed the tests in 4.2.2 unpackaged, it should be OK for all forms of storage/transport. The (reasonable) assumption here is that a test unpackaged is more severe than packaged, and therefore a single test unpackaged would be representative of both cases.

In this case, a manufacturer would normally state in the IFU and mark on the packaging a single set of conditions for "storage/transport" which covers both cases (before and after removal from packaging).

Assuming this is correct, then I guess the issue is with the controls applied to make sure they stay within these limits?

My opinion would be that the conditions meet in 4.2.2 are fairly extreme so no special controls are needed. Only if a part is sensitive e.g. limited to +40°C or 0°C or <80%RH then there would be procedures to monitor to ensure the temperature/RH stays within the allowable range. But the values in 4.2.2 is so wide that it's not important. Is that the issue?

Or is it labelling? Or lack of testing with the packaging?
 

SGquality

Quite Involved in Discussions
#3
Peter, sorry for not explaining clearly.

I believe the IEC 60601-1-11 requirements apply between the uses (at the Owner end) and not relating to prior to sale, correct?

There are no specific storage/transport requirements for the product and therefore not on the label/labeling that was cleared by FDA. The auditor is extending the IFU requirements to prior to sale and wanting us to map the temp, humidity and pressure requirements.

How do we justify?
 

Peter Selvey

Leader
Super Moderator
#4
Yes, IEC 60601-1-11 is actually for after removal from packaging.

Otherwise, normal "pre-use" transport and storage is covered by IEC 60601-1 Clause 7.2.17 and 7.9.3.1, which requires the range of permissible environmental conditions to be marked on the packaging and disclosed in the IFU. Maybe, the FDA/auditor are quoting the wrong standard?

In theory there will always be limits. However, I also understand that often the limits are way above those encountered in real world, so it's pointless to disclose them.

Unfortunately, IEC 60601-1 3rd edition does not allow for that, it says the range always needs to marked/declared in the IFU, there is no "if appropriate" preface in the requirement.

In practice for a home use device, it would be the same as the range specified for IEC 60601-1-11, Clause 4.2.2.

Note that in IEC 60601-1 2nd edition this was written in a different way and could be interpreted as needing this marking only if "special measures" are required. So if this is an older, legacy device, original tested to 2nd ed, upgraded to 3rd, this subtle change might not have been picked up.

Again we are looking at three issues here:
- labelling (package/IFU)
- testing (proof that the device can handle the range)
- controls (e.g. making sure the warehouse, delivery truck do not exceed the limit)

Which one(s) was the FDA/auditor raising an issue with?
 

SGquality

Quite Involved in Discussions
#5
Thanks for the detailed explanation here, Peter.

Unfortunately this was not the FDA but the ISO 13485 auditor who firstly, did not have a clue on what IEC 60601-1-11 is all about but he is getting fixated on the storage/transport conditions because it is in the Owner's Guide (aka IFU). We had the ISO 13485 audit for our Distribution Centers and he is trying to extend the requirements from the Owner's Guide to the storage/transportation prior to sale. And yes, the auditor wants us to map our distribution centers and monitor during transportation these "broad ranges" to make sure the warehouse, delivery truck do not exceed the limit. How do I support this requirement when there is no requirement on the product?

I tried to convey multiple times that the requirements from IFU is not "product" requirements and therefore not on the label/labeling but he is not getting the point :frust:
 
Last edited:

Peter Selvey

Leader
Super Moderator
#6
I agree with your conclusion (controls are not required), but not the method.

The IFU range limits need to be disclosed for both before and after use. You can't get around this. It's not possible to say "there are no limits". But that's not the really issue.

The issue here is whether controls are reasonable (e.g. mapping/monitoring/actions).

I used to be an ISO 13485 auditor but it's been a while and I don't have the latest edition of the standard. Even so, I'd be very confident that the standard has an "as appropriate" approach of handling this situation. Which is to say, if the range is limited and could reasonably be exceeded without controls, then controls are appropriate. Usually for this kind of thing, if you check the exact wording in the standard there should be the term "as appropriate" or similar.

In this case the range is so wide (-25°C to +70°C, water vapour pressure 50hPa) that controls would be, in my opinion, a waste of time. So, not appropriate.

Auditors are required to quote a specific standard/clause when raising a non-conformity. You should check what is the precise clause they are referring to and find out if it has an "as appropriate" or similar wording.

Note that the auditor cannot refer to IEC 60601-1-11 or IEC 60601-1 because neither of those standard have anything to do with controls that may or may not be applied in actual warehousing and distribution. If the auditor is quoting these standards as evidence that controls are required, then they are off their tree. So to speak.

