IEC 60601-1 and 60601-1-2 - Not Applicable for My Medical Device

R

rcjpedro

#1
Hi All!

Does any of you kind sirs possess a checklist or template were I could technically and in detail justify that 60601-1 and 60601-1-2 are not applicable to one device?

The mains basis for NA of the device would be:

As per 60601 Definitions:

3 * Terminology and definitions
For the purposes of this document, the following terms and definitions apply.
(…)
3.63
* MEDICAL ELECTRICAL EQUIPMENT
ME EQUIPMENT
electrical equipment having an APPLIED PART or
transferring energy to or from the PATIENT or
detecting such energy transfer to or from the PATIENT
and which is:
a) provided with not more than one connection to a particular SUPPLY MAINS; and
b) intended by its MANUFACTURER to be used:
1) in the diagnosis, treatment, or monitoring of a PATIENT; or
2) for compensation or alleviation of disease, injury or disability

My Device:

does not have “an APPLIED PART” and
does not “transfers energy to or from the PATIENT or detects such energy transfer to or from the PATIENT”

Of course all is consistent with Classification Rule and Clinical Evaluation and TF.

Thank you in advance
 
Elsmar Forum Sponsor

Mark Meer

Trusted Information Resource
#2
Very curious:

Your device is electrical, and intended for medical purposes, but doesn't contact the patient, nor detect or transfer energy?

What does it do?
 
R

rcjpedro

#5
Hi Mark,

Its classified as per

RULE 11
All active devices intended to administer and/or remove medicines, body liquids
or other substances to or from the body are in Class IIa,

Thanks for your interest
 
R

rcjpedro

#6
Hi Marcelo and yodon,

It is a Medical Device it has already been on the market but had to be withdrawn because I couldn't produce enough evidence for not falling under the scope of 60601

Thanks for your interest
 

Marcelo

Inactive Registered Visitor
#8
Hi Marcelo and yodon,

It is a Medical Device it has already been on the market but had to be withdrawn because I couldn't produce enough evidence for not falling under the scope of 60601

Thanks for your interest
It's really difficult to help if you don't say what the device is, including it's intended use.
 
Last edited:

Ronen E

Problem Solver
Moderator
#9
It can be an IVD or any other laboratory equipment (which are not medical electrical equipment under 60601).
I tend to agree with you, however I once worked on an active device that could be classified as an IVD (at least one of its variants) and had an argument with the development team - I argued that it was exempt from 60601 based on "no applied part" etc. and they argued that it had an applied part because otherwise the user wasn't able to handle it. The device was non-rechargeable battery powered, and via a rather convoluted arrangement was electrically in touch with a mains transformer/charger that could, in some rare situations, be part of the set-up. The charger wasn't part of the medical device, nor an accessory to it.
 

Marcelo

Inactive Registered Visitor
#10
I tend to agree with you, however I once worked on an active device that could be classified as an IVD (at least one of its variants) and had an argument with the development team - I argued that it was exempt from 60601 based on "no applied part" etc. and they argued that it had an applied part because otherwise the user wasn't able to handle it. The device was non-rechargeable battery powered, and via a rather convoluted arrangement was electrically in touch with a mains transformer/charger that could, in some rare situations, be part of the set-up. The charger wasn't part of the medical device, nor an accessory to it.
Well, I tried to make a general remark only. In fact there's IVDs that can be medical electrical equipment depending on the intended use (one common example are insulin strip analyzers, if the device requires that you cut your finger in a strip not attached to the device, it's not a medical electrical equipment. If the strip needs to be attached to the device, it's a medical electrical equipment).
 
Last edited:
Thread starter Similar threads Forum Replies Date
F IEC 60601-2-54:2009 Accuracy of CURRENT TIME PRODUCT IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C Sample qualification to be IEC 60601-1 series tested IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H Is testing according to IEC 60601-1 & IEC 60601-1-2 mandatory for a device being registered under MDR? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
G History of IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J IEC 60601-1 Clause 7.9.3.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S Safety testing IEC 60601 series Medical Device and FDA Regulations and Standards News 2
F Battery overload in IEC 60601-1 vs 62133-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Mandatory tests for IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
D IEC 60601-1-2: Is EMC immunity testing required for a device without essential performance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 25
E Clause 8.9.1.2 of AMD2:2020 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P IEC 60601-1-2 RF wireless technology IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 60601-2-8 cl.201.10.1.2.101 and cl.201.10.1.2.104 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
G Risk Management for IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 15
J IEC 60601 Applied Part - Probes intended to be covered by sheaths IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S IEC 60601-1-11 Requirements related to Storage and Distribution IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
shimonv IEC 60601-1-2:2014+A1:2020 (Ed. 4.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
G IEC 60601 Compliance IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W IEC 60601-1 clause 8.8.3 - Dielectric Strength IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
B IEC 60601-1 Clause 14.8 - Architecture IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R IEC 60601-2-57 light source classification IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
L Top opening of enclsoure against fire in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C EN 45502 vs IEC 60601-1-8 Other Medical Device Related Standards 0
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K IEC 60601-1:2005/AMD2:2020, Why this standard version is 3.0? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7

Similar threads

Top Bottom