This is the first time my company is designing an examination chair to the 3rd edition. As such there are many more safety criteria and tests with which we have to comply. My question revolves around clause 9.2.2.2 for acceptable gaps. Can someone please explain to me how a footrest on an electrical moving chair can meet the acceptable gap requirement for a foot and still meet human factor considerations? Our 3rd party inspector says that the footrest must be >100 mm from the ground to comply. Really? What purpose will the footrest serve then? Or is that the point? :confused:
I look forward to any insight!!
Elsmar Forum Sponsor
Mark Meer
Trusted Information Resource
#2
It sounds like, as you say, the inspector's expectations are contrary to the intended use of the footrest. As such, I would site 9.2.2.1 which states:
"If implementation of the above protective measures would be inconsistent with the INTENDED USE..., control of the relevant motion shall comply with 9.2.2.6."
(9.2.2.6. being specific controls for speed of movement).
Regardless of what you argue, it'll likely stem from your Risk Management file. If you haven't done so already, add a risk-evaluation line-item in your risk file to demonstrate that you've considered trapping hazards...
L
leroybrown
#3
Thanks for the help! I have another question now. I'm only familiar with the third edition requirements. Are the second edition requirements regarding trapping zones comparable? I'm trying to establish a baseline for comparing the trapping zone requirements between editions as additional trapping zones, according to our third party inspector, have been identified on our new chair.
P
PaulGr
#4
Second edition did not contain a list of acceptable gaps to determine if a trapping zone would lead to a hazard so your approach would probably be difficult.
Peter Selvey
Staff member
Super Moderator
#5
Clause 9.2.2.1 has a quad option "or" statement, i.e. four possible solutions, with the supplementary 5th solution if those are not practical (as Mark Meer already mentioned). Study each option carefully before turning to risk management.
You could be tricked (and the inspector also) into thinking that a trapping zone is not allowed, this is not true, keeping the distances outside the limits in Table 20 is just one of the 5 options.
If you do turn to risk management, take care to check the situation carefully and also offer an alternate solution (i.e. revised criteria, don't just say it is "not applicable").
L
leroybrown
#6
:thanks: Via a very thorough risk evaluation, additional info provided to the operator and some design modification, our exam chair passed the requirement.
I will now work on beefing up our risk management process to incorporate more of the 60601-1 requirements. This will definitely minimize the stress level for our next product submission in relation to testing.
Thread starter Similar threads Forum Replies Date
T List of Tests - IEC 61010-1, 3rd edition and IEC 60601-1, 3.1 edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
O Standards used for IEC 60601-1-2 3rd ed. Vs. 4th ed. CE Marking (Conformité Européene) / CB Scheme 1
L Differences in the requirements between IEC 60601-1 3rd ed. and EN 60601-1:2006 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
L IEC 60601-1 Amendment 1 3rd Edition and Brazil ABNT NBR IEC 60601-1:2010 IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
M Reducing cost of IEC 60601 3rd edition testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
B Did UL 60601-1 change when the IEC 60601-1 standard changed to the 3rd edition? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H Recognition of and Transition to IEC 60601-1 3rd Edition in Korea IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
T IEC 60601 3rd Labeling Requirements and Software GUI IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E FDA Implementation Date for IEC 60601 - 3rd Edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
N IEC 60601-1 , 3rd edition and the FDA - Special 510k to my updated Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 40
J Requirements to Ship Medical Devices to the EU - IEC 60601-1 3rd. Ed. IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S Are you struggling with your IEC 60601-1 3rd edition testing lab? IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
S IEC 60601-1 3rd Gap analysis template TO BUY - Does anyone know where? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N IEC 60601-1 3rd ed vs. IEC 60601-1 2nd ed Gap Analysis IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
T IEC 62366 included in 3rd ed IEC 60601? IEC 62366 - Medical Device Usability Engineering 1
A Essential Performance in IEC 3rd edition of the 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R IEC 62133 to satisfy 60601-1 3rd edition - What happen to 60086-4 Other Medical Device Related Standards 18
J EMC (Electromagnetic Compatibility) Requirements - IEC 60601 3rd edition Clause 17 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
thisby_ IEC 60601-1 3rd edition Transition Period - US vs. Europe US Food and Drug Administration (FDA) 5
M Rework Devices placed on the market - IEC 60601-1 Requirements 3rd edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
T IEC 60601-2-27 - Comparison of 2nd to 3rd edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
W Is there an IEC 60601 3rd Edition 1st Amendment? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D Relationship between IEC 60601-1 3rd Ed and IEC 60601-2-10 IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
B IEC 60601-1 3rd Edition adoption - I cant find this standard on the ISO page Reliability Analysis - Predictions, Testing and Standards 9
R Switching Power Supplies Certified to IEC 60601 3rd Edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
B IEC 60601-1 2nd Clause 17 and 3rd edition - Electrical Isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Current Products on the Market and IEC 60601 3rd Edition Safety Re-Test Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
N ISO 14971 vs. IEC 60601-1 3rd Ed. - What to use for the RM evaluation? ISO 14971 - Medical Device Risk Management 17
M 60601-1 3rd Edition Compliance with Particular Standards & Equivalency IEC/EN stds EU Medical Device Regulations 6
E Retrofitting Risk Management for IEC EN 60601 3rd Edition Compliance IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
C Component Requirements - IEC EN 60601-1 3rd ed. and latest IEC 60950-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Acceptance of IEC EN 60601-1 3rd edition in China China Medical Device Regulations 8
J IECEE Guidance Document for ISO 14971 Risk Analysis - IEC EN 60601-1:2005 3rd edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M UL (Underwriters Laboratories) Publishes IEC 60601 3rd edition FAQs IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
H IEC 60601-1 3rd Edition Certification - New Products Only? IEC 60601 - Medical Electrical Equipment Safety Standards Series 25
B IEC 60601 3rd Edition - Test Report Form UL uses for testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
J IEC EN 60601 3rd edition CF vs. CFD IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C Relationship of IEC 60601-1-6:2006 vs. IEC 60601-1 3rd Edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A IEC 60601-1 3rd Edition Recognition in Canada IEC 60601 - Medical Electrical Equipment Safety Standards Series 15
T Harmonization date of IEC 60601-1 , 3rd Edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 135
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K IEC 60601-1:2005/AMD2:2020, Why this standard version is 3.0? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Similar threads
Top Bottom