However, take care. Auditors are not always that bad. This could be a case of "shooting yourself in the foot". If a company procedure says that controls are required based on the IFU range, then the auditor can raise a valid non-conformity if there are no controls. The problem is not IFU range or the auditor, but the company procedure. The company procedure needs to be revised to incorporate the "as appropriate" terminology.
 
Thread starter Similar threads Forum Replies Date
P IEC 60601-2-25 requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
B Transformer Requirements - IEC 60601-1 vs. IEC 61558-2-15 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
rezayatmand IEC 60601-2-18 Medical electrical equipment - Part 2-18: Particular requirements for the basic safety and essential performance of endoscopic equipmen IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-1 - MOPP and MOOP Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
D IEC 60601-1-11 - Home Care Device Minimal requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J FDA regulations vs. IEC 60601-1 Labeling Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
E IEC 60601 Main requirements for the certification of a device with a Li-ion batt IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
J IEC 60601-2-2 - EMC Requirements - Surgical instrument manufacturer IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
M Ground stud requirements on our medical device in IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S IEC 60601 Main requirements for Lithium Ion Battery Powered Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M IEC 60601-1 and Active Current Protection Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A IEC 60601 Creepage and Clearance Requirements on PCB IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S IEC 60601-1 - Label Physical Requirements for Disposable (Single Use) Medical Devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M 15.3.3 Ball Impact Test requirements in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
L Differences in the requirements between IEC 60601-1 3rd ed. and EN 60601-1:2006 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K Interpretation of IEC 60601-1-2 Electromagnetic Compatibility – Requirements & tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
K Understanding IEC 60601-2-68 requirements ISO 13485:2016 - Medical Device Quality Management Systems 1
F Earthing system (IT, TT or TN) and IEC 60601 Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J What are the IEC 60601-1-2 requirements for Computed Radiography Devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C EMC Test Report Requirements to IEC 60601-1-2:2009 ISO 13485:2016 - Medical Device Quality Management Systems 4
M IEC 60601-1: 2012 - Subclause 11.3 - Constructional Requirements for Fire Enclosures IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
I Device Can't Pass IEC 60601-2 ESD Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601 Third Edition Labeling Requirements - Section 5.2.2.1 (g) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M IEC 60601 Compliance requirements for NON- ME Equpment/Systems IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T How to fulfill IEC 60601-1-2 EMC Testing Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
H IEC 60601-2-24 & ISO 8536-8 - Conflicting Requirements? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T IEC 60601 3rd Labeling Requirements and Software GUI IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
D Patient Leakage ISO 7176-14 vs. IEC 60601-1:2005 Reporting Requirements Other US Medical Device Regulations 1
A IEC 60601-1 SIP/SIO Isolation and Touch current requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-2-49 - Touch Current (Enclosure Leakage Current) requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M MEEE Expected Service Life - IEC 60601-1 Clause 4.4 Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
P IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
H IEC 60601-1 Requirements vs. UL 60601-1 Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
N Requirements for PCB Component Replacements in an IEC 60601-1 Compliant Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
J Requirements to Ship Medical Devices to the EU - IEC 60601-1 3rd. Ed. IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
R Power Input Marking requirements in IEC 60601 and IEC 60950 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
P IEC 60601-1 Clause 11.1.3 Winding Temperature Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601 Ground Bond Test Requirements for a Plastic Enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
R Class II Medical Device Marking - Different IEC 60950 & IEC 60601 Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 17
J EMC (Electromagnetic Compatibility) Requirements - IEC 60601 3rd edition Clause 17 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Rework Devices placed on the market - IEC 60601-1 Requirements 3rd edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
Z CB Certificate and IEC 60601-1 Test Report requirements - Korea Product Registration Other Medical Device Regulations World-Wide 6
J Marking (Labeling) Requirements on Medical Equipment per IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
T Charging Cradle to charge our Lithium Ion Batteries - IEC 60601 Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
D CSA Marking Requirements - IEC 60601-1 and CSA 622.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M Brazilian Requirements for Medical Device Production Testing to IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Current Products on the Market and IEC 60601 3rd Edition Safety Re-Test Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
D IEC 60601-2-18 (IEC 601-2-18) Particular Requirements for Endoscopic Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
R IEC 60601-1 - RS232 Connector Requirements - Female Connectors question IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
C Component Requirements - IEC EN 60601-1 3rd ed. and latest IEC 60950-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2

Similar threads

Top Bottom