manlongcheung
MCLong
The clause states:
"6.4.1 * ALARM SYSTEM delays
If the sum of the maximum ALARM CONDITION DELAY plus the maximum ALARM SIGNAL
GENERATION DELAY is greater than 10 s, then the statistics of each distribution or statistics of
the distribution of the sum shall be disclosed in the instructions for use."
What sort of statistics is expected to be disclosed in the IFU?
Currently, we display our total alarm delay time by saying it's less than a certain time. e.g. < 30 s
Would this capture the intent? or would you need a mean and standard deviation for each alarm times, which may be difficult to find out given the complexity of environment the device can experience.
Thank you very much!
Alarm condition | Priority | Delay |
Leak Detected | Medium | < 30s |