IEC 60601-1 and 60601-1-2 - Not Applicable for My Medical Device
rcjpedro
Hi All!
Does any of you kind sirs possess a checklist or template were I could technically and in detail justify that 60601-1 and 60601-1-2 are not applicable to one device?
The mains basis for NA of the device would be:
As per 60601 Definitions:
3 * Terminology and definitions
For the purposes of this document, the following terms and definitions apply.
3.63
* MEDICAL ELECTRICAL EQUIPMENT
ME EQUIPMENT
electrical equipment having an APPLIED PART or
transferring energy to or from the PATIENT or
detecting such energy transfer to or from the PATIENT
and which is:
a) provided with not more than one connection to a particular SUPPLY MAINS; and
b) intended by its MANUFACTURER to be used:
1) in the diagnosis, treatment, or monitoring of a PATIENT; or
2) for compensation or alleviation of disease, injury or disability
My Device:
does not have “an APPLIED PART” and
does not “transfers energy to or from the PATIENT or detects such energy transfer to or from the PATIENT”
Of course all is consistent with Classification Rule and Clinical Evaluation and TF.
Thank you in advance
Mark Meer
Very curious:
Your device is electrical, and intended for medical purposes, but doesn't contact the patient, nor detect or transfer energy?
What does it do?
rcjpedro
Hi Mark,
Its classified as per
RULE 11
All active devices intended to administer and/or remove medicines, body liquids
or other substances to or from the body are in Class IIa,
Thanks for your interest
rcjpedro
Hi Marcelo and yodon,
It is a Medical Device it has already been on the market but had to be withdrawn because I couldn't produce enough evidence for not falling under the scope of 60601
Marcelo
It's really difficult to help if you don't say what the device is, including it's intended use.
Ronen E
Problem Solver
Moderator
It can be an IVD or any other laboratory equipment (which are not medical electrical equipment under 60601).
I tend to agree with you, however I once worked on an active device that could be classified as an IVD (at least one of its variants) and had an argument with the development team - I argued that it was exempt from 60601 based on "no applied part" etc. and they argued that it had an applied part because otherwise the user wasn't able to handle it. The device was non-rechargeable battery powered, and via a rather convoluted arrangement was electrically in touch with a mains transformer/charger that could, in some rare situations, be part of the set-up. The charger wasn't part of the medical device, nor an accessory to it.
Marcelo
Well, I tried to make a general remark only. In fact there's IVDs that can be medical electrical equipment depending on the intended use (one common example are insulin strip analyzers, if the device requires that you cut your finger in a strip not attached to the device, it's not a medical electrical equipment. If the strip needs to be attached to the device, it's a medical electrical equipment).
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Question on whether IEC 60601-2-62 standard is applied IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Transformer Requirements - IEC 60601-1 vs. IEC 61558-2-15 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T Clearance and Creepage according to Subclause 8.9.1 for MOPP IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601-1 and IEC 60601-2-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3