SBS - The Best Value in QMS software

IEC 60601-1:2005/AMD2:2020, Why this standard version is 3.0?

IEC 60601-1:2005/AMD2:2020, Why this standard version is 3.0?

  • 3.0

    Votes: 0 0.0%
  • 3.2

    Votes: 0 0.0%

  • Total voters
    0
#1
Hello,

I have a question for the version of IEC 60601-1 series.
In IEC website, the version updates according to the below rule:
1615882412656.png

According the upper rules, since one amendment is applied, I think the version of IEC 60601-1:2005/AMD2:2020 have to be 3.2, but that standard's version is 3,0.
1615883071622.png

Please let me know what point i missed
 
Elsmar Forum Sponsor

Benjamin Weber

Trusted Information Resource
#2
AMD2:2020 itself relates to Ed. 3.0 (the main version) and contains only the changes, it is not the whole "new edition". The consolitated version (CSV) is Ed. 3.2 and contains covers the whole main standard contents with the changes of AMD1:2012 and AMD2:2020. You can see this also in the list right in the row under AMD2:2020.

Yet: It is confusing ;-)
 
#3
AMD2:2020 itself relates to Ed. 3.0 (the main version) and contains only the changes, it is not the whole "new edition". The consolitated version (CSV) is Ed. 3.2 and contains covers the whole main standard contents with the changes of AMD1:2012 and AMD2:2020. You can see this also in the list right in the row under AMD2:2020.

Yet: It is confusing ;-)
Thank you for your kindly reply!
Now i can understand clearly :D
 
Thread starter Similar threads Forum Replies Date
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K IEC 60601-1:2012 - Impact Assessment for changes from IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D Patient Leakage ISO 7176-14 vs. IEC 60601-1:2005 Reporting Requirements Other US Medical Device Regulations 1
S Separation Devices as required in 16.5 of IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R How to decide a suitable contact time from table 23 of IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
P Understanding IEC 60601-1:2005, Clause 8.1 b) - Short Circuiting Double Insulation IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
R Risk Assessment Template related to IEC 60601-1:2005 ISO 14971 - Medical Device Risk Management 4
E Humidity Pre-Conditioning IEC 60601-1 (2005) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
J IECEE Guidance Document for ISO 14971 Risk Analysis - IEC EN 60601-1:2005 3rd edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC/AAMI/ANSI 60601-1:2005 Checklist IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
D IEC 60601-1:2005 - Medical device standardization - Seeking information IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
L Top opening of enclsoure against fire in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C EN 45502 vs IEC 60601-1-8 Other Medical Device Related Standards 0
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-1 1988 - Device developed in 2012 with standard of 1988 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Different methods of achievement of the isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2

Similar threads